IEC 62304 Software changes - Minor labeling changes on the GUI

Myodoll

Registered
I am a software developer at a small startup and we are currently working on developing the software as part of an integrated device. The device is used to collect and analyze patient data in clinical studies. We have had the first version of the software released, and are currently working on changing the spftware for a different indication of use, and the only changes we will be making are minor labeling changes on the GUI. However, we want to keep both versions of software available for future use (so not obsoleting either one of them). For compliance with IEC 62304, shall we start a new set of DHF? Will we need to perform the risk analysis again, or is there a way to justify that there will be no additional risk involved by this change? As we have a very tight timeline we would like to minimize the amount of documentation work to do. Does anyone have any suggestions? Thank you in advance for your help!!
 

JPHD1

Registered
Without knowing enough about your product and the actual changes, I can only offer brief guidance. This is not legal advice. Please consult an SME who can look into your documentation and processes.

Sounds like you'll have two separate products with different intended uses?

Your documentation should demonstrate you've done your due diligence to ensure each product is safe and effective to use by intended users, in the intended context, for the intended purpose. I assume you've done all that for the previous software. Even if you think the changes to the new software are minor, they need to be looked at in the same holistic way with safety and efficacy in mind. And then documented.

For example, are there any new, previously not considered users? Does the change in the GUI introduce new potential use errors? Or can the incorrect software version be accidentally released to unintended users?

What may be helpful, although not completely equivalent, is to look at what FDA considers important in the light of 510(k) submission for a software change (search for Deciding When to Submit a 510(k) for a Software Change to an Existing Device Guidance for Industry and Food and Drug Administration Staff).

In a nutshell, the analysis goes like this:
  1. Is the change made solely to strengthen cybersecurity and does not have any other impact on the software or device?
  2. Is the change made solely to return the system into specification of the most recently cleared device?
  3. What are the impacts of any changes to risks associated with use of the device and the impacts of any changes to the risk controls for the device?
    • Does the change introduce a new risk or modify an existing risk that could result in significant harm and that is not effectively mitigated in the most recently cleared device?
    • Does the change create or necessitate a new risk control measure or a modification of an existing risk control measure for a hazardous situation that could result in significant harm?
  4. Could the change significantly affect clinical functionality or performance specifications that are directly associated with the intended use of the device?
You can see there are quite a few important questions you'll need to be able to answer. How you document your thought process and how you ensure (in production and post-production) your products are safe and effective is up to you.
 

Tidge

Trusted Information Resource
If you are subject to FDA oversight, allow me to direct you towards a somewhat recent 2019 FDA guidance: Policy for Device Software Functions and Mobile Medical Applications

Without knowing any details beyond what was written, you probably can expand the existing DHF to cover multiple products as a 'family' of devices.
If you do so, you will need to make sure that the elements in a single DHF support each device appropriately (i.e. requirements, specification, V&V, risk analysis). IF certain elements (e.g. risk analysis) get too confusing to keep in a single document, it would be appropriate to disentangle documents as necessary. This could be discrete sections of documents in the DHF, or it could be new documents in a single DHF.
 

yodon

Leader
Super Moderator
shall we start a new set of DHF?

Regarding this question, each product must have a complete DHF, but it can well be a 'virtual folder.' We have clients that have multiple software products all based on, essentially, a core library. There are unique aspects to each product but to compile the complete DHF for any product, the shared information is pulled from the library material. It's just a matter of organization.
 
Thread starter Similar threads Forum Replies Date
T Class B IEC 62304 - 5.4 Software Detailed Design IEC 62304 - Medical Device Software Life Cycle Processes 4
MaHoDie Is it possible to assign medical software to security class A (according to IEC 62304)? EU Medical Device Regulations 8
Y Software to Manage IEC 62304 Traceability Requirement IEC 62304 - Medical Device Software Life Cycle Processes 3
shimonv Working with a software developer who is not setup for IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 9
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 4
S IEC 62304 software costs and time Medical Device and FDA Regulations and Standards News 3
S IEC 62304 - Software verification cost IEC 62304 - Medical Device Software Life Cycle Processes 3
Sravan Manchikanti Software Risk Management & probability of occurrence as per IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 9
P IEC 62304:2006 A1:2015 - Software from the early 1990s IEC 62304 - Medical Device Software Life Cycle Processes 4
F IEC 62304 - Segregation and communication between software items IEC 62304 - Medical Device Software Life Cycle Processes 1
B Class IIB Device - IEC 62304 Software Classification IEC 62304 - Medical Device Software Life Cycle Processes 13
M IEC 62304 - Develop an Architecture for the Interfaces of Software Items IEC 62304 - Medical Device Software Life Cycle Processes 8
A SOP for software validation of software in medical device IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 5
D Changing software classification via software - IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
D Software as risk control - Confused on one aspect of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 20
Y Application of IEC/EN 62304 at an advanced stage of software development IEC 62304 - Medical Device Software Life Cycle Processes 4
C Software for Medical Devices - Requirements Content for compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 1
K IEC 62304 Amd 1 2015 - Figure 3 – Assigning Software Safety Classification IEC 62304 - Medical Device Software Life Cycle Processes 11
F FDA PMK 510(k) - IEC 62304 Software Components Segregation Other US Medical Device Regulations 3
M IEC 62304 Applicability - GUI Control Software IEC 62304 - Medical Device Software Life Cycle Processes 3
H ISO 14971 vs. IEC 62304 vs. 98/79/EC vs. ISO 13485 (Software Medical Device) ISO 14971 - Medical Device Risk Management 1
D A desperate call for help - IEC 62304 software IEC 62304 - Medical Device Software Life Cycle Processes 5
B IEC 62304:2006/AMD1:2015 Changes for Class A Software IEC 62304 - Medical Device Software Life Cycle Processes 3
K ISO 14971 and IEC 62304 - Medical Device Software House ISO 14971 - Medical Device Risk Management 9
S Software Test Report including IEC 62304 classification IEC 62304 - Medical Device Software Life Cycle Processes 4
A Requirements for Software to be CE Mark Compliant - IEC 62304 Medical Device Software IEC 62304 - Medical Device Software Life Cycle Processes 14
E IEC 62304 5.2.2 f - Installation and acceptance of device software IEC 62304 - Medical Device Software Life Cycle Processes 1
R IEC 62304 - 5.2.9 Verify Software Requirement IEC 62304 - Medical Device Software Life Cycle Processes 5
R IEC 62304 Section 5.3.6 Verify Software Architecture IEC 62304 - Medical Device Software Life Cycle Processes 4
T IEC 62304 required for CE Marking of Software? IEC 62304 - Medical Device Software Life Cycle Processes 5
S IEC 62304 Software Process Diagram IEC 62304 - Medical Device Software Life Cycle Processes 1
E Where to Buy IEC 62304 Software Life Cycle IEC 62304 - Medical Device Software Life Cycle Processes 7
C IEC 62304 conformity for 510k submission legacy product including software IEC 62304 - Medical Device Software Life Cycle Processes 2
R Software System Inputs and Outputs - IEC 62304 5.2.2 - Software Requirements Content IEC 62304 - Medical Device Software Life Cycle Processes 6
I IEC 62304:2006 Definitions - Software System, a Software Element and Software Unit IEC 62304 - Medical Device Software Life Cycle Processes 13
R How do we show our Software Develoment Life Cycle is similar to IEC 62304 Other US Medical Device Regulations 3
S IEC 62304 Clause 6. Medical Device Software Maintenance Plan IEC 62304 - Medical Device Software Life Cycle Processes 7
M IEC 62304 vs. General Principles of Software Validation IEC 62304 - Medical Device Software Life Cycle Processes 8
S One function, many devices? IEC 62304 and Medical Device Imaging Software IEC 62304 - Medical Device Software Life Cycle Processes 9
T IEC 62304 & FDA: Software Development Methodologies for Class II- DICOM device IEC 62304 - Medical Device Software Life Cycle Processes 8
sagai Process Model for IEC 62304 - Medical Device Software Life Cycle IEC 62304 - Medical Device Software Life Cycle Processes 4
T IEC 62304:2006: Medical device software SDLC- CE Vs. Fda 510(k) submission IEC 62304 - Medical Device Software Life Cycle Processes 16
B Tabular Traceability by Software Item ID - IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
A IEC 62304 Software Development Plan IEC 62304 - Medical Device Software Life Cycle Processes 28
W IEC 62304:2006 Medical device software - Software life cycle processes - Issued IEC 62304 - Medical Device Software Life Cycle Processes 10
L EN 62304:2006/AC:2008 vs IEC 62304:2006/Amd 1:2015 IEC 62304 - Medical Device Software Life Cycle Processes 1
T FDA level of concern vs IEC 62304 safety classification - 510(k) exempt device Other US Medical Device Regulations 13
B Should I buy IEC 62304:2006, IEC 62304:2006/AMD 1:2015 or both? EU Medical Device Regulations 1
M How to apply IEC 62304 Problem Resolution Process for bugs discovered prior to release i.e. during initial development? IEC 62304 - Medical Device Software Life Cycle Processes 2
T IEC 62304 : Risk control for SaMD IEC 62304 - Medical Device Software Life Cycle Processes 8

Similar threads

Top Bottom