IEC 62133 Test Report for Lithium Battery Certification

  • Thread starter karthika subramaniam
  • Start date
K

karthika subramaniam

One of our product is under IEC 60601-1 testing, in which we are using android device (Mobile).

We have developed android app which is installed on android device for controlling medical devices motions.

We have provided the Android device certification for IEC 60950-1 for IT equipment.

Regulatory Board, Test Lab is asking for IEC 62133 test report for Lithium battery and also USB charging cable certification that comes along with android device.

We have completed EMC/I testing. If we change MSO, there will be impact on EMC/I testing as well.


Please help
 

Mark Meer

Trusted Information Resource
What is "MSO"?

I presume that you've asked the supplier of your android device what kinds of documentation they can provide?

The USB cable will likely have a UL-listing (nothing to do with IEC 62133).

As for the battery, it might be a problem, as older consumer batteries may not have been tested IEC 62133 (instead, more likely something like UL1642).

If they DON'T have IEC 62133 testing done, there's a few options I can think of:

1. Negotiate with them (e.g. pay them) to get their batteries tested.
2. Find a battery supplier that can provide a battery with the same specification that you can have tested yourself. This is onerous, but has the advantage that you are in control of the battery.
3. Justify in your risk management documentation that whatever battery standards your current supplier uses is sufficient to ensure risk is acceptable. You can then justify not meeting that requirement of IEC 60601-1. In this case you wouldn't be claiming full compliance to the standard, but the tests can still be used as evidence for safety.

Best of luck!
MM
 
K

karthika subramaniam

Hi Mark

MSO - Multiple Socket Outlet

Regarding 3) How can we address in RMF, can you throw some light?

Thank you
 

Pads38

Moderator
I would suggest that the Android device is an accessory to the medical device and is NOT itself a medical device.

The 'app' should be considered a medical device so should satisfy the requirements of 60601-1 Clause 14 and 62304 (lifecycle).

There should be no requirement for the battery in the Android device.
 
Thread starter Similar threads Forum Replies Date
F Battery overload in IEC 60601-1 vs 62133-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M When is EN 62133-1 (new version i.e. IEC 62133:2017) mandatory in EU Other ISO and International Standards and European Regulations 3
M IEC 62133:2012 is now IEC 62133-1:2017 & IEC 62133-2:2017 Other ISO and International Standards and European Regulations 0
F IEC 62133 cell vs. Battery Pack - Certification on the cell only IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S IEC 62133 - CB Scheme Required by FDA 510k Other Medical Device Related Standards 3
T IEC 62133 Certified Battery Manufacturers Other Medical Device Related Standards 3
R IEC 62133 to satisfy 60601-1 3rd edition - What happen to 60086-4 Other Medical Device Related Standards 18
W When will Kuwait require Lithium Ion Batteries be tested to IEC 62133 standard Other Medical Device Regulations World-Wide 1
K Leakage current tests for battery powered device with IEC 62368 compliant charger IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Defining the need to perform the Proximity magnetic fields immunity tests (IEC 61000-4-39) according to the IEC 60601-1-2 ed.4.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601-1-8 clause 6.4.1 (statistics of the distribution of the sum of the maximum ALARM CONDITION + SIGNAL GENERATION DELAY) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
O ESD Testing IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
T Class B IEC 62304 - 5.4 Software Detailed Design IEC 62304 - Medical Device Software Life Cycle Processes 4
E Meaning of ALARM PRESETS (subclause 6.5.1 of IEC 60601-1-8:2020) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Franky Zhang123 The Electrical Standards GB9706.1:2020 (same as IEC 60601-1:2012) etc. will be implemented on May 1st, 2023 in China. China Medical Device Regulations 0
D IEC 61000-3-2 applied at 120V and 100V IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H IEC 62368-1 used in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Artificial hand (IEC 60601-1-2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L Differences between IEC 60601-1-2 and IEC TR 60601-4-2 US Food and Drug Administration (FDA) 3
A IEC 62366-1:2015+A1:2020 IEC 62366 - Medical Device Usability Engineering 3
L EN 62304:2006/AC:2008 vs IEC 62304:2006/Amd 1:2015 IEC 62304 - Medical Device Software Life Cycle Processes 1
A IEC 62366-1:2015+A1:2020 IEC 62366 - Medical Device Usability Engineering 2
H IEC 61010-1:2010/A1:2016 - Clause 10.5.3 - (b) - Interpretation CE Marking (Conformité Européene) / CB Scheme 6
H Regarding Double Insulation & Caution Symbol on Product according to IEC 61010-1 CE Marking (Conformité Européene) / CB Scheme 6
S PDF version of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P IEC 60601-2-25 requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Brexit IEC 61000-4-2 test point consideration IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
MaHoDie Is it possible to assign medical software to security class A (according to IEC 62304)? EU Medical Device Regulations 8
Richard Regalado Informational ISO/IEC 27001:2022 has been published IEC 27001 - Information Security Management Systems (ISMS) 0
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 62366 Hazard-related Use Scenario vs FDA Critical Task EU Medical Device Regulations 3
C Sample qualification to be IEC 60601-1 series tested IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H IEC 62368-1 Scope of Testing CE Marking (Conformité Européene) / CB Scheme 3
A IEC 18562-1 report EU Medical Device Regulations 5
H IEC 61010-1 Clause No. 10.5 Resistance to Heat - Sub Clause 10.5.3 Insulating Material CE Marking (Conformité Européene) / CB Scheme 6
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
I IEC 60812 or ISO 14971 for PFMEA? What should we use? ISO 14971 - Medical Device Risk Management 3
H Required explanation of sub clause 7.2.101.2 (Exception) of IEC 61010-2-020:2006 CE Marking (Conformité Européene) / CB Scheme 5
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
G History of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
H Pollution Degree (IEC 60664-1) CE Marking (Conformité Européene) / CB Scheme 2
T FDA level of concern vs IEC 62304 safety classification - 510(k) exempt device Other US Medical Device Regulations 13
J IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Y Software to Manage IEC 62304 Traceability Requirement IEC 62304 - Medical Device Software Life Cycle Processes 3
J IEC 60601-1 Clause 7.9.3.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 2
H Regarding Usability Engineering File according to IEC 62366-1 CE Marking (Conformité Européene) / CB Scheme 2
S Mandatory tests for IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
C Producing to Inactive ASTM and IEC Standards Manufacturing and Related Processes 11

Similar threads

Top Bottom