IEC 60601-2-18 (IEC 601-2-18) Particular Requirements for Endoscopic Equipment

D

DaGusta

Hello @ all,

is there anybody out there who work with the IEC601-2-18?

I have a question regarding 20.2 B-d ( and the annex 20.2). What is the reason why we have to test the dielectric strength test only on the endoscop. The standard says in the annex 20.2 "without connection to their supply units". That make for me no sense, because if the endoscope is inside the human body, the connector of the endoscope is not connected to the supply unit and lay on the floor. What should happen? The connector is made of a kind of medical plastic and if there is a hazardous voltage on the applied part... there is no closed electrical circuit.

For a response many thanks in advance and best regards
 
D

DaGusta

Re: Iec 601-2-18

Hello Harry,

yes thats right, that is the IEC 60601-2-18. There is an older version from 1996 .. on the standard is written: 601-2-18 IEC:1996. But i think that means the same.

Best regards
DG
 

Peter Selvey

Leader
Super Moderator
The comment in the Annex is informative only, and seems to indicate that the endoscope "can" be tested by itself, but this does not mean it "must".

In my experience I have seen both cases: Type BF insulation is in the endoscope, and Type BF insulation is in the equipment. Both have been OK in testing and accepted with CB scheme test reports and also for NRTL marking.

The Annex remark may have been historical basis on the assumption that endoscope might be interchangeable, but these days most endoscopes with internal circuits (CCD cameras or ultrasound etc) are custom designed for specific equipment.

Also, this comment seems to have been removed in the 2009 edition (3rd edition).
 
D

DaGusta

Hi Peter,

:thanx: for your statement. This helps me a lot! I thought this Annex is a "must" but if that is a "can" is everything ok. Because we have the insulation in the supply unit.

Yes you are right. In the third edition of the -2-18 this comment does not exist any more.

Best regards
DG
 
D

DaGusta

Does someone know when the DIN EN 60601-2-18 will be harmonized? The IEC 60601-2-18 is available since 2009 and since 2008 there is a draft. But i miss a date for the 3rd version.

For a short info many thanks in advance :)
Dagusta
 
K

KevinSun

Hi DaGusta,
I am glad can have a chance know you. I am kevin come from China one endoscopy accessories manufacturer. We have polypectomy snare Chinese report and need to translate into English. Now i have IEC 60601-1 and IEC 60601-2-2, i also need IEC 60601-2-18, can you send me English version of this full documents to my mail (old version is also ok), i appreciate you in advance. My mail is .

Thanks you, my friend!

kevin
 
K

KevinSun

Hi Marc Smith,
Thanks for your reply. I know that link but it is too expensive for me to download that documents. I am not expert, just need some lines which can corresponding our Chinese version, such as Clause 6, 6.1, 6.8.2 etc... help me translate our Chinese report into English report.
If you can help me will be appreciated!

Thanks you again.

Kevin Sun
 
Thread starter Similar threads Forum Replies Date
F Defining the need to perform the Proximity magnetic fields immunity tests (IEC 61000-4-39) according to the IEC 60601-1-2 ed.4.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601-1-8 clause 6.4.1 (statistics of the distribution of the sum of the maximum ALARM CONDITION + SIGNAL GENERATION DELAY) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
O ESD Testing IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
E Meaning of ALARM PRESETS (subclause 6.5.1 of IEC 60601-1-8:2020) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Franky Zhang123 The Electrical Standards GB9706.1:2020 (same as IEC 60601-1:2012) etc. will be implemented on May 1st, 2023 in China. China Medical Device Regulations 0
H IEC 62368-1 used in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Artificial hand (IEC 60601-1-2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L Differences between IEC 60601-1-2 and IEC TR 60601-4-2 US Food and Drug Administration (FDA) 3
S PDF version of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P IEC 60601-2-25 requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C Sample qualification to be IEC 60601-1 series tested IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
G History of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J IEC 60601-1 Clause 7.9.3.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 2
F Battery overload in IEC 60601-1 vs 62133-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Mandatory tests for IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
E Clause 8.9.1.2 of AMD2:2020 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P IEC 60601-1-2 RF wireless technology IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-2-8 cl.201.10.1.2.101 and cl.201.10.1.2.104 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G Risk Management for IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
J IEC 60601 Applied Part - Probes intended to be covered by sheaths IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S IEC 60601-1-11 Requirements related to Storage and Distribution IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
shimonv IEC 60601-1-2:2014+A1:2020 (Ed. 4.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
G IEC 60601 Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W IEC 60601-1 clause 8.8.3 - Dielectric Strength IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
B IEC 60601-1 Clause 14.8 - Architecture IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R IEC 60601-2-57 light source classification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
L Top opening of enclsoure against fire in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C EN 45502 vs IEC 60601-1-8 Other Medical Device Related Standards 0
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2

Similar threads

Top Bottom