IEC 60601-1 Clause 11.1.3 Winding Temperature Requirements
Pads38
Moderator
Clause 11.1.3 (60601-1 ed 3) gives the details of temperature measurements to be taken and gives details of the change of resistance method for windings.
This includes a constant of 234.5 along with the resistance readings and room temperature compensation.
The question is where does the 234.5 come from and what does it represent?
(the temp coefficient for copper is 0.394% per degree C so this doesn't seem to be the whole answer).
Peter Selvey
I've just pulled out my thesis from 1998 on this one. It appears the formula uses the co-efficient at 0degC rather than 20degC.
If the 20C value was used, the formula would appear as:
dT = (R2-R1)/R1 * 251.9 - 20 -(t2 -t1)
Your instinct might say that since the co-efficient changes so much between 0 and 20C, a first order (linear) approximation is not suitable. Or to put it another way, the above formula might give different results.
In actual case, in the region of interest copper is quite linear, but the linear approximation does not pass through the point (0,0). If it did pass through (0,0) the slope and temperature co-efficient would be the same. Since the line does not go through (0,0), the "co-efficient" (change in resistance / degC) is not the slope, and changes a lot with temperature. But as long as the slope does not change, the linear approximation holds and we can estimate temperature accurately.
Pads38
Moderator
Thanks Peter.
I suppose it doesn't really change how the test works, but it does help to resolve one more bit of 'head scratching'.
M IEC 60601-1-8 clause 6.4.1 (statistics of the distribution of the sum of the maximum ALARM CONDITION + SIGNAL GENERATION DELAY) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J IEC 60601-1 Clause 7.9.3.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E Clause 8.9.1.2 of AMD2:2020 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
W IEC 60601-1 clause 8.8.3 - Dielectric Strength IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
B IEC 60601-1 Clause 14.8 - Architecture IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
A Fire hazard classification - Clause in IEC 60601 about gauze and electricity ISO 14971 - Medical Device Risk Management 0
S ME system clause 16 of IEC 60601-1, 3.1 ed IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T What is PESS? Cited in Edition 3.1 of IEC 60601, clause 14.1 note 2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E ME (Medical Equipment) Systems - IEC 60601-1 Clause 16.1 Interpretation IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Question on IEC 60601 clause 15.5.1.1 (transformer overheating) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L IEC 60601-1 3rd edition Clause 9.2.2 Trapping Zone (Acceptable Gaps) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E IEC 60601-1 Clause 14 Questions IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R How to understand the exception of IEC 60601 sub-clause 8.11.5 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M MEEE Expected Service Life - IEC 60601-1 Clause 4.4 Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
R How to calculate the Air Flow Rate in clause 201.12.4.4.105 of IEC 60601-2-16:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J EMC (Electromagnetic Compatibility) Requirements - IEC 60601 3rd edition Clause 17 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
I Dielectric Test for HF Surgical Equipment (IEC 60601-2-2 clause 201.8.5.1.2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601 Clause 7.1.3 - Durability of Markings - Methylated Spirits IEC 60601 - Medical Electrical Equipment Safety Standards Series 19
P Understanding IEC 60601-1:2005, Clause 8.1 b) - Short Circuiting Double Insulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
P IEC 60601 Clause 57.3 / 8.11.3.2 - PVC Flexible Cords IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
G Single Fault Condition for 3-Phase Medical System - IEC 60601 Clause 8.1b IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B IEC 60601-1 2nd Clause 17 and 3rd edition - Electrical Isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
F Defining the need to perform the Proximity magnetic fields immunity tests (IEC 61000-4-39) according to the IEC 60601-1-2 ed.4.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
O ESD Testing IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
E Meaning of ALARM PRESETS (subclause 6.5.1 of IEC 60601-1-8:2020) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Franky Zhang123 The Electrical Standards GB9706.1:2020 (same as IEC 60601-1:2012) etc. will be implemented on May 1st, 2023 in China. China Medical Device Regulations 0
H IEC 62368-1 used in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Artificial hand (IEC 60601-1-2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L Differences between IEC 60601-1-2 and IEC TR 60601-4-2 US Food and Drug Administration (FDA) 3
S PDF version of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P IEC 60601-2-25 requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C Sample qualification to be IEC 60601-1 series tested IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
G History of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 2
F Battery overload in IEC 60601-1 vs 62133-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Mandatory tests for IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
P IEC 60601-1-2 RF wireless technology IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-2-8 cl.201.10.1.2.101 and cl.201.10.1.2.104 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G Risk Management for IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
J IEC 60601 Applied Part - Probes intended to be covered by sheaths IEC 60601 - Medical Electrical Equipment Safety Standards Series 3