IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge

Benjamin Weber

Trusted Information Resource
#1
Dear community,

cl. 8.4.3 says:

"ME EQUIPMENT or its parts intended to be connected to a power source by means of a plug (...)"

In my understanding this does not only apply to the mains plug, but also to all other plugs that have accessible pins (e.g. HF surgery electrode cables)?
 
Elsmar Forum Sponsor

Benjamin Weber

Trusted Information Resource
#3
I justed wanted to see, if there other interpretations. I have seen some test report from big test labs, where this clause was marked "N/A" saying, there are no plugs to be connected to mains :rolleyes:.

To me it is clear, that it does not matter, of the residual voltage is caused by mains, secondary or battery voltage.
 
Thread starter Similar threads Forum Replies Date
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P Scope of application for IEC 60601-1-11 Medical electrical equipment — Part 1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
rezayatmand IEC 60601-2-18 Medical electrical equipment - Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipmen IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D The necessity of a SECONDARY CIRCUITS in ME EQUIPMENT (IEC 60601-1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R IEC 60601-2-57 - Subclause 201.7.101.2 - Explanatory Label - LS Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
F Test Equipment for IEC 60601-2-27 related to the picture 201.105 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E ME (Medical Equipment) Systems - IEC 60601-1 Clause 16.1 Interpretation IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R IEC 60601-2-18 Endoscopic Medical Electronic Equipment question IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
K Definition of ME Equipment/Applied part according to IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
I IEC 60601-1 Least Favourable Working Conditions for Medical Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
I Dielectric Test for HF Surgical Equipment (IEC 60601-2-2 clause 201.8.5.1.2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J Marking (Labeling) Requirements on Medical Equipment per IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
D IEC 60601-2-18 (IEC 601-2-18) Particular Requirements for Endoscopic Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
J Medical Equipment (AC Adapter) Classification as per IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
N CB Scheme and Conformity of Medical Equipment to IEC/EN 60601 EU Medical Device Regulations 2
M IEC 60601 Medical electrical equipment - Send declaration and guidance to customers? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-2-6, Microwave Therapy Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Medical Electrical Equipment / System Peripherals - IEC 60601-1-3 IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
D Training for IEC 60601-1-8:2006 Alarm Systems in Medical equipment. IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Has anyone done an IEC 60601-1 gap analysis to IEC 60335? Medical Device and FDA Regulations and Standards News 4
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601-1 and IEC 60601-2-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601-1-2 EMC Tables IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Bookmarking my security protected IEC 60601-1 .pdf file IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E ECG device without ECG REPORTS - IEC 60601-2-25: 2011 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C IEC 60601-1-8, difference between table 4 and annex D IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601 7.1.3 Durability of Markings - Methylated Spirit or Ethanol? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C ABS plastic : which IEC 60601-1 tests to redo? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Q Isolation/Insulation Diagrams - IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Similar threads


















































Top Bottom