IEC 61010 Creepage and Clearence Requirements

F

Freddybaby

We have a sub-assembly we supply that has a push button on the front panel that the users second party safety assessor is requiring UL listing (or other) to the internal working voltage of 40KVDC.

This switch is in the 24V control system but is in the vicinity of a "simple voltage divider" and if breakdown of the resistor string occurs at the time someone pushes the button it's a shock hazard.

There are NO conductive parts including the switch plunger that extend beyond the protective earth guard (the front panel) but the creepage from switch terminals to the users finger is not adequate for the working voltage.

Hi-pot testing has proven that all HV arcs are diverted to the front panel.

My question is: Can we use a suitable (UL listed type 5) voltage clamp on each side of switch to the earthed panel to get around the creepage needed for the full voltage.

In other words, the voltage should never rise to 40KVDC as the clamp will divert and all conductive parts of the switch assembly are behind the earthed panel.
 

Stijloor

Leader
Super Moderator
A Quick Bump!

Can someone help?

If your questions were answered here, please consider helping others.

Thank you very much!
 

Hershal

Metrologist-Auditor
Trusted Information Resource
I have assessed IEC 61010-1 and several -2s many times. Generally the applicable -2 and parent paragraph in the -1 is pretty clear. However, your question is specific. I would work with your customer, lay out the design and circuitry, and let the UL Engineers do their thing. I know UL well, and many of their Engineers. Trust their call, and abide by it. But give them what they need to make the call.

Hope this helps.
 
Thread starter Similar threads Forum Replies Date
H IEC 61010-1:2010/A1:2016 - Clause 10.5.3 - (b) - Interpretation CE Marking (Conformité Européene) / CB Scheme 6
H Regarding Double Insulation & Caution Symbol on Product according to IEC 61010-1 CE Marking (Conformité Européene) / CB Scheme 6
H IEC 61010-1 Clause No. 10.5 Resistance to Heat - Sub Clause 10.5.3 Insulating Material CE Marking (Conformité Européene) / CB Scheme 6
H Required explanation of sub clause 7.2.101.2 (Exception) of IEC 61010-2-020:2006 CE Marking (Conformité Européene) / CB Scheme 5
B 3-phase supply, overvoltage catogory, line-to-neutral and working voltage (IEC 61010-1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M IEC 61010-1 requirements for motors Other Medical Device Related Standards 0
B IEC 61010, Touch currents related to burns at higher frequencies (cl. 6.3, A.3) Other Medical Device Related Standards 0
B IEC 61010-1 - Voltage / Dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J Universal Power Cord with Adapters - IVD IEC 61010-2-101 CE Marking (Conformité Européene) / CB Scheme 6
E IEC 61010-1 Dielectric strength test of inner power supply Other ISO and International Standards and European Regulations 0
T List of Tests - IEC 61010-1, 3rd edition and IEC 60601-1, 3.1 edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G IEC 61010 - Single Fault Condition - Protective Impedance Implementation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
I IEC 61010-1 and Flammable Liquids Other ISO and International Standards and European Regulations 1
K IEC 61010-2-100 Labeling question - IVD equipment Other ISO and International Standards and European Regulations 1
C IEC EN 61010 3rd Edition Safety Design Checklist Other ISO and International Standards and European Regulations 3
G IEC 61010 Humidity Preconditioning and Condensation CE Marking (Conformité Européene) / CB Scheme 4
O IEC 61010-1 Ed 2: "Automatic Disconnection of the supply" CE Marking (Conformité Européene) / CB Scheme 3
T Is IEC 61010 needed for a populated PCB within a unit? CE Marking (Conformité Européene) / CB Scheme 2
L Does IEC 61010-1 apply for Surgical Microscopes? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
L Temperature - IEC 61010-1 ed2 & ed3 at chapter 10 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
P IEC 61010 3rd Edition - When devices are listed to the 2nd edition CE Marking (Conformité Européene) / CB Scheme 5
S IEC 61010-1 and IEC 60601 - Borderline Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
B IEC 61010-1 Certification Requirements, and External Supply/ELV questions CE Marking (Conformité Européene) / CB Scheme 4
S Advantages of IEC 60601-1 over IEC 61010 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Is IEC 61010-1 required for my silly little device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 60601 vs. IEC 61010 - CE Mark for our IVD medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
R IEC 61010 vs IEC 60601-1 to ISO 15197 - Blood glucose monitoring device IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T CB Scheme IEC 61010 / IEC 60601...+ NRTL mark IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K Leakage current tests for battery powered device with IEC 62368 compliant charger IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Defining the need to perform the Proximity magnetic fields immunity tests (IEC 61000-4-39) according to the IEC 60601-1-2 ed.4.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601-1-8 clause 6.4.1 (statistics of the distribution of the sum of the maximum ALARM CONDITION + SIGNAL GENERATION DELAY) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
O ESD Testing IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
T Class B IEC 62304 - 5.4 Software Detailed Design IEC 62304 - Medical Device Software Life Cycle Processes 4
E Meaning of ALARM PRESETS (subclause 6.5.1 of IEC 60601-1-8:2020) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Franky Zhang123 The Electrical Standards GB9706.1:2020 (same as IEC 60601-1:2012) etc. will be implemented on May 1st, 2023 in China. China Medical Device Regulations 0
D IEC 61000-3-2 applied at 120V and 100V IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H IEC 62368-1 used in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Artificial hand (IEC 60601-1-2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L Differences between IEC 60601-1-2 and IEC TR 60601-4-2 US Food and Drug Administration (FDA) 3
A IEC 62366-1:2015+A1:2020 IEC 62366 - Medical Device Usability Engineering 3
L EN 62304:2006/AC:2008 vs IEC 62304:2006/Amd 1:2015 IEC 62304 - Medical Device Software Life Cycle Processes 1
A IEC 62366-1:2015+A1:2020 IEC 62366 - Medical Device Usability Engineering 2
S PDF version of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P IEC 60601-2-25 requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Brexit IEC 61000-4-2 test point consideration IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
MaHoDie Is it possible to assign medical software to security class A (according to IEC 62304)? EU Medical Device Regulations 8
Richard Regalado Informational ISO/IEC 27001:2022 has been published IEC 27001 - Information Security Management Systems (ISMS) 0
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 62366 Hazard-related Use Scenario vs FDA Critical Task EU Medical Device Regulations 3
C Sample qualification to be IEC 60601-1 series tested IEC 60601 - Medical Electrical Equipment Safety Standards Series 2

Similar threads

Top Bottom