IEC 62366 - Ethical approval
Julie_
Hi all,
Do you always need an approval from an ethical commission before you can do usability testing?
Even if there is nothing ethical about the scenarios that are part of it?
Tobias_HF
Hi all,
Yes, if you plan to submit it to FDA, later on.
IRB (ethical commission) is required when human subjects are involved. This is regardless if they are patients / lay persons or practitioners.
A usability study can cause harm, and the IRB will review your protocol / material if you did all necessary precautions to prevent this harm.
If you do a study in Europe, the IRB oversight is less strict.
Marcelo
IRB (ethical commission) is required when human subjects are involved..
Do you have any source that clearly says that an IRB approval is required for HF studies for all FDA submissions? I know that for some studies it will be required, but not all, and the only information I can find is exactly stating this.
Tobias_HF
No, I don`t have this. Unfortunately FDA is pretty vague about this in their official statements...
But, the Association of American Psychologists is clearer on this. They traditionally dealt with HF studies, and I assume this is why FDA is following their approach (Frequently Asked Questions about Institutional Review Boards). As I understand, if you are a registered psychologist and do a study that involves humans, you have to get IRB review. No matter what the subject of the study is.
I think somewhere in the CFR 820 it states that you should follow local and federal laws & regulations, and of course HIIPA. The IRB review also makes sure that you do that, so this would be your "certificate" that you are not breaking any federal laws, when doing a HF study (which might happen way easier than one might think...).
Also, one point to consider: if you have a good protocol, IRB review is usually fast, and not that expensive. So I`d do it anyway, as a "quality check" for a subject matter, that neither me nor my colleagues are experts in...
Marcelo
You are right that you need to perform any test/study under local law/regulations, but Associations usually do not prescribe regulations or anything unless a regulatory body says so So I'm not sure anything from Association of American Psychologists is mandatory (on a regulatory level). Also, the link you mentioned is only a FAQ on how an IRB works (so I'm not sure what you meant by " why FDA is following their approach").
This transcript of a FDA meeting about the last hum an factors guidance mentions:
The IRB question again is it depends on the studies and the study design. So that will be -- again that's a case-by-case decision as to whether or not your human factors study should be going through an IRB or not. So I don't have a direct answer to that. Again, I would suggest working with your reviewing division to identify if there is a need for that for your study type.
This guidance Human Factors Studies and Related Clinical Study Considerations in Combination Product Design and Development mentions that usability studies for IDEs would require IRB, because they would be under the same light as a clinical investigation.
Even the Usability Testing of Medical Devices book has a chapter on this, but also mentions that it[s not clear a requirement 9although it suggests that a best practice would be to go after it).
adir88 Amendment 1 of IEC 62366-1 - Details yet? IEC 62366 - Medical Device Usability Engineering 2
K Primary Operating Functions (POF) in IEC 62366-1:2015. IEC 62366 - Medical Device Usability Engineering 7
V Applicability of IEC 62366-1 usability to resorbable bone substitutes IEC 62366 - Medical Device Usability Engineering 9
pashah Looking for Usability File Template acc. IEC 62366-1 and IEC 60601-1-6 IEC 62366 - Medical Device Usability Engineering 4
C Usability IEC 62366-1:2015 and MDR 2017/745 - Risk based approach IEC 62366 - Medical Device Usability Engineering 1
E Usability Process - Looking for a template (IEC 62366) IEC 62366 - Medical Device Usability Engineering 1
J IEC 62366 - Summative Evaluation IEC 62366 - Medical Device Usability Engineering 3
T I need to make test reports according IEC 62304 & IEC 62366 IEC 62366 - Medical Device Usability Engineering 2
C IEC 60601-1-6 2010 maps to IEC 62366:2007 - What about 62366-1:2015? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K Usability of ME Equipment - IEC-62366-1:2015 IEC 62366 - Medical Device Usability Engineering 3
T IEC 62366 Test Report Form IEC 62366 - Medical Device Usability Engineering 20
M IEC 62366-1:2015 Usability Verification / Validation Plan example wanted IEC 62366 - Medical Device Usability Engineering 20
T IEC 62366:2015 Annex C - changed parts from 5.1-5.8 --> but what about 5.9? IEC 62366 - Medical Device Usability Engineering 3
S IEC 62366 - Usability Test Protocol format Imported Legacy Blogs 0
M IEC 62366-1 published - March 2015 IEC 62366 - Medical Device Usability Engineering 7
S IEC 62366 vs. FDA Human Factors Requirements Human Factors and Ergonomics in Engineering 2
Z Final Validation Unit Requirements - Annex D of IEC 62366 IEC 62366 - Medical Device Usability Engineering 6
Z IEC 62366 and Medical Device Usability Validation IEC 62366 - Medical Device Usability Engineering 10
C Application of IEC 62366 - Usability of Non-Active Medical Devices IEC 62366 - Medical Device Usability Engineering 15
T IEC 62366 included in 3rd ed IEC 60601? IEC 62366 - Medical Device Usability Engineering 1
M IEC 62366:2007 - Is FDA Usability Engineering Standard a Requirement in Canada? Canada Medical Device Regulations 5
S IEC 62366 - Embedded Software in separate Usability Engineering File? IEC 62366 - Medical Device Usability Engineering 1
J Where to buy standards such as IEC 62366:2007 Other Medical Device Related Standards 2
S IEC 62366 for Stand-Alone Medical Device Software and Usability Engineering IEC 62366 - Medical Device Usability Engineering 16
D Training on IEC 62366 (Medical Device Usability Engineering) for CE Mark? IEC 62366 - Medical Device Usability Engineering 5
M What do you think the revision of IEC 62366 should bring? IEC 62366 - Medical Device Usability Engineering 16
A IEC 60601-1-6 last release or IEC 62366? IEC 62366 - Medical Device Usability Engineering 17
K How to apply IEC 62366 and what shall be changed in establishing a Technical File ISO 13485:2016 - Medical Device Quality Management Systems 2
D IEC EN 62366 Usability Retroactive Compliance Other Medical Device Related Standards 1
R ISO IEC EN 62366 Usability Engineering Essential Requirements IEC 62366 - Medical Device Usability Engineering 8
P IEC 62366 vs. IEC 60601-1-6 - Has IEC 62366 now replaced IEC 60601-1-6? IEC 62366 - Medical Device Usability Engineering 19
Le Chiffre Online training available for ISO/IEC 17021-1: Requirements for bodies providing audit and certification of management systems Training - Internal, External, Online and Distance Learning 3
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 1
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T Relationship between ISO 9001 and ISO – IEC BS EN 870079- 34 2020 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
T Help with BS EN ISO - IEC 80079-34 2020 (Explosive atmospheres QMS) Other ISO and International Standards and European Regulations 0
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E IEC 62353 normalization IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
G Adopting old product - compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 9
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 61010-1 requirements for motors Other Medical Device Related Standards 0
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9