IEC 60601-1, Class 1 or Class 2 on emergency power system

U

uspe1

#1
Hi,

I need your help on this. I am working on the emergency power system (battery back up system)of the dialysis machine. Our dialysis machine is class 1, when it is operated on the main power supply. I need to know when there is an interruption in the main power supply and machine is powered by emergency power system. Now the machine is still on class 1 or class 2? We don't have earth protection in emergency power system, so it means class 2 or is there any option to create that.

Class I

Equipment protection against electric shock by (Earthed) additional protection to basic insulation through means of connecting exposed conductive parts to the protective Earth in the fixed wiring of the installation.

Class II

Also referred to as Double Insulated. Equipment protection against electric shock by additional protection to basic insulation through means of supplementary insulation are provided, there being no provision for the connection of exposed metalwork of the equipment to a protective conductor and no reliance upon precautions to be taken in the fixed wiring of the installation.

regards,
Usman
 
Elsmar Forum Sponsor
Thread starter Similar threads Forum Replies Date
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J IEC 60601-1 Electrical Safety testing - Class 1 ME externally powered - Earth resistance testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
JoCam Medical Device that is class II electrical and IEC 60601-1-11 EU Medical Device Regulations 5
W Is IEC 60601-1-2 compliance/testing required for battery powered Class I Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Class II Medical Device Marking - Different IEC 60950 & IEC 60601 Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 17
M IEC 60601 - List of required tests - Medical Device Class IB IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
C Sample qualification to be IEC 60601-1 series tested IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
G History of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J IEC 60601-1 Clause 7.9.3.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 2
F Battery overload in IEC 60601-1 vs 62133-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Mandatory tests for IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
E Clause 8.9.1.2 of AMD2:2020 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P IEC 60601-1-2 RF wireless technology IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-2-8 cl.201.10.1.2.101 and cl.201.10.1.2.104 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G Risk Management for IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
J IEC 60601 Applied Part - Probes intended to be covered by sheaths IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S IEC 60601-1-11 Requirements related to Storage and Distribution IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
shimonv IEC 60601-1-2:2014+A1:2020 (Ed. 4.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
G IEC 60601 Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W IEC 60601-1 clause 8.8.3 - Dielectric Strength IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
B IEC 60601-1 Clause 14.8 - Architecture IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R IEC 60601-2-57 light source classification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
L Top opening of enclsoure against fire in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C EN 45502 vs IEC 60601-1-8 Other Medical Device Related Standards 0
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1

Similar threads

Top Bottom