IEC 61010 Creepage and Clearence Requirements
F
Freddybaby
We have a sub-assembly we supply that has a push button on the front panel that the users second party safety assessor is requiring UL listing (or other) to the internal working voltage of 40KVDC.
This switch is in the 24V control system but is in the vicinity of a "simple voltage divider" and if breakdown of the resistor string occurs at the time someone pushes the button it's a shock hazard.
There are NO conductive parts including the switch plunger that extend beyond the protective earth guard (the front panel) but the creepage from switch terminals to the users finger is not adequate for the working voltage.
Hi-pot testing has proven that all HV arcs are diverted to the front panel.
My question is: Can we use a suitable (UL listed type 5) voltage clamp on each side of switch to the earthed panel to get around the creepage needed for the full voltage.
In other words, the voltage should never rise to 40KVDC as the clamp will divert and all conductive parts of the switch assembly are behind the earthed panel.
Stijloor
A Quick Bump!
Can someone help?
If your questions were answered here, please consider helping others.
Thank you very much!
Hershal
Metrologist-Auditor
I have assessed IEC 61010-1 and several -2s many times. Generally the applicable -2 and parent paragraph in the -1 is pretty clear. However, your question is specific. I would work with your customer, lay out the design and circuitry, and let the UL Engineers do their thing. I know UL well, and many of their Engineers. Trust their call, and abide by it. But give them what they need to make the call.
Hope this helps.
B 3-phase supply, overvoltage catogory, line-to-neutral and working voltage (IEC 61010-1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M IEC 61010-1 requirements for motors Other Medical Device Related Standards 0
B IEC 61010, Touch currents related to burns at higher frequencies (cl. 6.3, A.3) Other Medical Device Related Standards 0
B IEC 61010-1 - Voltage / Dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J Universal Power Cord with Adapters - IVD IEC 61010-2-101 CE Marking (Conformité Européene) / CB Scheme 6
E IEC 61010-1 Dielectric strength test of inner power supply Other ISO and International Standards and European Regulations 0
T List of Tests - IEC 61010-1, 3rd edition and IEC 60601-1, 3.1 edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G IEC 61010 - Single Fault Condition - Protective Impedance Implementation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
I IEC 61010-1 and Flammable Liquids Other ISO and International Standards and European Regulations 1
K IEC 61010-2-100 Labeling question - IVD equipment Other ISO and International Standards and European Regulations 1
C IEC EN 61010 3rd Edition Safety Design Checklist Other ISO and International Standards and European Regulations 3
G IEC 61010 Humidity Preconditioning and Condensation CE Marking (Conformité Européene) / CB Scheme 4
O IEC 61010-1 Ed 2: "Automatic Disconnection of the supply" CE Marking (Conformité Européene) / CB Scheme 3
T Is IEC 61010 needed for a populated PCB within a unit? CE Marking (Conformité Européene) / CB Scheme 2
L Does IEC 61010-1 apply for Surgical Microscopes? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
L Temperature - IEC 61010-1 ed2 & ed3 at chapter 10 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
P IEC 61010 3rd Edition - When devices are listed to the 2nd edition CE Marking (Conformité Européene) / CB Scheme 5
S IEC 61010-1 and IEC 60601 - Borderline Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
B IEC 61010-1 Certification Requirements, and External Supply/ELV questions CE Marking (Conformité Européene) / CB Scheme 4
S Advantages of IEC 60601-1 over IEC 61010 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Is IEC 61010-1 required for my silly little device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 60601 vs. IEC 61010 - CE Mark for our IVD medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
R IEC 61010 vs IEC 60601-1 to ISO 15197 - Blood glucose monitoring device IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T CB Scheme IEC 61010 / IEC 60601...+ NRTL mark IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E Test report to certify compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
B European Harmonized vs IEC standard IEC 62366 - Medical Device Usability Engineering 2
A ISO/DIS 15223-1:2020 - Country of manufacture label (IEC 60417 No. 6049) - Which national law requires this symbol? Other Medical Device Related Standards 0
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Reference to IEC 60950-1 in Amendment 2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Le Chiffre Online training available for ISO/IEC 17021-1: Requirements for bodies providing audit and certification of management systems Training - Internal, External, Online and Distance Learning 3
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 1
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T Relationship between ISO 9001 and ISO – IEC BS EN 870079- 34 2020 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
T Help with BS EN ISO - IEC 80079-34 2020 (Explosive atmospheres QMS) Other ISO and International Standards and European Regulations 0
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E IEC 62353 normalization IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
G Adopting old product - compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 9
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9