IEC 61010 - Single Fault Condition - Protective Impedance Implementation

G

guru.ganesheela

#1
I plan to test my equipment in single fault conditions as a part of pre-compliance testing. In that, the first condition is to create a protective impedance and run the equipment. I searched on the internet on how to create a protective impedance. But I didn't find any useful information. Can someone help me out?
 
Elsmar Forum Sponsor
M

MedMartin

#2
Hello,

what exactly do you want to test?
You can look at your circuit and identify single fault conditions and then test them.
Some examples would be short-circuits on connectors, heaters fully activated, processors in reset, ...
It depends on your device.

Best regards,
Martin
 
Thread starter Similar threads Forum Replies Date
H IEC 61010-1 Clause No. 10.5 Resistance to Heat - Sub Clause 10.5.3 Insulating Material CE Marking (Conformité Européene) / CB Scheme 6
H Required explanation of sub clause 7.2.101.2 (Exception) of IEC 61010-2-020:2006 CE Marking (Conformité Européene) / CB Scheme 5
B 3-phase supply, overvoltage catogory, line-to-neutral and working voltage (IEC 61010-1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M IEC 61010-1 requirements for motors Other Medical Device Related Standards 0
B IEC 61010, Touch currents related to burns at higher frequencies (cl. 6.3, A.3) Other Medical Device Related Standards 0
B IEC 61010-1 - Voltage / Dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J Universal Power Cord with Adapters - IVD IEC 61010-2-101 CE Marking (Conformité Européene) / CB Scheme 6
E IEC 61010-1 Dielectric strength test of inner power supply Other ISO and International Standards and European Regulations 0
T List of Tests - IEC 61010-1, 3rd edition and IEC 60601-1, 3.1 edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
I IEC 61010-1 and Flammable Liquids Other ISO and International Standards and European Regulations 1
K IEC 61010-2-100 Labeling question - IVD equipment Other ISO and International Standards and European Regulations 1
C IEC EN 61010 3rd Edition Safety Design Checklist Other ISO and International Standards and European Regulations 3
F IEC 61010 Creepage and Clearence Requirements Other ISO and International Standards and European Regulations 2
G IEC 61010 Humidity Preconditioning and Condensation CE Marking (Conformité Européene) / CB Scheme 4
O IEC 61010-1 Ed 2: "Automatic Disconnection of the supply" CE Marking (Conformité Européene) / CB Scheme 3
T Is IEC 61010 needed for a populated PCB within a unit? CE Marking (Conformité Européene) / CB Scheme 2
L Does IEC 61010-1 apply for Surgical Microscopes? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
L Temperature - IEC 61010-1 ed2 & ed3 at chapter 10 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
P IEC 61010 3rd Edition - When devices are listed to the 2nd edition CE Marking (Conformité Européene) / CB Scheme 5
S IEC 61010-1 and IEC 60601 - Borderline Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
B IEC 61010-1 Certification Requirements, and External Supply/ELV questions CE Marking (Conformité Européene) / CB Scheme 4
S Advantages of IEC 60601-1 over IEC 61010 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Is IEC 61010-1 required for my silly little device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 60601 vs. IEC 61010 - CE Mark for our IVD medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
R IEC 61010 vs IEC 60601-1 to ISO 15197 - Blood glucose monitoring device IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T CB Scheme IEC 61010 / IEC 60601...+ NRTL mark IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 62366 Hazard-related Use Scenario vs FDA Critical Task EU Medical Device Regulations 3
C Sample qualification to be IEC 60601-1 series tested IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H IEC 62368-1 Scope of Testing CE Marking (Conformité Européene) / CB Scheme 3
A IEC 18562-1 report EU Medical Device Regulations 5
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
I IEC 60812 or ISO 14971 for PFMEA? What should we use? ISO 14971 - Medical Device Risk Management 3
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
G History of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
H Pollution Degree (IEC 60664-1) CE Marking (Conformité Européene) / CB Scheme 2
T FDA level of concern vs IEC 62304 safety classification - 510(k) exempt device Other US Medical Device Regulations 13
J IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Y Software to Manage IEC 62304 Traceability Requirement IEC 62304 - Medical Device Software Life Cycle Processes 3
J IEC 60601-1 Clause 7.9.3.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 2
H Regarding Usability Engineering File according to IEC 62366-1 CE Marking (Conformité Européene) / CB Scheme 2
F Battery overload in IEC 60601-1 vs 62133-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Mandatory tests for IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
C Producing to Inactive ASTM and IEC Standards Manufacturing and Related Processes 11
K IEC Maximum Power Density for Muscle Stimulator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E Clause 8.9.1.2 of AMD2:2020 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
H Regarding confusion of the validity of the European implementation of IEC standards Other Medical Device Related Standards 1
P IEC 60601-1-2 RF wireless technology IEC 60601 - Medical Electrical Equipment Safety Standards Series 1

Similar threads

Top Bottom