IEC 60601-2-18 Endoscopic Medical Electronic Equipment question

Roland chung

Trusted Information Resource
#1
The Annex J out of IEC 60601-2-18 gives an insulation example for endoscopic equipment. In the example 2, the number "12" (separation between SIP/SOP and secondary/endoscopic circuit) requires 2 MOOP @MAINS VOLTAGE. Is this reasonable?

Thanks,
Roland
 

Attachments

Elsmar Forum Sponsor
#3
I think the reason is that in Endoscopic ME the SIP/SOP includes INTERFACE CONDITION where the existence of mains voltage is mostly probable. Therefore, the existence of MAINS VOLTAGE at SIP/SOP is not considered as SINGLE FAULT CONDITION and, hence, should be protected by 2 MOOP rather than 1 MOOP.

Confirmation welcomed.
 

Peter Selvey

Staff member
Super Moderator
#4
I'd be inclined to think it is a mistake. It's an informative annex so not checked as closely as the main text.

Under Clause 16.5 of the general standard, if a separation device is used as a solution for SIP/SOPs, then only 1 MOP is required. But there are many other solutions possible for Clause 16, it's just an option.

From a F-Type perspective, it could require 1 MOPP but it depends on the design of the patient interface area.

I find that even some of the diagrams in Annex J of the general standard don't make sense. But again it's just informative.
 
Thread starter Similar threads Forum Replies Date
rezayatmand IEC 60601-2-18 Medical electrical equipment - Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipmen IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D IEC 60601-2-18 (IEC 601-2-18) Particular Requirements for Endoscopic Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
G IEC 60601 Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W IEC 60601-1 clause 8.8.3 - Dielectric Strength IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B IEC 60601-1 Clause 14.8 - Architecture IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R IEC 60601-2-57 light source classification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
L Top opening of enclsoure against fire in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C EN 45502 vs IEC 60601-1-8 Other Medical Device Related Standards 0
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2

Similar threads

Top Bottom