Hello, i will prepare a 510k registration, but i'm not sure which tests are necessary to demonstrate the electric safety. The standard 60601-1 describes the electric safety for medcial devices.
The device is has only a small battery with less than 2 V and no external power. Can I easily explain that the device is safe, because no risk health exists? Or is there a specific standard for devices with low power/voltage?
Many thanks for all answers in advance!
 
Elsmar Forum Sponsor
S

SteveK

#2
I think you will find that IEC 60601-1 still applies to low voltage battery driven medical devices. It is just that a large majority of the clauses will be "not applicable" or be easy to state compliance. There is nothing to stop you doing what tests are required e.g. a rub test on markings (clause 7.1.3). Just because it is a low voltage battery device does not mean there are no risks (under ISO 14971) - consider the issues with lithium battery cells that have cropped up.:2cents:

Steve
 
Thread starter Similar threads Forum Replies Date
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Testing lab in India IEC 60601-2-25 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S CTS/CSE Waveforms Testing - Partial Recognition of IEC 60601-2-25 Standard by FDA IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K Is Compliance Testing of IEC 60601-2-63 mandatory for Europe - Dental X-Ray machine CE Marking (Conformité Européene) / CB Scheme 3
L IEC 60601-1-11 IP22 Compliance - Ingress protection testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S IEC 60601-1-2 Radiated E-Field Emission Testing - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S IEC 60601-1 Electrical safety testing - Dielectric Strength Test - 8.8.3 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1 Electrical Safety testing - Class 1 ME externally powered - Earth resistance testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Testing my ECG device as per IEC 60601-2-25 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
O EMC testing per IEC 60601-1-2 for considering new cables CE Marking (Conformité Européene) / CB Scheme 1
O EMC testing under RED and IEC 60601-1-2 CE Marking (Conformité Européene) / CB Scheme 1
D IEC 60601-2-44: 202.101 Immunity Testing of Essential Performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
K IEC 60601-1 Testing for Multiple Socket Outlet (MSO) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601-1-2 Testing Laboratories in the Middle East IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J IEC 60601 Medical Device Testing House Costs IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
T Testing Laboratories in Europe to IEC/EN 60601-2-47 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
M Parallel processing IEC 60601 testing and 510k submission 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D Radiated Immunity Testing as per IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Reducing cost of IEC 60601 3rd edition testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
T How get prepared for IEC 60601 series testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W Is IEC 60601-1-2 compliance/testing required for battery powered Class I Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-1 Cl. 14 - Report changes in Software after Testing to Test Lab? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
T How to fulfill IEC 60601-1-2 EMC Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S IEC 60601-1-2 EMC Testing Optional in US? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
Q Sample Size justification for IEC 60601-1 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J European Laboratories which can do IEC/EN 60601-2-50 Testing? (Close to Turkey) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K Identifying Required Testing to comply with IEC 60601 EU Medical Device Regulations 4
S Are you struggling with your IEC 60601-1 3rd edition testing lab? IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
S Issues with Dielectric Strength Testing - IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
J "Testing procedure: TMP", "WMT", "SMT" and "RMT" - IEC 60601-1 TRF nomenclature IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Current Limit in IEC 60601 Dielectric Strength Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601 Testing - Section 13.2.7 - Impairment of Cooling IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
M Brazilian Requirements for Medical Device Production Testing to IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S EN (IEC) 60601-1 ? Is in-house testing definitely OK? IEC 60601 - Medical Electrical Equipment Safety Standards Series 28
A Do HF accessories require EMC testing according to IEC 60601-1-2? IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
D Battery Powered Disposable Pump - What IEC 60601 testing is required, if any? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Does a 12V DC device need EMC (IEC 60601-1-2) Testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B IEC 60601 3rd Edition - Test Report Form UL uses for testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
E IEC/EN 60601 Testing and CE Marking for a Non Medical Device CE Marking (Conformité Européene) / CB Scheme 11
N Insulation Barrier Testing - According IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 22
J Clarification in UL Mark in Critical Component - IEC 60601 testing for CE mark in UL IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601-1 Testing required for low-voltage devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Predicate Device - Question on Biocompatibility and IEC 60601-1 Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
D Excessive Temperature Testing to IEC 60601-2-37 - Internal Probes IEC 60601 - Medical Electrical Equipment Safety Standards Series 20
D IEC 60601-2-37 - No Dielectric testing specifications? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5

Similar threads

Top Bottom