IEC 60601-2-2 - EMC Requirements - Surgical instrument manufacturer

#1
Hi All,

We are surgical instrument manufacturer and would like to ensure our instrument is meeting EMC Directive. Our instrument is work together with electrosurgical unit(ESU) which is not our product. In order to compliance to The EMC Requirement , do we need to perform IEC 60601-1-2 EMC Test sufficient or we need to perform both IEC 60601-1-2 EMC Test and IEC 60601-2-2 Annex BB ? On top of that, I was adviced that when performing IEC 60601-1-2 EMC Test , the sytem should be in idle mode and the ESU unit should not generate energy, is it Correct ? Thanks for your value input.

James
 
Last edited:
Elsmar Forum Sponsor

isolytical

Involved In Discussions
#2
The EMC testing is performed by an independent testing house not by the manufacturer. Once contracted, the testing house will help with all questions.
 
#3
Thanks for your reply. In our case, we intend to perform testing ourself. Is there a requirement that Medical testing must be performed by independant test provider ? Thanks. :)
 

isolytical

Involved In Discussions
#4
There is an internal path for meeting the EMC directive. A company can self-certify for the requirements of the EMC Directive. The EMC Directive indicates that a product must meet emissions and immunity requirements but it does not say what the limits or values of those requirements are. The European standards give the specific values for emissions and immunity for certain products which the directive does not give.
For me, not knowing the product it is difficult to say which direction the company should take for meeting the EMC directive. The EMC directive is a certification. If you do not know how to perform the certification, or do not have the calibrated equipment to perform it, how can you perform that certification?
Perhaps professional consulting help is needed?
 

Pads38

Trusted Information Resource
#5
Assuming that your product is a medical device, then talking of the EMC directive is perhaps misleading.

The accepted route to show EMC compliance for a medical device is to show that the requirements of 60601-1-2 are met, for both emissions and immunity.

It is possible to do this work 'in house' - as it is possible to test to the base 60601-1 standard. However, setting up and calibrating equipment and installations for the EMC tests is enormously complicated and expensive. For this reason the use of an external EMC test house tends to make a lot of sense.

A nice 'Certificate' from a reputable EMC test house, particularly one with ISO17025 certification, can enhance confidence in your customers and regulators.

There are many parts of 60601-1-2 that can be done as 'pre-compliance' tests. We often do this to try and eliminate potential difficulties before the large expense of the formal tests. For this we often rent the specialist equipment needed.

As for 60601-2-2 - I am not familiar with that standard. But I can see that your device would need to be tested to prove that concurrent use does not cause a problem. This would have to be part of your Risk Management process. There may be tests in that Particular standard that you could use to show satisfactory concurrent use.
 
#6
Hi guys,
Could you please help me to answer below mentioned query?
When there is swagged needle attached to suture and whole combination is supplied as finished product, should we consider needle as an instrument or not? And whether ISO:16061 is applicable or not?

I tried finding New Thread but I could not get it.

Thanks and regards,
RA World
 

isolytical

Involved In Discussions
#7
What part of the swagged suture is electrical? - none based on my knowledge. What IEC 60601 test do you think applies? To obtain FDA marketing clearance the evidence of 'substantially equivalent to a predicate device(s) under the FFDC Act' is required, and the device(s) chosen for that comparison should at least help with requirements. Perhaps some external consulting help is needed to identify the quickest route to market.
 
Thread starter Similar threads Forum Replies Date
M IEC 60601-1-2 EMC Tables IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
O EMC testing per IEC 60601-1-2 for considering new cables CE Marking (Conformité Européene) / CB Scheme 1
O EMC testing under RED and IEC 60601-1-2 CE Marking (Conformité Européene) / CB Scheme 1
C EMC Test Report Requirements to IEC 60601-1-2:2009 ISO 13485:2016 - Medical Device Quality Management Systems 4
P New EMC standard IEC 60601-1-2:2014 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T IEC 60601-1-2 EMC and WiFi IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
T How to fulfill IEC 60601-1-2 EMC Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S IEC 60601-1-2 EMC Testing Optional in US? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
C An ongoing EMC test in accordance with IEC/EN 60601-2-47 EMC IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
J EMC (Electromagnetic Compatibility) Requirements - IEC 60601 3rd edition Clause 17 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S EMI/EMC Failure Cause Analysis per IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Do HF accessories require EMC testing according to IEC 60601-1-2? IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
SteveK Does a 12V DC device need EMC (IEC 60601-1-2) Testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B IEC 60601-1: EMC problems: Confused about 915MHz ISM band in Europe IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Particular standard IEC 60601-2-40 references obsolete EMC standard IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601-1-2 - EMC Retesting Rational - Minor Component Changes IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Has anyone done an IEC 60601-1 gap analysis to IEC 60335? Medical Device and FDA Regulations and Standards News 4
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601-1 and IEC 60601-2-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Bookmarking my security protected IEC 60601-1 .pdf file IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E ECG device without ECG REPORTS - IEC 60601-2-25: 2011 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1

Similar threads

Top Bottom