We are a medical device manufacturer using an IEC62133 certified, single cell Li-ion battery pack (not user accessible) for a portable device. The pack we are using is no longer going to be available and we are in the market for a replacement. I am able to find replacement packs that have IEC62133 certification for the individual cell but not the assembled pack.

Does anyone have experience as to whether certification on the cell only is sufficient for a device that is IEC 60601-1 compliant or does the assembled pack need to carry the certification?

Thanks
 
Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
Pretty sure it has to be cell & pack since the pack will likely contain any safety circuitry. If you use a test lab, discuss with them.
 
Thread starter Similar threads Forum Replies Date
M When is EN 62133-1 (new version i.e. IEC 62133:2017) mandatory in EU Other ISO and International Standards and European Regulations 0
M IEC 62133:2012 is now IEC 62133-1:2017 & IEC 62133-2:2017 Other ISO and International Standards and European Regulations 0
S IEC 62133 - CB Scheme Required by FDA 510k Other Medical Device Related Standards 3
K IEC 62133 Test Report for Lithium Battery Certification IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T IEC 62133 Certified Battery Manufacturers Other Medical Device Related Standards 3
R IEC 62133 to satisfy 60601-1 3rd edition - What happen to 60086-4 Other Medical Device Related Standards 18
W When will Kuwait require Lithium Ion Batteries be tested to IEC 62133 standard Other Medical Device Regulations World-Wide 1
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 62471 vs IEC/TR 62778 CE Marking (Conformité Européene) / CB Scheme 0
B IEC 61010, Touch currents related to burns at higher frequencies (cl. 6.3, A.3) Other Medical Device Related Standards 0
M IEC 61000-4-2: Requirements for coating declared as insulating IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 7
B IEC 61010-1 - Voltage / Dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
adir88 Amendment 1 of IEC 62366-1 - Details yet? IEC 62366 - Medical Device Usability Engineering 0
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
B Has anyone done an IEC 60601-1 gap analysis to IEC 60335? Medical Device and FDA Regulations and Standards News 4
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
K Primary Operating Functions (POF) in IEC 62366-1:2015. IEC 62366 - Medical Device Usability Engineering 5
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 62304 Class A Project IEC 62304 - Medical Device Software Life Cycle Processes 15
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 4
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
MDD_QNA QR Code Standard ISO/IEC 15417:2007 - Does anyone use it? Other Medical Device Related Standards 3
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Universal Power Cord with Adapters - IVD IEC 61010-2-101 CE Marking (Conformité Européene) / CB Scheme 6
V Applicability of IEC 62366-1 usability to resorbable bone substitutes IEC 62366 - Medical Device Usability Engineering 9
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 3
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
V IS/ISO/IEC 17025:2017 Clause 7, sub clause 7.11 Control of data and information management ISO 17025 related Discussions 1
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
A IEC 60601-1 and IEC 60601-2-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601-1-2 EMC Tables IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Bookmarking my security protected IEC 60601-1 .pdf file IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
Similar threads


















































Top Bottom