IEC 60601-2-43 - Clause 203.6.103 - Physical button?

BargeSim

Starting to get Involved
#1
Hi,

Standard IEC 60601-2-43: 2010 requires to have an "IRRADIATION DISABLING SWITCH" @clause: 203.6.103.

"A switch shall be provided to disable/enable the LOADING STATE without affecting any other functions of the INTERVENTIONAL X-RAY EQUIPMENT. The operation of this switch shall not, in itself, be capable of initiating the LOADING STATE. The state of the IRRADIATION disabling switch shall be displayed at the working position of the OPERATOR. The switch should be configured to minimize the likelihood of accidental operation."


Does this mean a "physical button" or could it be a software setting/ button?


Thank you for your help
 
Elsmar Forum Sponsor

Benjamin Weber

Trusted Information Resource
#2
I would say, a physical button is not explicitely required. The standars says "a switch" and this is not explained/defined in detail.

So I think, yes, it could be a software button. But be aware, that software as a means of protection (MOP) has to be treated in the risk management process and the soultion shall be dafe in single fault condition. And implementing MOP via software can be very tricky!
 
Thread starter Similar threads Forum Replies Date
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A Fire hazard classification - Clause in IEC 60601 about gauze and electricity ISO 14971 - Medical Device Risk Management 0
S ME system clause 16 of IEC 60601-1, 3.1 ed IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T What is PESS? Cited in Edition 3.1 of IEC 60601, clause 14.1 note 2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E ME (Medical Equipment) Systems - IEC 60601-1 Clause 16.1 Interpretation IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Question on IEC 60601 clause 15.5.1.1 (transformer overheating) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L IEC 60601-1 3rd edition Clause 9.2.2 Trapping Zone (Acceptable Gaps) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E IEC 60601-1 Clause 14 Questions IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R How to understand the exception of IEC 60601 sub-clause 8.11.5 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M MEEE Expected Service Life - IEC 60601-1 Clause 4.4 Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
R How to calculate the Air Flow Rate in clause 201.12.4.4.105 of IEC 60601-2-16:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P IEC 60601-1 Clause 11.1.3 Winding Temperature Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J EMC (Electromagnetic Compatibility) Requirements - IEC 60601 3rd edition Clause 17 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
I Dielectric Test for HF Surgical Equipment (IEC 60601-2-2 clause 201.8.5.1.2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601 Clause 7.1.3 - Durability of Markings - Methylated Spirits IEC 60601 - Medical Electrical Equipment Safety Standards Series 19
P Understanding IEC 60601-1:2005, Clause 8.1 b) - Short Circuiting Double Insulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
P IEC 60601 Clause 57.3 / 8.11.3.2 - PVC Flexible Cords IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
G Single Fault Condition for 3-Phase Medical System - IEC 60601 Clause 8.1b IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B IEC 60601-1 2nd Clause 17 and 3rd edition - Electrical Isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1

Similar threads

Top Bottom