Hello everyone,

it's my first thread at this forum. :bigwave:

I am seeking tips related to IEC/EN 61010-1:2010 standard, clause 9.5 "Requirements for equipment containing or using flammable liquids".

The device is an electrical laboratory instrument that measures a property of liquid sample.

The simplest solution to bypass the clause would be to prohibit use of flammable sample liquids with the instrument. However, that is not ideal from business perspective. Sometimes customers will want to use liquids that are classified as flammable, such as ethanol.

Clause 9.5 offers three options to pass the clause:
a) temperature of sample liquid and materials contacting the liquid is limited to T+25 Celsius where T is fire point of liquid
b) quantity of liquid is limited to amount that does not cause spread of fire
c) fire enclosure

Option c) is ruled out at this point due to design. Option a) might be possible to implement as instruction for users, but liquid fire points are not generally listed at MSDS per my knowledge. Also this would rule out some flammable sample liquids, at least high concentrations of ethanol where the fire point is about 30 Celsius.

So I am left wondering how to verify conformance to option b). In a worst-case scenario the sample liquid volume could be 70 milli-litres, but that is more than enough for successful measurement. 10 milli-litres is a minimum needed for the measurement. Standard laboratory glass beakers are used as sample vessels.

Does anyone have experience with flammable liquids with similar instruments? How will a test laboratory verify conformance with clause 9.5b) ?

I highly appreciate any hints you might have.
 
Last edited:
Elsmar Forum Sponsor

Kevin Mader

One of THE Original Covers!
Staff member
Admin
#2
Hi ilkkahoo,

I don't have direct experience with your problem, but I've cleared a number of lab type equipment under IEC61010-1. My suggestion here is to contact your safety folks (UL, METLabs, etc....whoever you use) and talk it through with them. They have the experience and possible solutions (design) that might help you achieve what you desire relative to you customers' usages.

I've found them helpful over the years and most are willing to do the work upfront for no/little out of pocket expense (especially if you end up doing the testing/retesting with them in the near-term). And it's a great way to set the right expectations for all parties going forward.

Good luck!!

Kevin
 
Thread starter Similar threads Forum Replies Date
M IEC 61010-1 requirements for motors Other Medical Device Related Standards 0
B IEC 61010, Touch currents related to burns at higher frequencies (cl. 6.3, A.3) Other Medical Device Related Standards 0
B IEC 61010-1 - Voltage / Dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J Universal Power Cord with Adapters - IVD IEC 61010-2-101 CE Marking (Conformité Européene) / CB Scheme 6
E IEC 61010-1 Dielectric strength test of inner power supply Other ISO and International Standards and European Regulations 0
T List of Tests - IEC 61010-1, 3rd edition and IEC 60601-1, 3.1 edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G IEC 61010 - Single Fault Condition - Protective Impedance Implementation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K IEC 61010-2-100 Labeling question - IVD equipment Other ISO and International Standards and European Regulations 1
C IEC EN 61010 3rd Edition Safety Design Checklist Other ISO and International Standards and European Regulations 3
F IEC 61010 Creepage and Clearence Requirements Other ISO and International Standards and European Regulations 2
G IEC 61010 Humidity Preconditioning and Condensation CE Marking (Conformité Européene) / CB Scheme 4
O IEC 61010-1 Ed 2: "Automatic Disconnection of the supply" CE Marking (Conformité Européene) / CB Scheme 3
T Is IEC 61010 needed for a populated PCB within a unit? CE Marking (Conformité Européene) / CB Scheme 2
L Does IEC 61010-1 apply for Surgical Microscopes? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
L Temperature - IEC 61010-1 ed2 & ed3 at chapter 10 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
P IEC 61010 3rd Edition - When devices are listed to the 2nd edition CE Marking (Conformité Européene) / CB Scheme 5
S IEC 61010-1 and IEC 60601 - Borderline Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
B IEC 61010-1 Certification Requirements, and External Supply/ELV questions CE Marking (Conformité Européene) / CB Scheme 4
S Advantages of IEC 60601-1 over IEC 61010 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Is IEC 61010-1 required for my silly little device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 60601 vs. IEC 61010 - CE Mark for our IVD medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
R IEC 61010 vs IEC 60601-1 to ISO 15197 - Blood glucose monitoring device IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T CB Scheme IEC 61010 / IEC 60601...+ NRTL mark IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
G Adopting old product - compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 9
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
A IEC 62304 safety classification, External Controls and off-label use related risks IEC 62304 - Medical Device Software Life Cycle Processes 5
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C ISO/IEC 17021-1 clause 7.1.2 - Determination of competence criteria Document Control Systems, Procedures, Forms and Templates 2
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C ISO/ IEC 17021 Resource requirement (need help) Document Control Systems, Procedures, Forms and Templates 5
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
S IEC 62304 software costs and time Medical Device and FDA Regulations and Standards News 3
S IEC 62304 - Software verification cost IEC 62304 - Medical Device Software Life Cycle Processes 3
Sravan Manchikanti Software Risk Management & probability of occurrence as per IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
M IEC 62304 Software changes - Minor labeling changes on the GUI IEC 62304 - Medical Device Software Life Cycle Processes 3
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K IEC 62304 - Testing Independance IEC 62304 - Medical Device Software Life Cycle Processes 5
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Exclusions or justification for non-applicability of IEC standards Reliability Analysis - Predictions, Testing and Standards 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7

Similar threads

Top Bottom