Hi all,
We (medical device company) are working on a manufacturing process FMEA. The question has come up whether to follow IEC 60812 or ISO 14971. Our current pFMEA does not discuss harm at all. I know 14971 use to have a reference to 60812 but no longer does?
Looking for some overall guidance on what to do or how to think about this. Thank you!
yodon
Risk management for medical devices should be done in context of effects on the patient / user. I don't have 60812 but it looks like it gives a good framework for PFMEAs; as long as the right focus is maintained. 14971 doesn't prescribe any specific approaches.
Tidge
14971 is considered 'foundational' for 60601-1 (electrical devices) and 62304 (devices that include software) so I wouldn't eschew it if your device falls into either category.
Ronen E
Problem Solver
Moderator
Hi all,
IMO these are two separate aspects of risk management.
Assuming that "medical device company" means a finished medical devices legal Manufacturer (bearing the top level regulatory responsibility in e.g. the USA or the EU), I'd highly recommend applying ISO 14971 in full, regardless of your current, specific pFMEA endeavour. Such application is in principle technique-free, so you can use different published standards (or none) in support of it, as long as you follow the normative (binding) parts of 14971.
If you feel IEC 60812 might be helpful in your pFMEA (and not too onerous for your purposes) then go ahead and apply it; however, there's no regulatory obligation that I'm aware of to use it (or any other FMEA published standard, for that matter).
B Risk Management and FMEAs - IEC 60812, SRC-HDBK-1120 and SEQ-J1739 ISO 14971 - Medical Device Risk Management 3
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 62366 Hazard-related Use Scenario vs FDA Critical Task EU Medical Device Regulations 3
C Sample qualification to be IEC 60601-1 series tested IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H IEC 62368-1 Scope of Testing CE Marking (Conformité Européene) / CB Scheme 3
A IEC 18562-1 report EU Medical Device Regulations 5
H IEC 61010-1 Clause No. 10.5 Resistance to Heat - Sub Clause 10.5.3 Insulating Material CE Marking (Conformité Européene) / CB Scheme 6
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
H Required explanation of sub clause 7.2.101.2 (Exception) of IEC 61010-2-020:2006 CE Marking (Conformité Européene) / CB Scheme 5
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
G History of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
H Pollution Degree (IEC 60664-1) CE Marking (Conformité Européene) / CB Scheme 2
T FDA level of concern vs IEC 62304 safety classification - 510(k) exempt device Other US Medical Device Regulations 13
J IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Y Software to Manage IEC 62304 Traceability Requirement IEC 62304 - Medical Device Software Life Cycle Processes 3
J IEC 60601-1 Clause 7.9.3.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 2
H Regarding Usability Engineering File according to IEC 62366-1 CE Marking (Conformité Européene) / CB Scheme 2
F Battery overload in IEC 60601-1 vs 62133-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Mandatory tests for IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
C Producing to Inactive ASTM and IEC Standards Manufacturing and Related Processes 11
K IEC Maximum Power Density for Muscle Stimulator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E Clause 8.9.1.2 of AMD2:2020 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
H Regarding confusion of the validity of the European implementation of IEC standards Other Medical Device Related Standards 1
P IEC 60601-1-2 RF wireless technology IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Should I buy IEC 62304:2006, IEC 62304:2006/AMD 1:2015 or both? EU Medical Device Regulations 1
A IEC 60601-2-8 cl.201.10.1.2.101 and cl.201.10.1.2.104 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G Risk Management for IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
J IEC 60601 Applied Part - Probes intended to be covered by sheaths IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M How to apply IEC 62304 Problem Resolution Process for bugs discovered prior to release i.e. during initial development? IEC 62304 - Medical Device Software Life Cycle Processes 2
S IEC 60601-1-11 Requirements related to Storage and Distribution IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
shimonv IEC 60601-1-2:2014+A1:2020 (Ed. 4.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T IEC 62304 : Risk control for SaMD IEC 62304 - Medical Device Software Life Cycle Processes 8
G IEC 60601 Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W IEC 60601-1 clause 8.8.3 - Dielectric Strength IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
L ISO/IEC 20000-6 Technical Areas IT (Information Technology) Service Management 2
B IEC 60601-1 Clause 14.8 - Architecture IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T Infusion pump and IEC 62353 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F IEC 62304 agile development EU Medical Device Regulations 1
R IEC 60601-2-57 light source classification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
shimonv Working with a software developer who is not setup for IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 9
P Examples of quality plans in IEC 62304 US Food and Drug Administration (FDA) 2
L Top opening of enclsoure against fire in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S ISO/IEC 15408 - Is this is Certifiable Standard? Other ISO and International Standards and European Regulations 2
C EN 45502 vs IEC 60601-1-8 Other Medical Device Related Standards 0
Richard Regalado Informational ISO/IEC DIS 27002:2022, to be published soon. IEC 27001 - Information Security Management Systems (ISMS) 1
E Test report to certify compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 5
B 3-phase supply, overvoltage catogory, line-to-neutral and working voltage (IEC 61010-1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B European Harmonized vs IEC standard IEC 62366 - Medical Device Usability Engineering 2