IEC 62304 - Medical Device Wi-Fi Radio Requirements
RickQualityGuy
The company I work for designs wi-fi radios that we sell to a broad spectrum of customers. Some of these customers manufacture medical devices such as IV Infusion Pumps which our Wi-Fi radios are inserted into as a sub-componet. Our Wi-Fi radios have software embedded in them such as driver code and radio configuration utilities.
My question is should we be compliant with IEC 62304 in this context and/or should our medical device manufactuer customers expect us to be compliant?
Also, what would be the consensus be for the risk classification for this type of application:
Class A - minor effect after failure
Class B - moderate effect after failure
Class C - severe risk after failure
Thank you very much.
c.mitch
If I understand your message, your company isn't a manufacturer of MD and as such isn't subject neither to MD regulations, nor to standards like 62304.
-->My question is should we be compliant with IEC 62304 in this context
Yes, you may, on a voluntary basis. for your company this will be a marketing advantage compared to your competitors.
-->and/or should our medical device manufactuer customers expect us to be compliant?
Yes, they may ask you to be compliant. Again, it may be a good argument to say "yes, we are compiliant" to win a contract.
-->Also, what would be the consensus be for the risk classification for this type of application
We can't answer this question. It depends on too many factors.
I'd rather say it's the responsibility of your customers to assess the software safety class.
FYI I've seen components manufacturers declaring their components are developped with class C, to be able to answer any quote requested by their customers.
Cheers
Mitch.
MIREGMGR
I agree with C.Mitch's comments.
(...) risk classification for this type of application
The risk classification for a component depends on that of the device in which it is used, therefore is not a function of the component per se.
In the described application, malfunction of an infusion pump in some contexts could result in death. The particular range of applications of your customer's infusion pumps might be limited so that such a risk level would not apply, though.
A SOP for software validation of software in medical device IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 5
C Software for Medical Devices - Requirements Content for compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 1
H ISO 14971 vs. IEC 62304 vs. 98/79/EC vs. ISO 13485 (Software Medical Device) ISO 14971 - Medical Device Risk Management 1
M IEC 62304, ISO 14971 and FDA Medical Device SW Guidance 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
K ISO 14971 and IEC 62304 - Medical Device Software House ISO 14971 - Medical Device Risk Management 9
A Requirements for Software to be CE Mark Compliant - IEC 62304 Medical Device Software IEC 62304 - Medical Device Software Life Cycle Processes 14
S IEC 62304 Clause 6. Medical Device Software Maintenance Plan IEC 62304 - Medical Device Software Life Cycle Processes 3
S One function, many devices? IEC 62304 and Medical Device Imaging Software IEC 62304 - Medical Device Software Life Cycle Processes 9
sagai Process Model for IEC 62304 - Medical Device Software Life Cycle IEC 62304 - Medical Device Software Life Cycle Processes 4
T IEC 62304:2006: Medical device software SDLC- CE Vs. Fda 510(k) submission IEC 62304 - Medical Device Software Life Cycle Processes 16
W IEC 62304:2006 Medical device software - Software life cycle processes - Issued IEC 62304 - Medical Device Software Life Cycle Processes 10
K IEC 62304 - Functional and performance requirements for SOUP items IEC 62304 - Medical Device Software Life Cycle Processes 2
K IEC 62304 compliance - Code reviews as part of verification strategy IEC 62304 - Medical Device Software Life Cycle Processes 5
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
M IEC 62304 Class A Project IEC 62304 - Medical Device Software Life Cycle Processes 15
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 5
P IEC 62304 - evaluation of integration and system testing IEC 62304 - Medical Device Software Life Cycle Processes 4
P Risk acceptability alignment between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 6
D Required Checklist Showing Compliance to IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 11
P Proposed revision of IEC 62304 - 2019 IEC 62304 - Medical Device Software Life Cycle Processes 6
S Relationship between IEC 62304 problem resolution and ISO 13485 IEC 62304 - Medical Device Software Life Cycle Processes 8
P IEC 62304:2006 A1:2015 - Software from the early 1990s IEC 62304 - Medical Device Software Life Cycle Processes 4
B IEC 62304:2015 vs IEC 62304:2006 + AMD1 IEC 62304 - Medical Device Software Life Cycle Processes 4
F IEC 62304 - Segregation and communication between software items IEC 62304 - Medical Device Software Life Cycle Processes 1
B Class IIB Device - IEC 62304 Software Classification IEC 62304 - Medical Device Software Life Cycle Processes 13
B IEC 62304 - Update Checklist IEC 62304 - Medical Device Software Life Cycle Processes 2
L Connection between IEC 62304 and Chapter 14 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 62304 - Develop an Architecture for the Interfaces of Software Items IEC 62304 - Medical Device Software Life Cycle Processes 8
S Does IEC 62304 require documenting unresolved anomalies for all safety classes? IEC 62304 - Medical Device Software Life Cycle Processes 4
T I need to make test reports according IEC 62304 & IEC 62366 IEC 62366 - Medical Device Usability Engineering 2
D Changing software classification via software - IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
D Software as risk control - Confused on one aspect of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 20
K Trying to figure out what satisfies a few aspects of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 2
Y IEC 62304 Section 4.3(a) - 100% probability of failure IEC 62304 - Medical Device Software Life Cycle Processes 3
Y Application of IEC/EN 62304 at an advanced stage of software development IEC 62304 - Medical Device Software Life Cycle Processes 4
T Is there any requirement to be compliant with IEC 62304 while implementing ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 5
L Documentation Planning - IEC 62304 Clause 5.1.8 IEC 62304 - Medical Device Software Life Cycle Processes 2
W CPU BIST IEC 62304 - Embedded code has CPU instruction tests IEC 62304 - Medical Device Software Life Cycle Processes 2
K IEC 62304 Amd 1 2015 - Figure 3 – Assigning Software Safety Classification IEC 62304 - Medical Device Software Life Cycle Processes 11
K Risk Reduction by Risk Control: IEC:62304-Class C ISO 14971 - Medical Device Risk Management 15
C Per IEC 62304, are DHF documents Configuration Items? IEC 62304 - Medical Device Software Life Cycle Processes 5
P IEC 62304 AMD1:2015: What's new vs.the 2006 Edition? IEC 62304 - Medical Device Software Life Cycle Processes 4
F FDA PMK 510(k) - IEC 62304 Software Components Segregation Other US Medical Device Regulations 3
M IEC 62304 Applicability - GUI Control Software IEC 62304 - Medical Device Software Life Cycle Processes 3
B Our NB says that IEC 62304 is an ISO 14971 Requirement ISO 14971 - Medical Device Risk Management 1
B Clarification on interpretation of some EN ISO 14971:2012 & IEC 62304:2006 req's ISO 14971 - Medical Device Risk Management 46
D A desperate call for help - IEC 62304 software IEC 62304 - Medical Device Software Life Cycle Processes 5
B IEC 62304:2006/AMD1:2015 Changes for Class A Software IEC 62304 - Medical Device Software Life Cycle Processes 3
K IEC 62304 - Compliance steps IEC 62304 - Medical Device Software Life Cycle Processes 2