IEC 60601-1 Clause 7.9.3.1

#1
Dear All,
I am trying to get IEC-60601-1 standard Certification for Otoscopes. i am stuck at Clause 7.9.3.1 which says (All characteristics of ME Equipment - Range(s), accuracy, precision of displayed values (or where they can be found) )
remembering that the product that i am trying the certification is a simple ME Otoscope B/F Applied part, which function is to emits a Led light on the site of observation and given i clear vision to the observer so it don't not display any range of values or display any values.
please guide what should i do regarding this clause as its been a headache for me
 
Elsmar Forum Sponsor

Riya Gupta

Starting to get Involved
#3
Hi all,
I am looking for IEC 60601 compliance for my device. If i get compliance with IEC 60601-1-2 EMC tests, then do i also need to get tests done for IEC 60601-1.
kindly throw some light on this.

Thanks,
 

Peter Selvey

Leader
Super Moderator
#4
Yes, they are completely different standards, IEC 60601-1 is the general standard and covers things like marking, IFU, electrical safety, mechanical safety, temperatures and so on, while IEC 60601-1-2 is a "collateral" standard just for EMC, and there may be other applicable standards as well in the series.

Recommend to do a search for "IEC 60601 series" and you should find an overview of the standards.
 
Thread starter Similar threads Forum Replies Date
E Clause 8.9.1.2 of AMD2:2020 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
W IEC 60601-1 clause 8.8.3 - Dielectric Strength IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
B IEC 60601-1 Clause 14.8 - Architecture IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A Fire hazard classification - Clause in IEC 60601 about gauze and electricity ISO 14971 - Medical Device Risk Management 0
S ME system clause 16 of IEC 60601-1, 3.1 ed IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T What is PESS? Cited in Edition 3.1 of IEC 60601, clause 14.1 note 2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E ME (Medical Equipment) Systems - IEC 60601-1 Clause 16.1 Interpretation IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Question on IEC 60601 clause 15.5.1.1 (transformer overheating) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L IEC 60601-1 3rd edition Clause 9.2.2 Trapping Zone (Acceptable Gaps) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E IEC 60601-1 Clause 14 Questions IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R How to understand the exception of IEC 60601 sub-clause 8.11.5 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M MEEE Expected Service Life - IEC 60601-1 Clause 4.4 Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
R How to calculate the Air Flow Rate in clause 201.12.4.4.105 of IEC 60601-2-16:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P IEC 60601-1 Clause 11.1.3 Winding Temperature Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J EMC (Electromagnetic Compatibility) Requirements - IEC 60601 3rd edition Clause 17 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
I Dielectric Test for HF Surgical Equipment (IEC 60601-2-2 clause 201.8.5.1.2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601 Clause 7.1.3 - Durability of Markings - Methylated Spirits IEC 60601 - Medical Electrical Equipment Safety Standards Series 19
P Understanding IEC 60601-1:2005, Clause 8.1 b) - Short Circuiting Double Insulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
P IEC 60601 Clause 57.3 / 8.11.3.2 - PVC Flexible Cords IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
G Single Fault Condition for 3-Phase Medical System - IEC 60601 Clause 8.1b IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B IEC 60601-1 2nd Clause 17 and 3rd edition - Electrical Isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
H IEC 62368-1 used in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Artificial hand (IEC 60601-1-2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L Differences between IEC 60601-1-2 and IEC TR 60601-4-2 US Food and Drug Administration (FDA) 1
S PDF version of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P IEC 60601-2-25 requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C Sample qualification to be IEC 60601-1 series tested IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
G History of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 2
F Battery overload in IEC 60601-1 vs 62133-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Mandatory tests for IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
P IEC 60601-1-2 RF wireless technology IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-2-8 cl.201.10.1.2.101 and cl.201.10.1.2.104 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G Risk Management for IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
J IEC 60601 Applied Part - Probes intended to be covered by sheaths IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S IEC 60601-1-11 Requirements related to Storage and Distribution IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
shimonv IEC 60601-1-2:2014+A1:2020 (Ed. 4.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
G IEC 60601 Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-2-57 light source classification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
L Top opening of enclsoure against fire in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5

Similar threads

Top Bottom