IEC 61010-1 and Flammable Liquids

#1
Hello everyone,

it's my first thread at this forum. :bigwave:

I am seeking tips related to IEC/EN 61010-1:2010 standard, clause 9.5 "Requirements for equipment containing or using flammable liquids".

The device is an electrical laboratory instrument that measures a property of liquid sample.

The simplest solution to bypass the clause would be to prohibit use of flammable sample liquids with the instrument. However, that is not ideal from business perspective. Sometimes customers will want to use liquids that are classified as flammable, such as ethanol.

Clause 9.5 offers three options to pass the clause:
a) temperature of sample liquid and materials contacting the liquid is limited to T+25 Celsius where T is fire point of liquid
b) quantity of liquid is limited to amount that does not cause spread of fire
c) fire enclosure

Option c) is ruled out at this point due to design. Option a) might be possible to implement as instruction for users, but liquid fire points are not generally listed at MSDS per my knowledge. Also this would rule out some flammable sample liquids, at least high concentrations of ethanol where the fire point is about 30 Celsius.

So I am left wondering how to verify conformance to option b). In a worst-case scenario the sample liquid volume could be 70 milli-litres, but that is more than enough for successful measurement. 10 milli-litres is a minimum needed for the measurement. Standard laboratory glass beakers are used as sample vessels.

Does anyone have experience with flammable liquids with similar instruments? How will a test laboratory verify conformance with clause 9.5b) ?

I highly appreciate any hints you might have.
 
Last edited:
Elsmar Forum Sponsor

Kevin Mader

One of THE Original Covers!
Leader
Admin
#2
Hi ilkkahoo,

I don't have direct experience with your problem, but I've cleared a number of lab type equipment under IEC61010-1. My suggestion here is to contact your safety folks (UL, METLabs, etc....whoever you use) and talk it through with them. They have the experience and possible solutions (design) that might help you achieve what you desire relative to you customers' usages.

I've found them helpful over the years and most are willing to do the work upfront for no/little out of pocket expense (especially if you end up doing the testing/retesting with them in the near-term). And it's a great way to set the right expectations for all parties going forward.

Good luck!!

Kevin
 
Thread starter Similar threads Forum Replies Date
H IEC 61010-1 Clause No. 10.5 Resistance to Heat - Sub Clause 10.5.3 Insulating Material CE Marking (Conformité Européene) / CB Scheme 6
H Required explanation of sub clause 7.2.101.2 (Exception) of IEC 61010-2-020:2006 CE Marking (Conformité Européene) / CB Scheme 5
B 3-phase supply, overvoltage catogory, line-to-neutral and working voltage (IEC 61010-1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M IEC 61010-1 requirements for motors Other Medical Device Related Standards 0
B IEC 61010, Touch currents related to burns at higher frequencies (cl. 6.3, A.3) Other Medical Device Related Standards 0
B IEC 61010-1 - Voltage / Dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J Universal Power Cord with Adapters - IVD IEC 61010-2-101 CE Marking (Conformité Européene) / CB Scheme 6
E IEC 61010-1 Dielectric strength test of inner power supply Other ISO and International Standards and European Regulations 0
T List of Tests - IEC 61010-1, 3rd edition and IEC 60601-1, 3.1 edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G IEC 61010 - Single Fault Condition - Protective Impedance Implementation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K IEC 61010-2-100 Labeling question - IVD equipment Other ISO and International Standards and European Regulations 1
C IEC EN 61010 3rd Edition Safety Design Checklist Other ISO and International Standards and European Regulations 3
F IEC 61010 Creepage and Clearence Requirements Other ISO and International Standards and European Regulations 2
G IEC 61010 Humidity Preconditioning and Condensation CE Marking (Conformité Européene) / CB Scheme 4
O IEC 61010-1 Ed 2: "Automatic Disconnection of the supply" CE Marking (Conformité Européene) / CB Scheme 3
T Is IEC 61010 needed for a populated PCB within a unit? CE Marking (Conformité Européene) / CB Scheme 2
L Does IEC 61010-1 apply for Surgical Microscopes? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
L Temperature - IEC 61010-1 ed2 & ed3 at chapter 10 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
P IEC 61010 3rd Edition - When devices are listed to the 2nd edition CE Marking (Conformité Européene) / CB Scheme 5
S IEC 61010-1 and IEC 60601 - Borderline Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
B IEC 61010-1 Certification Requirements, and External Supply/ELV questions CE Marking (Conformité Européene) / CB Scheme 4
S Advantages of IEC 60601-1 over IEC 61010 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Is IEC 61010-1 required for my silly little device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 60601 vs. IEC 61010 - CE Mark for our IVD medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
R IEC 61010 vs IEC 60601-1 to ISO 15197 - Blood glucose monitoring device IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T CB Scheme IEC 61010 / IEC 60601...+ NRTL mark IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 62366 Hazard-related Use Scenario vs FDA Critical Task EU Medical Device Regulations 3
C Sample qualification to be IEC 60601-1 series tested IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H IEC 62368-1 Scope of Testing CE Marking (Conformité Européene) / CB Scheme 3
A IEC 18562-1 report EU Medical Device Regulations 5
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
I IEC 60812 or ISO 14971 for PFMEA? What should we use? ISO 14971 - Medical Device Risk Management 3
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
G History of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
H Pollution Degree (IEC 60664-1) CE Marking (Conformité Européene) / CB Scheme 2
T FDA level of concern vs IEC 62304 safety classification - 510(k) exempt device Other US Medical Device Regulations 13
J IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Y Software to Manage IEC 62304 Traceability Requirement IEC 62304 - Medical Device Software Life Cycle Processes 3
J IEC 60601-1 Clause 7.9.3.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 2
H Regarding Usability Engineering File according to IEC 62366-1 CE Marking (Conformité Européene) / CB Scheme 2
F Battery overload in IEC 60601-1 vs 62133-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Mandatory tests for IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
C Producing to Inactive ASTM and IEC Standards Manufacturing and Related Processes 11
K IEC Maximum Power Density for Muscle Stimulator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E Clause 8.9.1.2 of AMD2:2020 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
H Regarding confusion of the validity of the European implementation of IEC standards Other Medical Device Related Standards 1
P IEC 60601-1-2 RF wireless technology IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Should I buy IEC 62304:2006, IEC 62304:2006/AMD 1:2015 or both? EU Medical Device Regulations 1

Similar threads

Top Bottom