Dear all,

I have the latest german consolidated edition DIN EN 60601-2-10:2017 (incl. A1). In cl. 201.12.1.10" "Pulse parameters" it states a maximum allowed measurement error of only 1%. I have the previsous edition of the IEC version, where 10% measurement error were allowed.

In the change notices there is nothing mentioned regarding this change. Does anyone have the latest IEC version and can tell me if this is maybe onyl a mistake during translation to the german DIN EN version?
 
Elsmar Forum Sponsor

Pads38

Trusted Information Resource
#2
I have BS EN -2-10:2015 + A1:2016

from that:
...shall not deviate by more than ± 20 % when measured with a load resistance within the range specified in the ACCOMPANYING DOCUMENTS
Compliance is checked by measurement with an error not exceeding ± 10 %.
 

Peter Selvey

Staff member
Moderator
#4
It seems to be a double mistake. A general rule of thumb is to have measurement equipment that is 5 times more accurate than the specification. So, if the allowable limit for the equipment is ±20%, the reference equipment (in total) should be ±4%. Neither 10% or 1% make sense ....
 
Thread starter Similar threads Forum Replies Date
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B Has anyone done an IEC 60601-1 gap analysis to IEC 60335? Medical Device and FDA Regulations and Standards News 4
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601-1 and IEC 60601-2-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601-1-2 EMC Tables IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Bookmarking my security protected IEC 60601-1 .pdf file IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E ECG device without ECG REPORTS - IEC 60601-2-25: 2011 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C IEC 60601-1-8, difference between table 4 and annex D IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Scope of application for IEC 60601-1-11 Medical electrical equipment — Part 1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601 7.1.3 Durability of Markings - Methylated Spirit or Ethanol? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C ABS plastic : which IEC 60601-1 tests to redo? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Q Isolation/Insulation Diagrams - IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-2-2 - Defib-proof test amendment to base standard IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
pashah Looking for Usability File Template acc. IEC 62366-1 and IEC 60601-1-6 IEC 62366 - Medical Device Usability Engineering 4
P Testing lab in India IEC 60601-2-25 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Electromagnetic Compatibility Test IEC 60601-1-2: 2014 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
rezayatmand IEC 60601-2-18 Medical electrical equipment - Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipmen IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Fire hazard classification - Clause in IEC 60601 about gauze and electricity ISO 14971 - Medical Device Risk Management 0
K IEC 60601-1 Ed. 3.1 b:2012: Replaced IEC 60601-1 Ed. 3.1 en:2012? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Impact test on glass touch controller IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D IEC 60601 for a Software only Medical Device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C Is IEC 60601 applicable to our smart bed? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S CTS/CSE Waveforms Testing - Partial Recognition of IEC 60601-2-25 Standard by FDA IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S IEC 60601-1 - MOPP and MOOP Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-25 ECG 4 electrodes and derived leads IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
T IEC 60601-1 vs IEC 60601-1-2 4th Edition Other ISO and International Standards and European Regulations 1
P IEC 60601 11.2.2 - Oxygen Rich Environment sources of ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601 subclause 8.9.1.12 - Earthed secondary circuits IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
T Tests IEC 60601-1 - Grounding Resistance, Leakage Current and Dielectric Rigidity in an automated way IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J IEC 60601-1, ed 3.1 - Blue icon question IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S IEC 60601-1-2 Radiated Immunity Test - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Similar threads


















































Top Bottom