IEC 60601-2-24 & ISO 8536-8 - Conflicting Requirements?

H

hevak

#1
Dear All,

Is anyone able to help provide some clarification/assistance regarding what appear to be conflicting requirements for infusion set/pump performance and labeling requirements as described in ISO 8536-8 and IEC 60601-2-24?

Specifically, ISO 8536-8, under 6.14 describes that storage volume is to be determined in accordance with IEC 60601-2-24 and declared in labeling in accordance with ISO 8536-8 10.1(h). However IEC 60601-2-24 describes that the testing is to be performed at 20C +/- 2C / 65%RH +/- 5%RH, while ISO 8536-8 10.1(h) describes that the storage volume should be stated at 40C.

Any assistance on this would be greatly appreciated.
 
Elsmar Forum Sponsor

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#2
A quick "Bump". I know this thread is kind of old, but any help will be appreciated! My Thanks in advance!
 
H

hevak

#3
Thank you for the "Bump" Marc.

I've received a written response from ISO that they are aware of the conflict and are currently working on a technical document to provide guidance.

In the meantime I have advised my client that the best approach in this particular case should be risk-based and linked with the intended use of the device. In this case they also export the device to the European market and therefore they should also be aware that the ISO standard is a harmonized standard and that the IEC standard is not.

As soon as I'm aware that the technical document has been published I will add a link on the forum for reference purposes for other members.
 
Thread starter Similar threads Forum Replies Date
D IEC 60601-1 and ISO 14971 Assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
M ISO 14971, IEC 60601 Satisfy 98/37/EC, 2006/95/EC, 2004/108/EC Directives? Other ISO and International Standards and European Regulations 3
E Upcoming changes for ISO 10993 and IEC 60601 in South Korea Other Medical Device Regulations World-Wide 2
D Patient Leakage ISO 7176-14 vs. IEC 60601-1:2005 Reporting Requirements Other US Medical Device Regulations 1
B IEC 60601-1 3rd Edition adoption - I cant find this standard on the ISO page Reliability Analysis - Predictions, Testing and Standards 9
N ISO 14971 vs. IEC 60601-1 3rd Ed. - What to use for the RM evaluation? ISO 14971 - Medical Device Risk Management 17
J IECEE Guidance Document for ISO 14971 Risk Analysis - IEC EN 60601-1:2005 3rd edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L MDD and Mandatory Standards (IEC/EN 60601-1-4, 60601-1-6 and ISO 14971) EU Medical Device Regulations 6
R IEC 61010 vs IEC 60601-1 to ISO 15197 - Blood glucose monitoring device IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
W What kind of materials should I prepare for passing ISO IEC 60601-1-4 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1

Similar threads

Top Bottom