IEC 60601-1 - Risk assessment to determine the liquid - 11.6.3
Tiago Alfenas
The equipment uses deionized water for cooling, we had not considered, but now let's consider, the question is how to determine the test to be done? volume, height?
Tks!
yodon
The standard lays out the testing:
A quantity of liquid is poured steadily on a point on the top of the ME EQUIPMENT. The type of liquid, volume, duration of the spill, and
location (point) are determined through analysis. Test conditions that simulate the worst case for spillage shall be documented in the RISK
MANAGEMENT FILE.
After these PROCEDURES, the ME EQUIPMENT is to pass the appropriate dielectric strength and LEAKAGE CURRENT tests and is to show no signs of wetting of uninsulated electrical parts or electrical insulation of parts that could result in the loss of BASIC SAFETY or ESSENTIAL PERFORMANCE in NORMAL CONDITION or in combination with a SINGLE FAULT CONDITION (based on a visual inspection).
You have to determine volume, duration, height based on your assessment of worst case conditions.
If you have a closed loop cooling system then your risk may be limited to leakage issues rather than spillage.
Unless, say, the user has to 'top-up' a reservoir? Spillage may occur then.
So, hazard = spillage of top-up fluid
possible severity to be tested - 5 litres deionised water from a jug 1 metre above unit (just a guess!)
probability - only done during commissioning, so once per units lifespan
Result - is Basic Safety or Essential Performance affected?
Doe the unit still pass the dielectric strength and leakage current tests of clause 8?
R IEC 60601-2-57 Risk Group 2 OHD IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Ed Panek Transition to IEC 60601 4th Edition - Risk Analysis and test submissions CE Marking (Conformité Européene) / CB Scheme 2
D IEC 60601-1 & Risk Management issue IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Experience of IEC OD-2004 2nd Ed., 60601-1 and Risk Management Files? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
R Risk Assessment Template related to IEC 60601-1:2005 ISO 14971 - Medical Device Risk Management 4
E Retrofitting Risk Management for IEC EN 60601 3rd Edition Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
J IECEE Guidance Document for ISO 14971 Risk Analysis - IEC EN 60601-1:2005 3rd edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Risk of fire in an Oxygen rich environment - Fire Prevention - IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
C New edition of IEC 60601-1 - Risk assessment to determine the liquid - 11.6.3 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2