Hello All,

We have a Class C medical device. I was reviewing some of the testing documentation and find that some of the software system tests have been performed by the developers themselves.

The test cases were reviewed and approved by different people but some of them were still performed by the developers. Would this be allowed?
 
Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
While it's probably not the best practice (due to a variety of reasons), there's nothing in the standards that would explicitly disallow it. Testing to requirements should generate objective evidence that the requirement was met. If, in fact, the testing did generate objective evidence of compliance, you should be in a pretty defensible position. As noted, there may be some landmines with this approach. If the requirement is open some to interpretation, the developers would, of course, be biased by how they implemented. Similarly, developers generally think "happy path" so the testing may not be as rigorous as you'd want, especially for a Class C device. So be careful.
 

Tidge

Quite Involved in Discussions
#3
If I were trying to detonate some of the landmines suggested by @yodon I would ask:
  • Were the test criteria established prior to the developers doing testing?
  • Did the testing occur on a controlled release of the software system?
If either of these questions are answered "no", then it is very likely that the development model of 62304 was not followed.
 

KrishQA

Starting to get Involved
#4
@yodon
I have asked the team to retest those specific test cases again and get them approved.

There is evidence of performing the test, and the results have been captured, however there are no screenshots, logs or other evidence that can be looked at and verified.

@Tidge
The test criteria was established prior to testing and it has been done in a controlled release. For this retesting, we are creating a change request to make the changes to documentation as well, since the release notes are already signed.
 

Tidge

Quite Involved in Discussions
#5
@KrishQA It sounds like you have the bases covered. Some companies may not have the resources to allow for a completely different group of people to do the system testing, even if it is a best practice.

If you have pre-approved protocols, with instructions and pre-determined expected results, you may not need screenshots. My own bias is to rely on screenshots only for those results where it would otherwise be too much trouble to record the information. Examples include: GUI testing, or if there are multiple items simultaneously displayed. Keep in mind that if screenshots and/or logs are used as evidence of system testing that those need to be preserved.

My experience has been that there is a certain generation of folks that are rather obsessive about screenshots. Too many screenshots can clutter up test results; screenshots are not impossible to manipulate so it isn't as if they are a 'better' form of proof.
 

yodon

Staff member
Super Moderator
#6
There is evidence of performing the test, and the results have been captured, however there are no screenshots, logs or other evidence that can be looked at and verified.
My approach is that the tests should gather sufficient evidence that an independent reviewer (with sufficient knowledge) should be able to draw the same conclusion that the tester made.

Too many screenshots can clutter up test results;
AMEN! I once had a client that insisted on a screenshot for every screen transition. At the end, we had stacks of paper over a foot high and 5 feet long. (And, honestly, most of the screenshots did nothing to support the evidence that a requirement was met.) Be judicious!
 
Thread starter Similar threads Forum Replies Date
P IEC 62304 - evaluation of integration and system testing IEC 62304 - Medical Device Software Life Cycle Processes 4
S IEC 62304 software costs and time Medical Device and FDA Regulations and Standards News 2
S IEC 62304 - Software verification cost IEC 62304 - Medical Device Software Life Cycle Processes 3
Sravan Manchikanti Software Risk Management & probability of occurrence as per IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
M IEC 62304 Software changes - Minor labeling changes on the GUI IEC 62304 - Medical Device Software Life Cycle Processes 3
K IEC 62304 - Functional and performance requirements for SOUP items IEC 62304 - Medical Device Software Life Cycle Processes 2
K IEC 62304 compliance - Code reviews as part of verification strategy IEC 62304 - Medical Device Software Life Cycle Processes 5
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
M IEC 62304 Class A Project IEC 62304 - Medical Device Software Life Cycle Processes 15
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 5
P Risk acceptability alignment between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 6
D Required Checklist Showing Compliance to IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 11
P Proposed revision of IEC 62304 - 2019 IEC 62304 - Medical Device Software Life Cycle Processes 6
S Relationship between IEC 62304 problem resolution and ISO 13485 IEC 62304 - Medical Device Software Life Cycle Processes 8
P IEC 62304:2006 A1:2015 - Software from the early 1990s IEC 62304 - Medical Device Software Life Cycle Processes 4
B IEC 62304:2015 vs IEC 62304:2006 + AMD1 IEC 62304 - Medical Device Software Life Cycle Processes 4
F IEC 62304 - Segregation and communication between software items IEC 62304 - Medical Device Software Life Cycle Processes 1
B Class IIB Device - IEC 62304 Software Classification IEC 62304 - Medical Device Software Life Cycle Processes 13
B IEC 62304 - Update Checklist IEC 62304 - Medical Device Software Life Cycle Processes 2
L Connection between IEC 62304 and Chapter 14 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 62304 - Develop an Architecture for the Interfaces of Software Items IEC 62304 - Medical Device Software Life Cycle Processes 8
S Does IEC 62304 require documenting unresolved anomalies for all safety classes? IEC 62304 - Medical Device Software Life Cycle Processes 4
A SOP for software validation of software in medical device IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 5
T I need to make test reports according IEC 62304 & IEC 62366 IEC 62366 - Medical Device Usability Engineering 2
D Changing software classification via software - IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
D Software as risk control - Confused on one aspect of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 20
K Trying to figure out what satisfies a few aspects of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 2
Y IEC 62304 Section 4.3(a) - 100% probability of failure IEC 62304 - Medical Device Software Life Cycle Processes 3
Y Application of IEC/EN 62304 at an advanced stage of software development IEC 62304 - Medical Device Software Life Cycle Processes 4
T Is there any requirement to be compliant with IEC 62304 while implementing ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 5
L Documentation Planning - IEC 62304 Clause 5.1.8 IEC 62304 - Medical Device Software Life Cycle Processes 2
C Software for Medical Devices - Requirements Content for compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 1
W CPU BIST IEC 62304 - Embedded code has CPU instruction tests IEC 62304 - Medical Device Software Life Cycle Processes 2
K IEC 62304 Amd 1 2015 - Figure 3 – Assigning Software Safety Classification IEC 62304 - Medical Device Software Life Cycle Processes 11
K Risk Reduction by Risk Control: IEC:62304-Class C ISO 14971 - Medical Device Risk Management 15
C Per IEC 62304, are DHF documents Configuration Items? IEC 62304 - Medical Device Software Life Cycle Processes 5
P IEC 62304 AMD1:2015: What's new vs.the 2006 Edition? IEC 62304 - Medical Device Software Life Cycle Processes 4
F FDA PMK 510(k) - IEC 62304 Software Components Segregation Other US Medical Device Regulations 3
M IEC 62304 Applicability - GUI Control Software IEC 62304 - Medical Device Software Life Cycle Processes 3
B Our NB says that IEC 62304 is an ISO 14971 Requirement ISO 14971 - Medical Device Risk Management 1
B Clarification on interpretation of some EN ISO 14971:2012 & IEC 62304:2006 req's ISO 14971 - Medical Device Risk Management 46
H ISO 14971 vs. IEC 62304 vs. 98/79/EC vs. ISO 13485 (Software Medical Device) ISO 14971 - Medical Device Risk Management 1
D A desperate call for help - IEC 62304 software IEC 62304 - Medical Device Software Life Cycle Processes 5
B IEC 62304:2006/AMD1:2015 Changes for Class A Software IEC 62304 - Medical Device Software Life Cycle Processes 3
M IEC 62304, ISO 14971 and FDA Medical Device SW Guidance 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
K IEC 62304 - Compliance steps IEC 62304 - Medical Device Software Life Cycle Processes 5
K ISO 14971 and IEC 62304 - Medical Device Software House ISO 14971 - Medical Device Risk Management 9
S Software Test Report including IEC 62304 classification IEC 62304 - Medical Device Software Life Cycle Processes 4
A Mapping of IEC 62304 artefacts (SRS, SAD, etc) to the 820.30 phases IEC 62304 - Medical Device Software Life Cycle Processes 5

Similar threads

Top Bottom