Hello,

I am looking for suggestions and recommendations on how to address the following requirements of IEC 62304.

IEC 62304 Cl 5.3.6 - Verify SW Architecture:

5.3.6.a The MANUFACTURER has verified and documented that the ARCHITECTURE of the software implements SYSTEM and software requirements including those relating to RISK CONTROL.

5.3.6.b The MANUFACTURER has verified and documented that the SOFTWARE ARCHITECTURE is able to support interfaces between SOFTWARE ITEMS and between SOFTWARE ITEMS and hardware.

5.3.6.c The MANUFACTURER has verified and documented that the MEDICAL DEVICE ARCHITECTURE supports proper operation of any SOUP items.


IEC 62304 Annex B.5.3 states : Verification of the architecture is generally done by technical evaluation.

What activities in a development life cycle constitute a technical evaluation to verify architecture?
 
Elsmar Forum Sponsor
#3
How did you verify your SW architecture if it wasn't by a technical evaluation?
I'm asking because you've done your verification and I think you're probably reading something into this, that isn't there.
 

glork98

Involved In Discussions
#4
The verification activities should be described in the SOPs, Work Instructions and in the Software Development Plan, first of all.

What class is your software?

To verify the software architecture you would use a variety of review and test approaches as appropriate for your code and company. I'd expect that this would include a multi-pronged approach:

* Create SW architecture document and review it
* Create SW detailed design documents and review those
* Review code against the SW architecture and designs
* Write system, unit and integration tests
* Trace the above tests to the Software Requirements and the SW Req to the System requirements

And don't forget the risk analysis of the architecture, design and code. Make sure the mitigations (tests, code additions, reviews, more...) are followed through on.
 
R

raqark

#5
Thanks everyone.

Its good to get reassurance that the verification, design review activities, traceability, etc all provide evidence of architecture verification. I might have been reading too much into it. I understand it better now. :thanks:
 
Thread starter Similar threads Forum Replies Date
Y IEC 62304 Section 4.3(a) - 100% probability of failure IEC 62304 - Medical Device Software Life Cycle Processes 3
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 1
G Adopting old product - compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 9
A IEC 62304 safety classification, External Controls and off-label use related risks IEC 62304 - Medical Device Software Life Cycle Processes 5
S IEC 62304 software costs and time Medical Device and FDA Regulations and Standards News 3
S IEC 62304 - Software verification cost IEC 62304 - Medical Device Software Life Cycle Processes 3
Sravan Manchikanti Software Risk Management & probability of occurrence as per IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
M IEC 62304 Software changes - Minor labeling changes on the GUI IEC 62304 - Medical Device Software Life Cycle Processes 3
K IEC 62304 - Testing Independance IEC 62304 - Medical Device Software Life Cycle Processes 5
K IEC 62304 - Functional and performance requirements for SOUP items IEC 62304 - Medical Device Software Life Cycle Processes 2
K IEC 62304 compliance - Code reviews as part of verification strategy IEC 62304 - Medical Device Software Life Cycle Processes 5
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
M IEC 62304 Class A Project IEC 62304 - Medical Device Software Life Cycle Processes 15
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 5
P SOUP anomaly evaluation for MMA (Mobile Medical Application) IEC 62304 clause 7.1.3 IEC 62304 - Medical Device Software Life Cycle Processes 6
P IEC 62304 - evaluation of integration and system testing IEC 62304 - Medical Device Software Life Cycle Processes 4
P Risk acceptability alignment between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 6
D Required Checklist Showing Compliance to IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 11
P Proposed revision of IEC 62304 - 2019 IEC 62304 - Medical Device Software Life Cycle Processes 6
S Relationship between IEC 62304 problem resolution and ISO 13485 IEC 62304 - Medical Device Software Life Cycle Processes 8
P IEC 62304:2006 A1:2015 - Software from the early 1990s IEC 62304 - Medical Device Software Life Cycle Processes 4
B IEC 62304:2015 vs IEC 62304:2006 + AMD1 IEC 62304 - Medical Device Software Life Cycle Processes 4
F IEC 62304 - Segregation and communication between software items IEC 62304 - Medical Device Software Life Cycle Processes 1
B Class IIB Device - IEC 62304 Software Classification IEC 62304 - Medical Device Software Life Cycle Processes 13
B IEC 62304 - Update Checklist IEC 62304 - Medical Device Software Life Cycle Processes 2
L Connection between IEC 62304 and Chapter 14 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 62304 - Develop an Architecture for the Interfaces of Software Items IEC 62304 - Medical Device Software Life Cycle Processes 8
S Does IEC 62304 require documenting unresolved anomalies for all safety classes? IEC 62304 - Medical Device Software Life Cycle Processes 4
A SOP for software validation of software in medical device IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 5
T I need to make test reports according IEC 62304 & IEC 62366 IEC 62366 - Medical Device Usability Engineering 2
D Changing software classification via software - IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
D Software as risk control - Confused on one aspect of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 20
K Trying to figure out what satisfies a few aspects of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 2
Y Application of IEC/EN 62304 at an advanced stage of software development IEC 62304 - Medical Device Software Life Cycle Processes 4
T Is there any requirement to be compliant with IEC 62304 while implementing ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 5
L Documentation Planning - IEC 62304 Clause 5.1.8 IEC 62304 - Medical Device Software Life Cycle Processes 2
C Software for Medical Devices - Requirements Content for compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 1
W CPU BIST IEC 62304 - Embedded code has CPU instruction tests IEC 62304 - Medical Device Software Life Cycle Processes 2
K IEC 62304 Amd 1 2015 - Figure 3 – Assigning Software Safety Classification IEC 62304 - Medical Device Software Life Cycle Processes 11
K Risk Reduction by Risk Control: IEC:62304-Class C ISO 14971 - Medical Device Risk Management 15
C Per IEC 62304, are DHF documents Configuration Items? IEC 62304 - Medical Device Software Life Cycle Processes 8
P IEC 62304 AMD1:2015: What's new vs.the 2006 Edition? IEC 62304 - Medical Device Software Life Cycle Processes 4
F FDA PMK 510(k) - IEC 62304 Software Components Segregation Other US Medical Device Regulations 3
M IEC 62304 Applicability - GUI Control Software IEC 62304 - Medical Device Software Life Cycle Processes 3
B Our NB says that IEC 62304 is an ISO 14971 Requirement ISO 14971 - Medical Device Risk Management 1
B Clarification on interpretation of some EN ISO 14971:2012 & IEC 62304:2006 req's ISO 14971 - Medical Device Risk Management 46
H ISO 14971 vs. IEC 62304 vs. 98/79/EC vs. ISO 13485 (Software Medical Device) ISO 14971 - Medical Device Risk Management 1
D A desperate call for help - IEC 62304 software IEC 62304 - Medical Device Software Life Cycle Processes 5
B IEC 62304:2006/AMD1:2015 Changes for Class A Software IEC 62304 - Medical Device Software Life Cycle Processes 3

Similar threads

Top Bottom