IEC 60601 Clause 57.3 / 8.11.3.2 - PVC Flexible Cords
phase90
I am going through 601 and I am having a bit of an issue with clause 57.3 (2 nd edition) and 8.11.3.2 (3rd edition) regarding cord types. Do cord sets need to be certified to UL 227 or just use a robust PVC? The cord we are using is a hospital grade cord (green dot and all) that uses UL listed connectors at both ends and UL listed cord. The manufacturer says the cords are made in compliance with UL 817.
Do the two tie together (817 and 227)?
raymond02302
Re: Clause 57.3 / 8.11.3.2 - PVC
IEC 60601 is the international standard for medical equipment. Since IEC issues the standards for flexible cable (IEC 60227 for PVC jacket and IEC 60245 for rubber jacket),IEC 60601-1 specifies requirement flexible should be certified to relevant standards (IEC 60227 or IEC 60245).
Based on national deviation, UL 60601-1 corresponding to IEC 60601-1 2nd edition specify its own requirment for flexible cable (UL 62) or cord set(UL 817).
When you apply an NRTL mark such as UL,ETL,CSA etc, The cord set should be certified to UL 817.
phase90
Thank you for your quick response.
I am looking at UL 60601-1 (2nd edition), clause 57.3b (types). It also refers to IEC 227 designation 53. (the ANSI AAMI 3rd edition, clause 8.11.3.2 also refers to IEC 60227, Annex A, designation 53). I do not see where it refers to UL817. Where is this deviation? The only place that I found UL 817 is in Annex DVA where it just says cord sets UL817.
If a cord set is certified to UL817, does that mean that it is robust enough to satisify the conditions of UL 227 / IEC 60227?
Thanks.
Peter Selvey
Power cables (cords and plugs) are normally covered by separate regulations in each country. Thus, it really does not matter what IEC 60601-1 says, you need to comply with those power cord regulations.
Fortunately, most suppliers of cord sets know these regulations well, so if you buy a cord set with a US plug, it will comply with US regulations. If you buy one with an Australian plug, it will comply with Australian regulations. And if you want to be sure, contact the supplier of the cord set.
Also note that it is possible to argue that in IEC 60601-1, for equipment that uses a detachable cord set, the cord set itself is not part of the "medical electrical equipment".
This is based on 57.3 which says that "EQUIPMENT ... shall be provided with either a POWER SUPPLY CORD or an APPLIANCE INLET".
So, if the equipment has an appliance inlet (for a detachable cord), then the power supply cord is not part of the equipment. This makes sense since of course in the real world it can be used with any detachable cord set.
A Fire hazard classification - Clause in IEC 60601 about gauze and electricity ISO 14971 - Medical Device Risk Management 0
S ME system clause 16 of IEC 60601-1, 3.1 ed IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T What is PESS? Cited in Edition 3.1 of IEC 60601, clause 14.1 note 2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E ME (Medical Equipment) Systems - IEC 60601-1 Clause 16.1 Interpretation IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Question on IEC 60601 clause 15.5.1.1 (transformer overheating) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L IEC 60601-1 3rd edition Clause 9.2.2 Trapping Zone (Acceptable Gaps) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E IEC 60601-1 Clause 14 Questions IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R How to understand the exception of IEC 60601 sub-clause 8.11.5 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M MEEE Expected Service Life - IEC 60601-1 Clause 4.4 Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
R How to calculate the Air Flow Rate in clause 201.12.4.4.105 of IEC 60601-2-16:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P IEC 60601-1 Clause 11.1.3 Winding Temperature Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J EMC (Electromagnetic Compatibility) Requirements - IEC 60601 3rd edition Clause 17 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
I Dielectric Test for HF Surgical Equipment (IEC 60601-2-2 clause 201.8.5.1.2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601 Clause 7.1.3 - Durability of Markings - Methylated Spirits IEC 60601 - Medical Electrical Equipment Safety Standards Series 19
P Understanding IEC 60601-1:2005, Clause 8.1 b) - Short Circuiting Double Insulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
G Single Fault Condition for 3-Phase Medical System - IEC 60601 Clause 8.1b IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B IEC 60601-1 2nd Clause 17 and 3rd edition - Electrical Isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2