IEC 60601-1 cl.15.4.2.1 - Tubular Heating Element (Protection against Overheating)

A

antonoia

#1
Hi All!!
I am working on reviewing the design of a class I device that has a tubular heating element powered with live and neutral trough a relais, the scope is to verify if complies with 60601-1. I found the clause above specified that states: "equipment that incorporates tubular heating elements shall have protection against overheating in both leads where a conductive connection to earth could result in overheating".
I did not fully understand the reason of this requiremnent, from what it is written I believe that if not both leads are disconnected currents can still pass trough the live conductor to earth trough the tubular. If this is the reason why is this requirement of disconnecting both leads applied only to tubular heating element and not also to some other components? Hope someone can help to understand and to verify if this will be applicable also in my case. Many Thanks!!
 
Elsmar Forum Sponsor

3dBUnderThe Limit

Starting to get Involved
#2
Re: IEC 60601-1 cl.15.4.2.1 - Tubular Heating Element (Protection against Overheating

I think that your analysis is essentially correct.
For a heater this fault would lead to uncontrolled heating. As the heating element could operate with a normal current, any overcurrent protection will not be activated, that would be the difference between this type of component compared with an power supply where the same type of error would trip the fuse.

This is a known failure mode and I would say that disconnecting both leads are necessary
 
A

antonoia

#3
Re: IEC 60601-1 cl.15.4.2.1 - Tubular Heating Element (Protection against Overheating

Thank you 3DBunderthelimit, fortunately in 10 years of working with tubular heating element I did not have occasion of experiencing such failures, perhaps I should try to simulate in a bench test. Are you aware of other devices where this requirement can be still used?
 

3dBUnderThe Limit

Starting to get Involved
#4
Re: IEC 60601-1 cl.15.4.2.1 - Tubular Heating Element (Protection against Overheating

For a heater you would get uncontrolled heating for a device in standby mode. A similar scenario in a motor could lead to uncontrolled movement. If this could lead to a hazardous situation, disconnection of both live and neutral could be an appropriate mitigation.

So I think that it's appropriate to consider this type of failure for other devices in the risk analysis
 
A

antonoia

#5
Re: IEC 60601-1 cl.15.4.2.1 - Tubular Heating Element (Protection against Overheating

Thank you again, I was thinking if a RCD breaker is installed on the building , would not be this enough to mitigate this risk? In case of current on the earth conductor this will disconnect all the poles, or do you think it will be better to have a RCD installed on the device? will this release me from the complying from the requirements of this famous clause? Regards!
 

3dBUnderThe Limit

Starting to get Involved
#6
Re: IEC 60601-1 cl.15.4.2.1 - Tubular Heating Element (Protection against Overheating

I feel that I?m getting in to a grey zone, where I don?t know enough about the system to give a good answer. But I will give you my opinion anyway :)
To me, using an RCD is adding much more complexity in the system than needed. Most likely a requirement of regular testing of the RCD is needed in the documentation. Personally I would opt for a design where I have a switching element in both live and neutral to the heater.
 
Thread starter Similar threads Forum Replies Date
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 0
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Has anyone done an IEC 60601-1 gap analysis to IEC 60335? Medical Device and FDA Regulations and Standards News 4
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601-1 and IEC 60601-2-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601-1-2 EMC Tables IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Bookmarking my security protected IEC 60601-1 .pdf file IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E ECG device without ECG REPORTS - IEC 60601-2-25: 2011 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C IEC 60601-1-8, difference between table 4 and annex D IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Scope of application for IEC 60601-1-11 Medical electrical equipment — Part 1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601 7.1.3 Durability of Markings - Methylated Spirit or Ethanol? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C ABS plastic : which IEC 60601-1 tests to redo? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Q Isolation/Insulation Diagrams - IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-2-2 - Defib-proof test amendment to base standard IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
pashah Looking for Usability File Template acc. IEC 62366-1 and IEC 60601-1-6 IEC 62366 - Medical Device Usability Engineering 4
P Testing lab in India IEC 60601-2-25 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Electromagnetic Compatibility Test IEC 60601-1-2: 2014 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
rezayatmand IEC 60601-2-18 Medical electrical equipment - Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipmen IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Fire hazard classification - Clause in IEC 60601 about gauze and electricity ISO 14971 - Medical Device Risk Management 0
K IEC 60601-1 Ed. 3.1 b:2012: Replaced IEC 60601-1 Ed. 3.1 en:2012? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Impact test on glass touch controller IEC 60601 - Medical Electrical Equipment Safety Standards Series 2

Similar threads

Top Bottom