IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge

Benjamin Weber

Trusted Information Resource
#1
Dear community,

cl. 8.4.3 says:

"ME EQUIPMENT or its parts intended to be connected to a power source by means of a plug (...)"

In my understanding this does not only apply to the mains plug, but also to all other plugs that have accessible pins (e.g. HF surgery electrode cables)?
 
Elsmar Forum Sponsor

Benjamin Weber

Trusted Information Resource
#3
I justed wanted to see, if there other interpretations. I have seen some test report from big test labs, where this clause was marked "N/A" saying, there are no plugs to be connected to mains :rolleyes:.

To me it is clear, that it does not matter, of the residual voltage is caused by mains, secondary or battery voltage.
 
Thread starter Similar threads Forum Replies Date
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P Scope of application for IEC 60601-1-11 Medical electrical equipment — Part 1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
rezayatmand IEC 60601-2-18 Medical electrical equipment - Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipmen IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D The necessity of a SECONDARY CIRCUITS in ME EQUIPMENT (IEC 60601-1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R IEC 60601-2-57 - Subclause 201.7.101.2 - Explanatory Label - LS Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
F Test Equipment for IEC 60601-2-27 related to the picture 201.105 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E ME (Medical Equipment) Systems - IEC 60601-1 Clause 16.1 Interpretation IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R IEC 60601-2-18 Endoscopic Medical Electronic Equipment question IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
K Definition of ME Equipment/Applied part according to IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
I IEC 60601-1 Least Favourable Working Conditions for Medical Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
I Dielectric Test for HF Surgical Equipment (IEC 60601-2-2 clause 201.8.5.1.2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Marking (Labeling) Requirements on Medical Equipment per IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
D IEC 60601-2-18 (IEC 601-2-18) Particular Requirements for Endoscopic Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
J Medical Equipment (AC Adapter) Classification as per IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
N CB Scheme and Conformity of Medical Equipment to IEC/EN 60601 EU Medical Device Regulations 2
M IEC 60601 Medical electrical equipment - Send declaration and guidance to customers? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-2-6, Microwave Therapy Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Medical Electrical Equipment / System Peripherals - IEC 60601-1-3 IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
D Training for IEC 60601-1-8:2006 Alarm Systems in Medical equipment. IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3

Similar threads

Top Bottom