IEC 62304 - evaluation of integration and system testing
PaulG
In IEC 62304:2006/AMD1:2015, the updated clause 5.6.5 simply states "The manufacturer shall evaluate the integration test procedures for adequacy." Based on the standard's definition of "evaluation" (a systematic determination of the extent to which an entity meets its specified criteria) does this mean that each test procedure should be run with sample cases to demonstrate the outcomes are as expected? Or is a qualitative assessment/rationale for the test sufficient?
Similarly, 5.7.4 states that "The manufacturer shall evaluate the adequacy of verification strategies and test procedures." I'd be interested to hear how people have interpreted and satisfied these requirements in practice.
kreid
I think this can be covered by evidence that the test procedures have been reviewed and approved before running the tests.
I have seen several instances where this is not done and reviews only take place on the completed test results document.
yodon
I mostly agree with @kreid but we don't typically approve (in the traditional sense) integration tests. We do review them "for adequacy" and document the review (with a conclusion that they are adequate... with justification). We also do the same for software system test protocols although they do go through a formal review / approval process). You could argue that the review process covers requirement for appropriateness but we expand the documented review to assert that the protocol will drive capture of the required test record contents (5.7.5).
PaulG
OK, this makes sense. We are planning to capture the testing in a test plan document that will be reviewed and approved beforehand, so I think this will cover it. Thanks for everyone's feedback.
Tidge
In my experience, less mature (i.e. reactionary) ME software development teams struggle with the concept of integration testing more than any other element of 62304. If a team struggles with this concept, it can look like they aren't living up to this particular part of the standard. The folksy advice I offer about integration testing is that it shouldn't appear to be nearly indistinguishable from either unit or system-level testing, and that the purpose of the testing is to identify defects that aren't otherwise found at the unit level before they become issues at the system level.
I have found that a formal review and approval of the integration tests prior to execution (as suggested by kreid) to both satisfy the OP's concern and leads to more useful testing, at all levels. I typically don't expect test plans to go into the level of detail necessary to demonstrate adequacy of the specific tests at the discrete levels. For example, if the ME software was software safety class C: the test plan may be as simple as identifying the need for unit, integration, and system level testing of the software, and then describing the general approaches to be used at the different levels. This would be sufficient to begin the planning process of dedicating resources, etc. but would not require revision when issues begin to crop up in more specified detail test protocols.
If the software development plan which governs a specific software package (with a previously approved, dedicated test plan) has prescribed which tests are to be executed during integration and has predetermined the appropriateness of those tests, then I feel like the development team wouldn't have to revisit those formally prescribed tests. My experience with software development and testing is that it is often ad hoc and/or the development team bristles at "being told the way to do things", so the formal review and approval of test protocols is usually necessary.
P SOUP anomaly evaluation for MMA (Mobile Medical Application) IEC 62304 clause 7.1.3 IEC 62304 - Medical Device Software Life Cycle Processes 6
E Test report to certify compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 1
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 1
G Adopting old product - compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 9
A IEC 62304 safety classification, External Controls and off-label use related risks IEC 62304 - Medical Device Software Life Cycle Processes 5
S IEC 62304 software costs and time Medical Device and FDA Regulations and Standards News 3
S IEC 62304 - Software verification cost IEC 62304 - Medical Device Software Life Cycle Processes 3
Sravan Manchikanti Software Risk Management & probability of occurrence as per IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
M IEC 62304 Software changes - Minor labeling changes on the GUI IEC 62304 - Medical Device Software Life Cycle Processes 3
K IEC 62304 - Testing Independance IEC 62304 - Medical Device Software Life Cycle Processes 5
K IEC 62304 - Functional and performance requirements for SOUP items IEC 62304 - Medical Device Software Life Cycle Processes 2
K IEC 62304 compliance - Code reviews as part of verification strategy IEC 62304 - Medical Device Software Life Cycle Processes 5
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
M IEC 62304 Class A Project IEC 62304 - Medical Device Software Life Cycle Processes 15
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 5
P Risk acceptability alignment between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 6
D Required Checklist Showing Compliance to IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 11
P Proposed revision of IEC 62304 - 2019 IEC 62304 - Medical Device Software Life Cycle Processes 6
S Relationship between IEC 62304 problem resolution and ISO 13485 IEC 62304 - Medical Device Software Life Cycle Processes 8
P IEC 62304:2006 A1:2015 - Software from the early 1990s IEC 62304 - Medical Device Software Life Cycle Processes 4
B IEC 62304:2015 vs IEC 62304:2006 + AMD1 IEC 62304 - Medical Device Software Life Cycle Processes 4
F IEC 62304 - Segregation and communication between software items IEC 62304 - Medical Device Software Life Cycle Processes 1
B Class IIB Device - IEC 62304 Software Classification IEC 62304 - Medical Device Software Life Cycle Processes 13
B IEC 62304 - Update Checklist IEC 62304 - Medical Device Software Life Cycle Processes 2
L Connection between IEC 62304 and Chapter 14 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 62304 - Develop an Architecture for the Interfaces of Software Items IEC 62304 - Medical Device Software Life Cycle Processes 8
S Does IEC 62304 require documenting unresolved anomalies for all safety classes? IEC 62304 - Medical Device Software Life Cycle Processes 4
A SOP for software validation of software in medical device IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 5
T I need to make test reports according IEC 62304 & IEC 62366 IEC 62366 - Medical Device Usability Engineering 2
D Changing software classification via software - IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
D Software as risk control - Confused on one aspect of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 20
K Trying to figure out what satisfies a few aspects of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 2
Y IEC 62304 Section 4.3(a) - 100% probability of failure IEC 62304 - Medical Device Software Life Cycle Processes 3
Y Application of IEC/EN 62304 at an advanced stage of software development IEC 62304 - Medical Device Software Life Cycle Processes 4
T Is there any requirement to be compliant with IEC 62304 while implementing ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 5
L Documentation Planning - IEC 62304 Clause 5.1.8 IEC 62304 - Medical Device Software Life Cycle Processes 2
C Software for Medical Devices - Requirements Content for compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 1
W CPU BIST IEC 62304 - Embedded code has CPU instruction tests IEC 62304 - Medical Device Software Life Cycle Processes 2
K IEC 62304 Amd 1 2015 - Figure 3 – Assigning Software Safety Classification IEC 62304 - Medical Device Software Life Cycle Processes 11
K Risk Reduction by Risk Control: IEC:62304-Class C ISO 14971 - Medical Device Risk Management 15
C Per IEC 62304, are DHF documents Configuration Items? IEC 62304 - Medical Device Software Life Cycle Processes 8
P IEC 62304 AMD1:2015: What's new vs.the 2006 Edition? IEC 62304 - Medical Device Software Life Cycle Processes 4
F FDA PMK 510(k) - IEC 62304 Software Components Segregation Other US Medical Device Regulations 3
M IEC 62304 Applicability - GUI Control Software IEC 62304 - Medical Device Software Life Cycle Processes 3
B Our NB says that IEC 62304 is an ISO 14971 Requirement ISO 14971 - Medical Device Risk Management 1
B Clarification on interpretation of some EN ISO 14971:2012 & IEC 62304:2006 req's ISO 14971 - Medical Device Risk Management 46
H ISO 14971 vs. IEC 62304 vs. 98/79/EC vs. ISO 13485 (Software Medical Device) ISO 14971 - Medical Device Risk Management 1
D A desperate call for help - IEC 62304 software IEC 62304 - Medical Device Software Life Cycle Processes 5
B IEC 62304:2006/AMD1:2015 Changes for Class A Software IEC 62304 - Medical Device Software Life Cycle Processes 3