IEC 60601 vs. IEC 61010 - CE Mark for our IVD medical device

M

menlomed

#1
We are in the process of doing a gap analysis for getting a CE Mark for our IVD medical device which is currently cleared in the US. One question that came up is which electrical safety standard we need to be following. For the US, we comply with 60601. For Europe it seems like there is a separate standard for IVDs - 61010. Can anyone confirm this?

Thanks.
 
Elsmar Forum Sponsor

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#2
Re: IEC 60601 vs IEC 61010

It appears that ISO/IEC 60101 may be more for medical devices, while ISO/IEC/UL/CSA 61010 is for laboratory equipment.

Best bet is to find a CB Scheme recognized lab and talk to them. Examples in North America are SGS, UL, CSA, ITS, and I suspect others as well.

Hope this helps.
 

Peter Selvey

Staff member
Super Moderator
#3
According to the scope, IEC 60601-1 is only for medical devices with patient contact, this may include some IVD devices where the sample is taken using the same device (such as home use blood sugar level testers).

However, for the vast majority of IVDs the patient is well removed, and hence IEC 60601-1 is not applicable. The correct standard is IEC 61010-1, and any particulars that apply to your device.

There is persistent and incorrect use of IEC 60601-1 for anything that is electrical and medical because of the title of the standard. The scope itself which refers to patient contact are buried inside the standard.

The FDA complicates this, because IEC 60601-1 is listed as a recognized consensus standard, while IEC 61010-1 is not. Thus, many feel more comfortable using IEC 60601-1 in the US even though it is not the correct standard.

In Europe, you will definately have to use the correct standard: IEC 61010-1 and any particulars that are applicable.
 

Marcelo

Inactive Registered Visitor
#4
According to the scope, IEC 60601-1 is only for medical devices with patient contact
Just to correct the information a little, IEC 60601-1 is for electrical medical devices, let´s say it this way, which have patient contact (applied parts by the definition of the standard), or transfer energy to or from the patient or detect this transfer (which means the equipment really does not have to contact the patient).

Still, IVDs devices, as you say, generally do not fit this scope.
 
B

biomargam

#5
Product Safety in Medical Device

Hi All
I am trying to compile the Product Safety standard and testing requirements for Medical Devices. Would appreciate knowing if any of you have any presentation or reference materials on the above.

Thanks
 
S

SRavi

#6
Hello All,

I am working on a IVD medical device. The device is an implantable pressure measurement. The implant communicates with the reader device via RFID means. I would like to know what standards will apply to the reader device antenna. The antenna is in contact with skin. Is IEC 60601 an appropriate one or should I also explore IEC 61010.

Thank you,
Ram
 
#7
Thanks everyone for your input. I’m working with a client that manufactures a product that heats water in which a heat pad is placed. After the heat pad has warmed up, the heat pad is removed and placed on the patient. Should the product be compliant to 60601-1, or 61010-1? It does not come in direct contact with the patient, but the heat pad it warms up does contact the patient. I’m in contact with Intertek and they are telling me 61010-1, but I’m not quite sure that’s correct. Also my client tells me that some competitors products are 60601-1 compliant, while other competitor products are 61010-1 compliant.
 

Michael Ayers

Starting to get Involved
#8
Thanks everyone for your input. I’m working with a client that manufactures a product that heats water in which a heat pad is placed. After the heat pad has warmed up, the heat pad is removed and placed on the patient. Should the product be compliant to 60601-1, or 61010-1? It does not come in direct contact with the patient, but the heat pad it warms up does contact the patient. I’m in contact with Intertek and they are telling me 61010-1, but I’m not quite sure that’s correct. Also my client tells me that some competitors products are 60601-1 compliant, while other competitor products are 61010-1 compliant.
I think the distinction here is whether the water heater is designed to be used with the device that has the heat pad (which I am assuming is itself is a medical device and complaint with 60601). If the water heater is designed to be used with the heat pad device (or with these types of medical devices), it would be classified as an accessory to a medical device, and regulated in the same was as a medical device itself would be (e.g. by following 60601). If it is not intended to be used with a heat pad device, it would not be classified as an accessory to a medical device (for example, a domestic microwave 'could' be used to heat up a heat pad device, but since it is not explicitly designed to do so by the manufacturer, it will not be regulated as a medical device).

The definition of a medical device accessory in the MDD is
an article which whilst not being a device is intended specifically by its manufacturer to be used together with a device to enable it to be used in accordance with the use of the device intended by the manufacturer of the device
I suspect the same applies with regards to why some similar products are complaint with 61010 (e.g. they are accessories to devices that are themselves complaint with 61010, as opposed to 60601), although someone else may have to clarify that.
 
Thread starter Similar threads Forum Replies Date
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Has anyone done an IEC 60601-1 gap analysis to IEC 60335? Medical Device and FDA Regulations and Standards News 4
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601-1 and IEC 60601-2-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3