IEC 62304 5.2.2 f - Installation and acceptance of device software
E
El Guapo
#1
I am confused about section f) of IEC 62304 5.2.2:
As appropriate to the MEDICAL DEVICE SOFTWARE, the MANUFACTURER shall include in the
software requirements: f) installation and acceptance requirements of the delivered MEDICAL DEVICE SOFTWARE at the operation and maintenance site or sites;
Does that mean we just need a simple statement about the required capabilities of the installation target machine? What about the acceptance requirement for installation -- or is that more geared for contractors?
We've just got some imaging software for retinal imaging. Can run with or without cameras attached to the computer. Comes on a data stick we provide or as a downloaded -- same package does a fresh install or an upgrade. Can't see how this part of the standard applies beyond "Do you have XP/Win7, 4g RAM, 20g free disk?". Maybe "User must comprehend English."?
Elsmar Forum Sponsor
c.mitch
Quite Involved in Discussions
#2
Yes, I would say so.
Your acceptance requirements may look like those you wrote at the end of your post: Win XP and so on.
The acceptance requirements may be tests to:
-demonstrate that the installation of your software runs on this kind of machine + OS (testing the installer bundle if you have one, or testing the installation procedure).
-demonstrate that it works on such platform (test of main functions of software)
The info to provide to end-users/contractors may be the minimal HW requirements to have your SW run on their machines.
Thread starter Similar threads Forum Replies Date
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 7
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
M IEC 62304 Class A Project IEC 62304 - Medical Device Software Life Cycle Processes 15
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 4
P IEC 62304 - evaluation of integration and system testing IEC 62304 - Medical Device Software Life Cycle Processes 4
P Risk acceptability alignment between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 6
D Required Checklist Showing Compliance to IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 11
P Proposed revision of IEC 62304 - 2019 IEC 62304 - Medical Device Software Life Cycle Processes 6
S Relationship between IEC 62304 problem resolution and ISO 13485 IEC 62304 - Medical Device Software Life Cycle Processes 8
P IEC 62304:2006 A1:2015 - Software from the early 1990s IEC 62304 - Medical Device Software Life Cycle Processes 4
B IEC 62304:2015 vs IEC 62304:2006 + AMD1 IEC 62304 - Medical Device Software Life Cycle Processes 4
F IEC 62304 - Segregation and communication between software items IEC 62304 - Medical Device Software Life Cycle Processes 1
B Class IIB Device - IEC 62304 Software Classification IEC 62304 - Medical Device Software Life Cycle Processes 13
B IEC 62304 - Update Checklist IEC 62304 - Medical Device Software Life Cycle Processes 2
L Connection between IEC 62304 and Chapter 14 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 62304 - Develop an Architecture for the Interfaces of Software Items IEC 62304 - Medical Device Software Life Cycle Processes 8
S Does IEC 62304 require documenting unresolved anomalies for all safety classes? IEC 62304 - Medical Device Software Life Cycle Processes 4
A SOP for software validation of software in medical device IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 5
T I need to make test reports according IEC 62304 & IEC 62366 IEC 62366 - Medical Device Usability Engineering 2
D Changing software classification via software - IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
D Software as risk control - Confused on one aspect of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 20
K Trying to figure out what satisfies a few aspects of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 2
Y IEC 62304 Section 4.3(a) - 100% probability of failure IEC 62304 - Medical Device Software Life Cycle Processes 3
Y Application of IEC/EN 62304 at an advanced stage of software development IEC 62304 - Medical Device Software Life Cycle Processes 4
T Is there any requirement to be compliant with IEC 62304 while implementing ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 5
L Documentation Planning - IEC 62304 Clause 5.1.8 IEC 62304 - Medical Device Software Life Cycle Processes 2
C Software for Medical Devices - Requirements Content for compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 1
W CPU BIST IEC 62304 - Embedded code has CPU instruction tests IEC 62304 - Medical Device Software Life Cycle Processes 2
K IEC 62304 Amd 1 2015 - Figure 3 – Assigning Software Safety Classification IEC 62304 - Medical Device Software Life Cycle Processes 11
K Risk Reduction by Risk Control: IEC:62304-Class C ISO 14971 - Medical Device Risk Management 15
C Per IEC 62304, are DHF documents Configuration Items? IEC 62304 - Medical Device Software Life Cycle Processes 5
P IEC 62304 AMD1:2015: What's new vs.the 2006 Edition? IEC 62304 - Medical Device Software Life Cycle Processes 4
F FDA PMK 510(k) - IEC 62304 Software Components Segregation Other US Medical Device Regulations 3
M IEC 62304 Applicability - GUI Control Software IEC 62304 - Medical Device Software Life Cycle Processes 3
B Our NB says that IEC 62304 is an ISO 14971 Requirement ISO 14971 - Medical Device Risk Management 1
B Clarification on interpretation of some EN ISO 14971:2012 & IEC 62304:2006 req's ISO 14971 - Medical Device Risk Management 46
H ISO 14971 vs. IEC 62304 vs. 98/79/EC vs. ISO 13485 (Software Medical Device) ISO 14971 - Medical Device Risk Management 1
D A desperate call for help - IEC 62304 software IEC 62304 - Medical Device Software Life Cycle Processes 5
B IEC 62304:2006/AMD1:2015 Changes for Class A Software IEC 62304 - Medical Device Software Life Cycle Processes 3
M IEC 62304, ISO 14971 and FDA Medical Device SW Guidance 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
K IEC 62304 - Compliance steps IEC 62304 - Medical Device Software Life Cycle Processes 2
K ISO 14971 and IEC 62304 - Medical Device Software House ISO 14971 - Medical Device Risk Management 9
S Software Test Report including IEC 62304 classification IEC 62304 - Medical Device Software Life Cycle Processes 4
A Mapping of IEC 62304 artefacts (SRS, SAD, etc) to the 820.30 phases IEC 62304 - Medical Device Software Life Cycle Processes 5
W IEC 62304 vs. IMDRF SaMD Guideline Risk Class IEC 62304 - Medical Device Software Life Cycle Processes 5
C New IEC/TR 80002-3 Guidance for IEC 62304 - June 2014 IEC 62304 - Medical Device Software Life Cycle Processes 2
R IEC 62304 was brought up during an FDA Inspection/Audit IEC 62304 - Medical Device Software Life Cycle Processes 6
O Electronic Fever Thermometer - Why not IEC 62304 Class C? IEC 62304 - Medical Device Software Life Cycle Processes 7
R IEC 62304 - Medical Device Wi-Fi Radio Requirements IEC 62304 - Medical Device Software Life Cycle Processes 5
A Dental Laser Certification IEC 62304 for a Firmware IEC 62304 - Medical Device Software Life Cycle Processes 1
Similar threads
Top Bottom