I plan to test my equipment in single fault conditions as a part of pre-compliance testing. In that, the first condition is to create a protective impedance and run the equipment. I searched on the internet on how to create a protective impedance. But I didn't find any useful information. Can someone help me out?
 
Elsmar Forum Sponsor
M

MedMartin

#2
Hello,

what exactly do you want to test?
You can look at your circuit and identify single fault conditions and then test them.
Some examples would be short-circuits on connectors, heaters fully activated, processors in reset, ...
It depends on your device.

Best regards,
Martin
 
Thread starter Similar threads Forum Replies Date
M IEC 61010-1 requirements for motors Other Medical Device Related Standards 0
B IEC 61010, Touch currents related to burns at higher frequencies (cl. 6.3, A.3) Other Medical Device Related Standards 0
B IEC 61010-1 - Voltage / Dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J Universal Power Cord with Adapters - IVD IEC 61010-2-101 CE Marking (Conformité Européene) / CB Scheme 6
E IEC 61010-1 Dielectric strength test of inner power supply Other ISO and International Standards and European Regulations 0
T List of Tests - IEC 61010-1, 3rd edition and IEC 60601-1, 3.1 edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
I IEC 61010-1 and Flammable Liquids Other ISO and International Standards and European Regulations 1
K IEC 61010-2-100 Labeling question - IVD equipment Other ISO and International Standards and European Regulations 1
C IEC EN 61010 3rd Edition Safety Design Checklist Other ISO and International Standards and European Regulations 3
F IEC 61010 Creepage and Clearence Requirements Other ISO and International Standards and European Regulations 2
G IEC 61010 Humidity Preconditioning and Condensation CE Marking (Conformité Européene) / CB Scheme 4
O IEC 61010-1 Ed 2: "Automatic Disconnection of the supply" CE Marking (Conformité Européene) / CB Scheme 3
T Is IEC 61010 needed for a populated PCB within a unit? CE Marking (Conformité Européene) / CB Scheme 2
L Does IEC 61010-1 apply for Surgical Microscopes? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
L Temperature - IEC 61010-1 ed2 & ed3 at chapter 10 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
P IEC 61010 3rd Edition - When devices are listed to the 2nd edition CE Marking (Conformité Européene) / CB Scheme 5
S IEC 61010-1 and IEC 60601 - Borderline Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
B IEC 61010-1 Certification Requirements, and External Supply/ELV questions CE Marking (Conformité Européene) / CB Scheme 4
S Advantages of IEC 60601-1 over IEC 61010 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Is IEC 61010-1 required for my silly little device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 60601 vs. IEC 61010 - CE Mark for our IVD medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
R IEC 61010 vs IEC 60601-1 to ISO 15197 - Blood glucose monitoring device IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T CB Scheme IEC 61010 / IEC 60601...+ NRTL mark IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
A IEC 62304 safety classification, External Controls and off-label use related risks IEC 62304 - Medical Device Software Life Cycle Processes 5
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C ISO/IEC 17021-1 clause 7.1.2 - Determination of competence criteria Document Control Systems, Procedures, Forms and Templates 2
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C ISO/ IEC 17021 Resource requirement (need help) Document Control Systems, Procedures, Forms and Templates 5
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
S IEC 62304 software costs and time Medical Device and FDA Regulations and Standards News 3
S IEC 62304 - Software verification cost IEC 62304 - Medical Device Software Life Cycle Processes 3
Sravan Manchikanti Software Risk Management & probability of occurrence as per IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
M IEC 62304 Software changes - Minor labeling changes on the GUI IEC 62304 - Medical Device Software Life Cycle Processes 3
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K IEC 62304 - Testing Independance IEC 62304 - Medical Device Software Life Cycle Processes 5
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Exclusions or justification for non-applicability of IEC standards Reliability Analysis - Predictions, Testing and Standards 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7

Similar threads

Top Bottom