Recently, the test house told me that the female RS232 connector mounted on our equipment is not recommended, a male RS232 connector which the pins can not be touched directly is more reasonable. However, I can not find any relevant requirement in IEC 60601-1 does not allow the female connector provided that the connector meets the clause 16 a)5) and 16 e).

Please kindly advise it, thanks.
 
Elsmar Forum Sponsor

Marcelo

Inactive Registered Visitor
#3
Maybe they were thinking about 56.3 e) (which does not seem the case).

Anyway, I would be in principle againt connectors with pins touchable without the use of a tool because they would have to pass all requirements for accessible parts.
 

Roland chung

Trusted Information Resource
#4
The female connector surely meets the requirement of 56.3 since such SELV connector complies with 17 g)4). It seems no technical reasons to reject female connector.
 
Last edited:

Marcelo

Inactive Registered Visitor
#5
Sorry, i tried to say that 56.3 c) is not applicable in this case (no conductive connection to a patient).

And, if it was applicable, it would not pass, because the pins can be directly touched.

Anyway, i think that maybe this was the wrong interpretation of your test house.

What was their justification for that? You mentioned reccomentation, does not seem a requirement to me.
 

Roland chung

Trusted Information Resource
#6
Yes, the test house did not use "MUST", but "strongly recommend". It seems they would not compromise. They also do not have reasonable explanation. One just felt the connector with pins can be touched without the aid of tool is not safety enough. Most products are employed male connector.
 
S

Shine Zhang

#7
I think it's irresponsible of saying "strongly recommend" without any reasonable explanations.

some explanations may come as below:

1.It reduces the probability of the operator touching the contacts of the male connector and the patient simultaneously. even if the saied construction complied with 16e.

2. avoiding unneccessary electro-static discharge on your device through such male connector, which may influence the reliability and quality of the product.

3. if it is a low-signal equipment, avoiding contacing internal circuit through such male connetor is reasonable or desirable.

4.....
 

Peter Selvey

Staff member
Super Moderator
#8
As mentioned, any connector located on the rear of the device is normally exempted from leakage and isolation requirements based on Clause 16e.

But, it may not be immediately apparent that this exemption is necessary for both male and female connectors, as most female connectors do not comply with isolation requirements such as 3.4mm creepage. If the exemption is not applied, the internal circuits must be treated as live and tested for double insulation (typically 500Vdc, 3.4mm).

So a test house that insists a male connector fails while a female connector passes would appear to have some more explaining to do.
 

Roland chung

Trusted Information Resource
#9
I am sorry for the editorial error. The correct statement shall be male RS232 connector mounted on our equipment. My question is if male connector can be accepted when such male connector meets clause 16 e)?

Actually, either male connector or female connector does not meet the 3.4 mm creepage. It is unreasonable to consider any SELV connector as hazardous live part since such connector has nothing to do with the patients.

Peter, I can not understand your point very well. Could you explain in detail?
 

Peter Selvey

Staff member
Super Moderator
#10
Just to clarify first: we may be mixing male and female. A male connector is normally the one that has directly accessible pins; a female connector is usually not accessible but the spacing is very small (e.g. ~1mm).

My thinking was that the test house "fails" the male connector because the pins are accessible. However, to be consistent they should also "fail" the female connectors, as a spacing of ~1mm is also not enough.

Actually the 2nd edition is very messy on this point. Strictly, both male and female connectors do not comply with Clause Clause 17g (see the compliance statement, 2nd paragraph, which allows shorting of any cr/cl below limits).

The exclusion under 16a)5) still requires basic insulation, which most connectors do not have (male or female). And the exclusion in 16e)1) requires a cover, which again most connectors do not have.

However, the principle is that if contact with the patient is unlikely, then they can be excluded. Most test labs understand this and quietly consider signal connectors excluded under 16e)1). This quiet exclusion should apply to both male and female connectors.

This area has been fixed in the 3rd edition. Clause 8.4.2 c) sets up the basic exclusion (using risk management), and specifically refers to contacts of connectors, whether male or female.

The flow from 8.4 to requirements in the standard is:
8.4 --> 8.7 (leakage currents), and
8.4 --> 8.5 (MOP) --> 8.8 (dielectric strength, cr/cl)

Therefore, once the basic exclusion is set up under 8.4.2 c), the part is exempt from all tests for leakage, dielectric strength and cr/cl.

So to summarize:

Male and female connectors formally fail the 2nd edition, but it is industry practice to accept under 16e;

Male and female connectors pass the 3rd edition, provided location is such that direct or indirect contact with the patient is unlikely (e.g. rear of the device).
 
Thread starter Similar threads Forum Replies Date
L Top opening of enclsoure against fire in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C EN 45502 vs IEC 60601-1-8 Other Medical Device Related Standards 0
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Has anyone done an IEC 60601-1 gap analysis to IEC 60335? Medical Device and FDA Regulations and Standards News 4
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0

Similar threads

Top Bottom