IEC 60601-1 Impact test on glass touch controller
We are designing a portable medical device with a glass face which has two separate subareas: a) the display area and b) the touch controller/buttons area.
According to the IEC 60601-1 standard, the display area should be excluded from the 15.3.3 Impact test.
"The test is not applied to flat panel displays, to the platen glass of ME EQUIPMENT (for example
film scanners), or to cathode ray tubes (see 9.5.2)."
We also interpret that the "platen glass" term above includes the whole glass face of our device including the touch controller area.
Could you please confirm if the total glass face should be excluded from this test?
Thanks.
Marcelo
It's difficult to answer such questions without a knowledge of the device.
But please bear in mind that:
- Product standards usually have requirements to deal with specific hazards/risks;
- Tests are usually ways to verify implementation and efficacy of risk control measures;
- Standards are usually voluntary;
- What you have to comply with is usually the mandatory regulatory requirements.
In your case, I would ask - what is the problem (hazard/risk) here? What should I do to prevent the problem? What should I do to show that I included a control to prevent the problem?
Hello Marcelo,
Thank you for your response.
I understand that a risk-based approach shall be used in order to select the appropriate pass/fail verdict of the impact test.
Please let me rephrase the question.
If we ask a test-lab to perform the 60601-1 impact test, can we specifically request them to avoid hitting the glass area of the enclosure (display+control pannel) when they perform this test, as the standard states "The test is not applied [...] to the platen glass of ME EQUIPMENT"?
Independently of the outcome and risk.
Thanks.
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M 15.3.3 Ball Impact Test requirements in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K IEC 60601-1:2012 - Impact Assessment for changes from IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J IEC 60601 (Steel Ball Impact Test) Other ISO and International Standards and European Regulations 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 0
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601-1 and IEC 60601-2-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601-1-2 EMC Tables IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Bookmarking my security protected IEC 60601-1 .pdf file IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E ECG device without ECG REPORTS - IEC 60601-2-25: 2011 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C IEC 60601-1-8, difference between table 4 and annex D IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Scope of application for IEC 60601-1-11 Medical electrical equipment — Part 1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601 7.1.3 Durability of Markings - Methylated Spirit or Ethanol? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C ABS plastic : which IEC 60601-1 tests to redo? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Q Isolation/Insulation Diagrams - IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-2-2 - Defib-proof test amendment to base standard IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
pashah Looking for Usability File Template acc. IEC 62366-1 and IEC 60601-1-6 IEC 62366 - Medical Device Usability Engineering 4
P Testing lab in India IEC 60601-2-25 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Electromagnetic Compatibility Test IEC 60601-1-2: 2014 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7