Our medical device company is currently in the process of creating a device which can be classified as a INTERNALLY POWERED ELECTRICAL EQUIPMENT. The device is powered by three 3V coin cell batteries which form part of the the ME Equipment.
I have read through the IEC 60601-1 (tried to atleast) and there is a plethora of electrical testing.
My initial reaction is that for such a low-voltage battery, the risks are minimal and the filters of IEC 60601 are not designed for this device and therefore is not relevant.
Any IEC 60601 experts out there who can guide me through this standard.
Is the electrical testing relevant for such a device? I have tried to find clauses in the standard which would dictate the minimum voltage for an Internally Powered Equipment that can be tested by this standard but haven't found any information.
Using a worst case scenario, imagine a medical device that is only three 3v coin cell batteries stacked on top of each other with leads sticking out of them. You can consider the leads sticking out as the APPLIED PARTS. Essentially a 9v battery which is being considered here as a medical device. Would any of testing in IEC 60601 help?
Any help would be greatly appreciated!.
Thank you
edited: Leads defined as APPLIED parts
Marcelo
As weird as it seems, IEC 60601-1 is still applicable to this kind of equipment.
As you noted, there´s no exemption for minimum voltage (and if there was, it would be in the scope).
Surely a lot os tests would not be applicable in different cases, but it´s also not very clear which ones would not.
Finally, what exactly is the problem? Standards are generally voluntary, so if you do not want to apply it, generally you can (but you surely would need a good justification).
Peter Selvey
For a device like this there remains some simple and reasonable requirements you should check:
- labelling (6) - external/internal & instructions for use
- battery insertion: what happens if reversed (56.something)?
- battery short: any problems (52.5.9)
There are some agencies that may go to town on patient isolation, leakage currents, dieletric strength but none of this is justified based on the fundamentals.
3Vdc can be potentially dangerous in the following condition:
- applied directly to a patient with low impedance
(e.g. metal/gel electrodes)
- applied for a long period
In this case "dc burns" may occur which are not very nice.
But all electrical current through a patient requires 2 points (in and out), the probability of these conditions occuring at both sites is non-existent for most 3V types of devices that I know of.
You are probably stuck sequentially checking all of the ~500 requirements from IEC 60601-1 just to be sure, but around 98% will be not applicable.
Peter
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Testing lab in India IEC 60601-2-25 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S CTS/CSE Waveforms Testing - Partial Recognition of IEC 60601-2-25 Standard by FDA IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K Is Compliance Testing of IEC 60601-2-63 mandatory for Europe - Dental X-Ray machine CE Marking (Conformité Européene) / CB Scheme 3
L IEC 60601-1-11 IP22 Compliance - Ingress protection testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S IEC 60601-1-2 Radiated E-Field Emission Testing - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S IEC 60601-1 Electrical safety testing - Dielectric Strength Test - 8.8.3 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1 Electrical Safety testing - Class 1 ME externally powered - Earth resistance testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Testing my ECG device as per IEC 60601-2-25 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
O EMC testing per IEC 60601-1-2 for considering new cables CE Marking (Conformité Européene) / CB Scheme 1
O EMC testing under RED and IEC 60601-1-2 CE Marking (Conformité Européene) / CB Scheme 1
D IEC 60601-2-44: 202.101 Immunity Testing of Essential Performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
K IEC 60601-1 Testing for Multiple Socket Outlet (MSO) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601-1-2 Testing Laboratories in the Middle East IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J IEC 60601 Medical Device Testing House Costs IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
T Testing Laboratories in Europe to IEC/EN 60601-2-47 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
M Parallel processing IEC 60601 testing and 510k submission 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D Radiated Immunity Testing as per IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Reducing cost of IEC 60601 3rd edition testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
T How get prepared for IEC 60601 series testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W Is IEC 60601-1-2 compliance/testing required for battery powered Class I Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-1 Cl. 14 - Report changes in Software after Testing to Test Lab? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
T How to fulfill IEC 60601-1-2 EMC Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S IEC 60601-1-2 EMC Testing Optional in US? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
Q Sample Size justification for IEC 60601-1 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 60601-1 testing for device with its own 2V battery power supply necessary? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
J European Laboratories which can do IEC/EN 60601-2-50 Testing? (Close to Turkey) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K Identifying Required Testing to comply with IEC 60601 EU Medical Device Regulations 4
S Are you struggling with your IEC 60601-1 3rd edition testing lab? IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
S Issues with Dielectric Strength Testing - IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
J "Testing procedure: TMP", "WMT", "SMT" and "RMT" - IEC 60601-1 TRF nomenclature IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Current Limit in IEC 60601 Dielectric Strength Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601 Testing - Section 13.2.7 - Impairment of Cooling IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
M Brazilian Requirements for Medical Device Production Testing to IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S EN (IEC) 60601-1 ? Is in-house testing definitely OK? IEC 60601 - Medical Electrical Equipment Safety Standards Series 28
A Do HF accessories require EMC testing according to IEC 60601-1-2? IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
D Battery Powered Disposable Pump - What IEC 60601 testing is required, if any? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Does a 12V DC device need EMC (IEC 60601-1-2) Testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B IEC 60601 3rd Edition - Test Report Form UL uses for testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
E IEC/EN 60601 Testing and CE Marking for a Non Medical Device CE Marking (Conformité Européene) / CB Scheme 11
N Insulation Barrier Testing - According IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 22
J Clarification in UL Mark in Critical Component - IEC 60601 testing for CE mark in UL IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Predicate Device - Question on Biocompatibility and IEC 60601-1 Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
D Excessive Temperature Testing to IEC 60601-2-37 - Internal Probes IEC 60601 - Medical Electrical Equipment Safety Standards Series 20
D IEC 60601-2-37 - No Dielectric testing specifications? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3