I'm curious if anyone here has any experience utilizing subclause 5.1 when obtaining IEC 60601-1 certification from a test lab. For those unfamiliar, the relevant section of sublcause 5.1 here reads:
A test need not be performed if analysis shows that the condition being tested has been adequately evaluated by other tests or methods.
We are looking to swap out a critical component in our device. Both the old and new critical components are themselves IEC 60601-1 certified. We went through those 60601-1 test lab reports to pull relevant test data out and make a technical comparison of behavioral difference that we feel--per 5.1--adequately demonstrates the change wouldn't result in us failing to continue to meet the specs called out in 60601-1 for the whole system. What I'm trying to get from the test lab is an explanation of why our evaluation was not adequate, but can't seem to get a straight answer. Is there something I'm missing?

I know that in 60601-1 test reports for certified medical device components, it commonly explicitly states that certain sections of 60601-1 need to be evaluated on the full system instead of just that component, but would an adequate evaluation per 5.1 meet that requirement? Ultimately I just want to understand how everything here ties together, or why our evaluation is not adequate.

Not expecting in-depth technical analysis here (obviously, since I'm not providing specifics), just want to know if anyone here has experience with utilizing 5.1.
 
Elsmar Forum Sponsor
#2
I worked for a test house before moving on to a notified body to audit medical device manufacturers. In the handful of cases clause 5.1 was leveraged successfully, most pertained to temperature tests of low power (<15W) components downstream of the 60601-1 ed 3.0 (or 3.1) Switch Mode Power Supply. Typically, it is quicker to simply retest (leakage, worst-case temp test for that component from the previous evaluation, the loading tests for suspended masses, and/or dielectric) than to get an exception via 5.1 signed off by reviewers, perhaps even chief engineer in the case of mechanical support systems or MAINS connected components.

All that said, it depends on the component, accessibility to the user/patient, and risk analysis. By definition, critical components are necessary for either safety or essential performance and are therefore subject to more scrutiny. If the new component performs differently than the original component, then the device may likewise perform in a manner that is less safe or that does not meet essential performance requirements. Perhaps the new component has an unidentified failure mode that the current risk management file does not account for.

If you have test reports from an accredited lab of the component in similar or worse-case scenarios, the test house may accept those in lieu of performing a new test. Most labs I have worked with adhere to some form of common sense engineering, even the more risk-averse ones.
I'm curious about the genesis of the question. Are you seeking to save on certification costs or reduce time to market?
 

DitchDigger

Starting to get Involved
#3
Well, you're right about it being quicker to simply re-test as opposed to work a 5.1 exception. Luckily I'm in a position where we have a few months until we'll be forced to switch to the new component.
 
Thread starter Similar threads Forum Replies Date
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E IEC 60601 subclause 8.9.1.12 - Earthed secondary circuits IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
R IEC 60601-2-57 - Subclause 201.7.101.2 - Explanatory Label - LS Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
kevinkre IEC EN 60601-2-25 subclause 201.12.4.106.2 Channel Crosstalk (ECG Multichannel Crosstalk) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A What kind of test is described in subclause 56.3(c) of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1: 2012 - Subclause 11.3 - Constructional Requirements for Fire Enclosures IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J IEC EN 60601-2-27 subclause 50.102.5 ECG Multichannel Crosstalk IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3

Similar threads

Top Bottom