I'm writing an SRS document for a class A medical device according to IEC 62304. I've wrote all the software requirements and I'm stuck in the verification section.
What am I supposed to write there?
Do I write the actual tests for each requirement, or do I just write a short description on how the test will be performed?
yodon
Presume you mean 5.2.6?
If you look at what the standard is addressing for this activity, I believe it's more review-based. Here are snippets from the list:
a) implement parent requirements (including risk mitigations allocated to software) ;
b) aren't contradictory (within);
c) are unambiguous;
d) are verifiable (testable);
e) are uniquely identified; and
f) are traceable to parent specifications.
This is all pretty review-based, IMO. Of course, you need to have things in place like unique identification and a trace back to the parent specifications. If you take a top-down look from the parent (which may be implemented in multiple software requirements), you determine if the sum of software requirements (fully) implement the parent specification. The review would also confirm no contradictions and the ABILITY to verify (through test) the requirement (e.g., having a requirement that the software be "fast" is not verifiable).
Marcelo
This is realted to a requirements engineering process, if you use a formal one, you probably is already doing everything the standards is requiring.
If not (which seems to be your case), you need to verify that each requirement fulfills the requirements in 5.2.6.
Itns a and f are the more complicated because they require that you trace the software requirements to the device (system) requirements, including risk control measures implemented by software, so you need to link the software requirements to the device requirements and device risk management.
rumplefish
Thank you for your replies.
I'm kind of new at this so you'll have to excuse me if I ask stupid questions
:D
So I've wrote all the software/system (I'm not sure that there's any difference in my case) requirements, I went over the risk analysis document and added that to the requirement, and now I have to verify that the requirements are up to the standard.
I don't have any higher level requirements to trace my requirements to, except maybe the risk analysis. Do I need to list the requirements and specify from where they were derived?
Also, I assume we'll do a review on the requirements, does this mean a seperate document?
Sorry:eek:
R IEC 62304 Section 5.3.6 Verify Software Architecture IEC 62304 - Medical Device Software Life Cycle Processes 4
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 1
G Adopting old product - compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 9
A IEC 62304 safety classification, External Controls and off-label use related risks IEC 62304 - Medical Device Software Life Cycle Processes 5
S IEC 62304 software costs and time Medical Device and FDA Regulations and Standards News 3
S IEC 62304 - Software verification cost IEC 62304 - Medical Device Software Life Cycle Processes 3
Sravan Manchikanti Software Risk Management & probability of occurrence as per IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
M IEC 62304 Software changes - Minor labeling changes on the GUI IEC 62304 - Medical Device Software Life Cycle Processes 3
K IEC 62304 - Testing Independance IEC 62304 - Medical Device Software Life Cycle Processes 5
K IEC 62304 - Functional and performance requirements for SOUP items IEC 62304 - Medical Device Software Life Cycle Processes 2
K IEC 62304 compliance - Code reviews as part of verification strategy IEC 62304 - Medical Device Software Life Cycle Processes 5
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
M IEC 62304 Class A Project IEC 62304 - Medical Device Software Life Cycle Processes 15
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 5
P SOUP anomaly evaluation for MMA (Mobile Medical Application) IEC 62304 clause 7.1.3 IEC 62304 - Medical Device Software Life Cycle Processes 6
P IEC 62304 - evaluation of integration and system testing IEC 62304 - Medical Device Software Life Cycle Processes 4
P Risk acceptability alignment between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 6
D Required Checklist Showing Compliance to IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 11
P Proposed revision of IEC 62304 - 2019 IEC 62304 - Medical Device Software Life Cycle Processes 6
S Relationship between IEC 62304 problem resolution and ISO 13485 IEC 62304 - Medical Device Software Life Cycle Processes 8
P IEC 62304:2006 A1:2015 - Software from the early 1990s IEC 62304 - Medical Device Software Life Cycle Processes 4
B IEC 62304:2015 vs IEC 62304:2006 + AMD1 IEC 62304 - Medical Device Software Life Cycle Processes 4
F IEC 62304 - Segregation and communication between software items IEC 62304 - Medical Device Software Life Cycle Processes 1
B Class IIB Device - IEC 62304 Software Classification IEC 62304 - Medical Device Software Life Cycle Processes 13
B IEC 62304 - Update Checklist IEC 62304 - Medical Device Software Life Cycle Processes 2
L Connection between IEC 62304 and Chapter 14 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 62304 - Develop an Architecture for the Interfaces of Software Items IEC 62304 - Medical Device Software Life Cycle Processes 8
S Does IEC 62304 require documenting unresolved anomalies for all safety classes? IEC 62304 - Medical Device Software Life Cycle Processes 4
A SOP for software validation of software in medical device IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 5
T I need to make test reports according IEC 62304 & IEC 62366 IEC 62366 - Medical Device Usability Engineering 2
D Changing software classification via software - IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
D Software as risk control - Confused on one aspect of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 20
K Trying to figure out what satisfies a few aspects of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 2
Y IEC 62304 Section 4.3(a) - 100% probability of failure IEC 62304 - Medical Device Software Life Cycle Processes 3
Y Application of IEC/EN 62304 at an advanced stage of software development IEC 62304 - Medical Device Software Life Cycle Processes 4
T Is there any requirement to be compliant with IEC 62304 while implementing ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 5
L Documentation Planning - IEC 62304 Clause 5.1.8 IEC 62304 - Medical Device Software Life Cycle Processes 2
C Software for Medical Devices - Requirements Content for compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 1
W CPU BIST IEC 62304 - Embedded code has CPU instruction tests IEC 62304 - Medical Device Software Life Cycle Processes 2
K IEC 62304 Amd 1 2015 - Figure 3 – Assigning Software Safety Classification IEC 62304 - Medical Device Software Life Cycle Processes 11
K Risk Reduction by Risk Control: IEC:62304-Class C ISO 14971 - Medical Device Risk Management 15
C Per IEC 62304, are DHF documents Configuration Items? IEC 62304 - Medical Device Software Life Cycle Processes 8
P IEC 62304 AMD1:2015: What's new vs.the 2006 Edition? IEC 62304 - Medical Device Software Life Cycle Processes 4
F FDA PMK 510(k) - IEC 62304 Software Components Segregation Other US Medical Device Regulations 3
M IEC 62304 Applicability - GUI Control Software IEC 62304 - Medical Device Software Life Cycle Processes 3
B Our NB says that IEC 62304 is an ISO 14971 Requirement ISO 14971 - Medical Device Risk Management 1
B Clarification on interpretation of some EN ISO 14971:2012 & IEC 62304:2006 req's ISO 14971 - Medical Device Risk Management 46
H ISO 14971 vs. IEC 62304 vs. 98/79/EC vs. ISO 13485 (Software Medical Device) ISO 14971 - Medical Device Risk Management 1
D A desperate call for help - IEC 62304 software IEC 62304 - Medical Device Software Life Cycle Processes 5