IEC 61000-4-2: Requirements for coating declared as insulating
Hi everyone,
IEC 61000-4-2 contains a statement regarding painted surfaces, "Coating declared as insulating by the manufacturer shall only be
submitted to the air discharge. The contact discharge test shall not be applied to such surfaces."
Is there any reference to the insulating coating characteristics like min impedance, coating type, thickness, etc?
All the best,
Meir
Benjamin Weber
Hey Meir,
IEC 60601-1 has some notes regarding coatings as insulation, but no particular requirements. You chould consider the coating as solid insulation. The corresponding requirements should then be identified from the relevant standard sections.
To be honest, I think such a coating can be used as solid insulation, but it may be difficult to specify it as a means of protection against electric shock. In this case there might be requirements, the coating maybe could not fulfill, e.g. mechanical strength tests (impact test!) and maybe also minimum thickness of 0,4mm if the working voltage is above 70V!
So be careful, how you specify the coating!
And even if you specifiy it "only" to comply with EMC requirements. Does your device implement any critical functions maybe even essential performance? If yes: You should evaluate how reliable the coating is over the whole expected lifetime of your device. If the coating degrades over time or might be damaged due to external influences (e.g. mechanical hazards), than the contact discharge on the uncoated surface is maybe more relevant! Your device should pass the ESD tests in this as well.
Thank you Benjamin.
I'm aware of the need to ensure esential performance over expected lifetime of out product.
I was looking for someone with practical experience who could share data on existing coating technologies and durability over time and use.
I Formula in the Annex of IEC 61000-4-3 - How is k=7 determined? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Radiated Immunity Device Test Orientations - IEC 61000-4-3 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B What is our requirement for maintaining the EMC series of standards? IEC 61000 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
L Build own Power Magnetic Field Test Equipment for IEC 61000-4-8 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Is LED STREP 12v with power supply Class C - IEC 61000 3-2 CE Marking (Conformité Européene) / CB Scheme 2
K IEC 62304 - Testing Independance IEC 62304 - Medical Device Software Life Cycle Processes 5
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Exclusions or justification for non-applicability of IEC standards Reliability Analysis - Predictions, Testing and Standards 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
T ISO/IEC 17065 certification scheme Help Other ISO and International Standards and European Regulations 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
K IEC 62304 - Functional and performance requirements for SOUP items IEC 62304 - Medical Device Software Life Cycle Processes 2
K IEC 62304 compliance - Code reviews as part of verification strategy IEC 62304 - Medical Device Software Life Cycle Processes 5
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 62471 vs IEC/TR 62778 CE Marking (Conformité Européene) / CB Scheme 0
B IEC 61010, Touch currents related to burns at higher frequencies (cl. 6.3, A.3) Other Medical Device Related Standards 0
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
B IEC 61010-1 - Voltage / Dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M When is EN 62133-1 (new version i.e. IEC 62133:2017) mandatory in EU Other ISO and International Standards and European Regulations 0
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
adir88 Amendment 1 of IEC 62366-1 - Details yet? IEC 62366 - Medical Device Usability Engineering 2
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
K Primary Operating Functions (POF) in IEC 62366-1:2015. IEC 62366 - Medical Device Usability Engineering 5
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 62304 Class A Project IEC 62304 - Medical Device Software Life Cycle Processes 15
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 5
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
MDD_QNA QR Code Standard ISO/IEC 15417:2007 - Does anyone use it? Other Medical Device Related Standards 3