IEC 60601 for Non-Medical Research Purpose Devices

Q

Qara5

#1
Hello! This is my first post so I hope someone can help me.

At my company we sell a medical device into many markets including the EU and US. We are CE marked in the EU and FDA approved in the US. We tested to 60601 on the recommendation of our NB (As an aside I'm not 100% convinced this was correct as whilst we are definitely medical equipment we are never in the vicinity of the patient so maybe 61010 would have been more appropriate).

We are now looking to sell the same device into the life sciences market. Outside a medical setting our device would appear to be laboratory equipment. So my question is this - does our 60601 testing cover our medical device for non-medical research purposes or do we need to retest it against 61010?
 
Elsmar Forum Sponsor
K

KTcons

#3
I think that a comparative analysis between the two standards could be sufficient: requirements of EN 60601 should be more restrictive...
:cfingers:
Katia
 
S

Spazz

#4
I agree with KTcons that a comparative analysis between the two standards should be enough. Essentially you'd be performing a gap analysis on data that's missing for 61010.

But before you get that far...

At my company we sell a medical device into many markets including the EU and US. We are CE marked in the EU and FDA approved in the US.
...
We are now looking to sell the same device into the life sciences market. Outside a medical setting our device would appear to be laboratory equipment.
For CE Marking at least, I'd double check which directive(s) would apply now that the market (and intended use, I assume) has changed. 61010 may not be applicable for that directive.
 
Q

Qara5

#5
Many thanks both for your replies.

I'd double check which directive(s) would apply now that the market (and intended use, I assume) has changed. 61010 may not be applicable for that directive.
Good point! I shall look into that.

Whilst we are in the topic - we have a lot of circular discussions in our company about whether we even need 60601 or if 61010 would be sufficient for our MDs. Some people's opinion is that it is a medical device therefore must have 60601. Others feel that 60601 should only apply to a device that is in the vicinity of the patient which ours is definitely not. The information online seems to mirror this - I can't find anything definitive and it seems very open to interpretation. Our notified body sent us down the route of 60601 but I notice that competitors products are only 61010! Would be interested in the opinions of other forum users.
 
S

Spazz

#6
Whilst we are in the topic - we have a lot of circular discussions in our company about whether we even need 60601 or if 61010 would be sufficient for our MDs.
We were in a similar situation not long ago at my company and the participants on this forum were very helpful in pointing out that it all depends on whether or not the device met the scope of the standards (Sec. 1 which usually is available for free preview). That said, if your NB is asking for a specific standard, you probably have very little choice...
 

Pads38

Trusted Information Resource
#7
Directives versus standards!

EN 60601 is a harmonised standard under the medical device directive EEC/93/42 (MDD).

EN 61010 is not.

That means that you would have a very difficult time in persuading a notified body that testing to EN61010 shows that the Essential Requirements of the MDD have been met.

Notified Bodies assess your device and Technical Files to the requirements of the MDD, not standards.
 
Q

Qara5

#8
Directives versus standards!

EN 60601 is a harmonised standard under the medical device directive EEC/93/42 (MDD).

EN 61010 is not.
Thank you! That is very helpful. I think I am starting to understand after reading these responses and some similar posts on this forum. So for our medical devices we must conform to the MDD. 60601 is the harmonised standard under that so shows that me meet the essential requirements of the MDD. For our non-medical devices the MDD no longer applies so 60601 will no longer imply compliance but we can still use it with some justification. Please correct me if I've misinterpreted and thanks again for all the help!
 
Thread starter Similar threads Forum Replies Date
M IEC 60601 Compliance requirements for NON- ME Equpment/Systems IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC/EN 60601 Testing and CE Marking for a Non Medical Device CE Marking (Conformité Européene) / CB Scheme 11
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Has anyone done an IEC 60601-1 gap analysis to IEC 60335? Medical Device and FDA Regulations and Standards News 4
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601-1 and IEC 60601-2-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601-1-2 EMC Tables IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Bookmarking my security protected IEC 60601-1 .pdf file IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E ECG device without ECG REPORTS - IEC 60601-2-25: 2011 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C IEC 60601-1-8, difference between table 4 and annex D IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Scope of application for IEC 60601-1-11 Medical electrical equipment — Part 1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601 7.1.3 Durability of Markings - Methylated Spirit or Ethanol? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C ABS plastic : which IEC 60601-1 tests to redo? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Q Isolation/Insulation Diagrams - IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-2-2 - Defib-proof test amendment to base standard IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
pashah Looking for Usability File Template acc. IEC 62366-1 and IEC 60601-1-6 IEC 62366 - Medical Device Usability Engineering 4
P Testing lab in India IEC 60601-2-25 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Kuldeep Singh Electromagnetic Compatibility Test IEC 60601-1-2: 2014 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
rezayatmand IEC 60601-2-18 Medical electrical equipment - Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipmen IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Fire hazard classification - Clause in IEC 60601 about gauze and electricity ISO 14971 - Medical Device Risk Management 0
K IEC 60601-1 Ed. 3.1 b:2012: Replaced IEC 60601-1 Ed. 3.1 en:2012? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Impact test on glass touch controller IEC 60601 - Medical Electrical Equipment Safety Standards Series 2