Z Zameer Registered Oct 9, 2020 #1 Oct 9, 2020 #1 Hi everyone, Could anyone please explain clause 201.12.4.107.1.1.2 Low frequency (impulse ) response test. And also calculation method to determine PASS/FAIL. Best regards Zameer
Hi everyone, Could anyone please explain clause 201.12.4.107.1.1.2 Low frequency (impulse ) response test. And also calculation method to determine PASS/FAIL. Best regards Zameer
P Peter Selvey Leader Super Moderator Nov 1, 2020 #2 Nov 1, 2020 #2 Have a look at Articles — MEDTEQ There's a section on ECG including an article on ECG filters which will give you a background, and an update to point out the errors in that article (ECG Low Frequency Response Impulse test (0.05Hz, 3mV/100ms test). It turns out to be a complicated test!
Have a look at Articles — MEDTEQ There's a section on ECG including an article on ECG filters which will give you a background, and an update to point out the errors in that article (ECG Low Frequency Response Impulse test (0.05Hz, 3mV/100ms test). It turns out to be a complicated test!