IEC 60601-1 - Label Physical Requirements for Disposable (Single Use) Medical Devices

S

SmallD

We have a medical device accessory (say electrode) needed for a medical device to work. It has its own packaging (and label). The packaging has print (text/instructions).
Since this is a single use accessory, does the package print need to be fully compliant with section 7.1.3 Durability of markings?
We have problems with the print because of the material of the packaging.

Big thanks for any help!!
 

Pads38

Moderator
Re: 60601-1 label physical requirements for disposable (single use) devices

It may be useful to realise that the rub tests of 7.1.3 only applies to those markings as required by 7.2, 7.4, 7.5 and 7.6.

Each of those relates to markings/symbols/signs that are on the outside of the device itself. They are not requirements for packaging labelling / information.
 

Peter Selvey

Leader
Super Moderator
Re: IEC 60601-1 - Label Physical Requirements for Disposable (Single Use) Medical Dev

It's a good point and appears to be a mistake in the standard, something for the committee to consider.

There are a number of requirements in 7.2 which are permitted to be marked on the packaging if justified, but there is no mention of any exclusion for the durability tests.

Note that it makes sense that the marking is durable, the problem is not the requirement but the test being not representative of normal use (packaging would not normally be cleaned with alcohol). There should still be some minimum level of durability e.g. poor quality marking that easily flakes or rubs off with handling would not be OK even for packaging.

Maybe a water only test would be reasonable?

The other option would be to exclude accessories from the scope of evaluation. The standard is wrong for this point as well: accessories that are placed on the market as a separate medical device should be treated separately and only considered for the system related aspects. Non-system aspects such as marking should be outside of the scope of IEC 60601-1. These are covered by other standards.
 

Pads38

Moderator
Re: IEC 60601-1 - Label Physical Requirements for Disposable (Single Use) Medical Dev

We recently had our shipping cartons tested, to ASTM D4169-14. They came back rather "second hand".
If we did have required markings on the carton we could repeat the legibility test of 7.1.2

In the case of packaging, such as double peel pouches often used for sterile items, I can't think what tests you could perform to replace the (7.1.3) durability tests. Perhaps pre-printed pouches could survive the rub tests even though the test is not really relevant. But paper self adhesive labels would need something different.
 
1

13lla

Re: IEC 60601-1 - Label Physical Requirements for Disposable (Single Use) Medical Dev

Have this same question at our company. Is there an industry standard practice for the testing of (IEC 60601-1) 7.1.3. Durability of Markings on disposable ME accessory packaging?

Is it the conservative approach which is to apply the tests outlined in 7.1.3 a & b which appear to be designed with the intent to test permanently affixed labels to ME equipment?
 

nanotek

Registered
Hi,
We have the same problem, our product is an electrode single use, for intraoperative use that shall be used with a active ME for recording.
We applied the EN 45502-1 §10.3 even if it is not applicable:
The MARKINGS on the SALES PACKAGING of an ACTIVE IMPLANTABLE MEDICAL DEVICE shall be indelible.
Test: The package shall be placed so that the MARKINGS under test are uppermost and in a horizontal plane.
Ten millilitres of water shall be dispensed onto the centre of the area. After one minute, the MARKINGS shall be
wiped clear of surface water using a wet, soft cloth.
Compliance is checked by inspection to determine that all MARKINGS remain clearly legible. If the MARKINGS are
on a LABEL, the adhesive fixing the LABEL shall not have loosened and the LABEL shall not have become curled
at any edge.
 
Thread starter Similar threads Forum Replies Date
R IEC 60601-2-57 - Subclause 201.7.101.2 - Explanatory Label - LS Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
F Defining the need to perform the Proximity magnetic fields immunity tests (IEC 61000-4-39) according to the IEC 60601-1-2 ed.4.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601-1-8 clause 6.4.1 (statistics of the distribution of the sum of the maximum ALARM CONDITION + SIGNAL GENERATION DELAY) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
O ESD Testing IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
E Meaning of ALARM PRESETS (subclause 6.5.1 of IEC 60601-1-8:2020) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Franky Zhang123 The Electrical Standards GB9706.1:2020 (same as IEC 60601-1:2012) etc. will be implemented on May 1st, 2023 in China. China Medical Device Regulations 0
H IEC 62368-1 used in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Artificial hand (IEC 60601-1-2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L Differences between IEC 60601-1-2 and IEC TR 60601-4-2 US Food and Drug Administration (FDA) 3
S PDF version of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P IEC 60601-2-25 requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C Sample qualification to be IEC 60601-1 series tested IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
G History of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J IEC 60601-1 Clause 7.9.3.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 2
F Battery overload in IEC 60601-1 vs 62133-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Mandatory tests for IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
E Clause 8.9.1.2 of AMD2:2020 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P IEC 60601-1-2 RF wireless technology IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-2-8 cl.201.10.1.2.101 and cl.201.10.1.2.104 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G Risk Management for IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
J IEC 60601 Applied Part - Probes intended to be covered by sheaths IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S IEC 60601-1-11 Requirements related to Storage and Distribution IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
shimonv IEC 60601-1-2:2014+A1:2020 (Ed. 4.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
G IEC 60601 Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W IEC 60601-1 clause 8.8.3 - Dielectric Strength IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
B IEC 60601-1 Clause 14.8 - Architecture IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R IEC 60601-2-57 light source classification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
L Top opening of enclsoure against fire in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C EN 45502 vs IEC 60601-1-8 Other Medical Device Related Standards 0
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9

Similar threads

Top Bottom