A Quick Bump!

Can someone help?

If your questions were answered here, please consider helping others.

Thank you very much!
 

Mark Meer

Trusted Information Resource
#3
As far as using IEC 62366 as a consensus standard, you can lookup the standard here, and it will tell you which parts are not recognized by the FDA.

If you lookup IEC 62366, you will get the following exceptions:
  • Clause 5.5 (USABILITY SPECIFICATION) Note 1 is not recognized
  • Table D.5 (Examples of objective USABILITY goals) is not recognized
  • Table D.6 (Examples of subjective USABILITY goals) is not recognized
  • Annex G (USABILITY goals: Illustrative example for a home parenteral infusion pump) is not recognized.
 
Thread starter Similar threads Forum Replies Date
M IEC 62366:2007 - Is FDA Usability Engineering Standard a Requirement in Canada? Canada Medical Device Regulations 5
adir88 Amendment 1 of IEC 62366-1 - Details yet? IEC 62366 - Medical Device Usability Engineering 2
K Primary Operating Functions (POF) in IEC 62366-1:2015. IEC 62366 - Medical Device Usability Engineering 7
V Applicability of IEC 62366-1 usability to resorbable bone substitutes IEC 62366 - Medical Device Usability Engineering 9
pashah Looking for Usability File Template acc. IEC 62366-1 and IEC 60601-1-6 IEC 62366 - Medical Device Usability Engineering 4
C Usability IEC 62366-1:2015 and MDR 2017/745 - Risk based approach IEC 62366 - Medical Device Usability Engineering 1
E Usability Process - Looking for a template (IEC 62366) IEC 62366 - Medical Device Usability Engineering 1
J IEC 62366 - Ethical approval IEC 62366 - Medical Device Usability Engineering 5
J IEC 62366 - Summative Evaluation IEC 62366 - Medical Device Usability Engineering 3
T I need to make test reports according IEC 62304 & IEC 62366 IEC 62366 - Medical Device Usability Engineering 2
C IEC 60601-1-6 2010 maps to IEC 62366:2007 - What about 62366-1:2015? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K Usability of ME Equipment - IEC-62366-1:2015 IEC 62366 - Medical Device Usability Engineering 3
T IEC 62366 Test Report Form IEC 62366 - Medical Device Usability Engineering 20
M IEC 62366-1:2015 Usability Verification / Validation Plan example wanted IEC 62366 - Medical Device Usability Engineering 20
T IEC 62366:2015 Annex C - changed parts from 5.1-5.8 --> but what about 5.9? IEC 62366 - Medical Device Usability Engineering 3
S IEC 62366 - Usability Test Protocol format Imported Legacy Blogs 0
M IEC 62366-1 published - March 2015 IEC 62366 - Medical Device Usability Engineering 7
Z Final Validation Unit Requirements - Annex D of IEC 62366 IEC 62366 - Medical Device Usability Engineering 6
Z IEC 62366 and Medical Device Usability Validation IEC 62366 - Medical Device Usability Engineering 10
C Application of IEC 62366 - Usability of Non-Active Medical Devices IEC 62366 - Medical Device Usability Engineering 15
T IEC 62366 included in 3rd ed IEC 60601? IEC 62366 - Medical Device Usability Engineering 1
S IEC 62366 - Embedded Software in separate Usability Engineering File? IEC 62366 - Medical Device Usability Engineering 1
J Where to buy standards such as IEC 62366:2007 Other Medical Device Related Standards 2
S IEC 62366 for Stand-Alone Medical Device Software and Usability Engineering IEC 62366 - Medical Device Usability Engineering 16
D Training on IEC 62366 (Medical Device Usability Engineering) for CE Mark? IEC 62366 - Medical Device Usability Engineering 5
M What do you think the revision of IEC 62366 should bring? IEC 62366 - Medical Device Usability Engineering 16
A IEC 60601-1-6 last release or IEC 62366? IEC 62366 - Medical Device Usability Engineering 17
K How to apply IEC 62366 and what shall be changed in establishing a Technical File ISO 13485:2016 - Medical Device Quality Management Systems 2
D IEC EN 62366 Usability Retroactive Compliance Other Medical Device Related Standards 1
R ISO IEC EN 62366 Usability Engineering Essential Requirements IEC 62366 - Medical Device Usability Engineering 8
P IEC 62366 vs. IEC 60601-1-6 - Has IEC 62366 now replaced IEC 60601-1-6? IEC 62366 - Medical Device Usability Engineering 19
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C ISO/IEC 17021-1 clause 7.1.2 - Determination of competence criteria Document Control Systems, Procedures, Forms and Templates 2
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C ISO/ IEC 17021 Resource requirement (need help) Document Control Systems, Procedures, Forms and Templates 5
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 0
S IEC 62304 software costs and time Medical Device and FDA Regulations and Standards News 3
S IEC 62304 - Software verification cost IEC 62304 - Medical Device Software Life Cycle Processes 3
Sravan Manchikanti Software Risk Management & probability of occurrence as per IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
M IEC 62304 Software changes - Minor labeling changes on the GUI IEC 62304 - Medical Device Software Life Cycle Processes 3
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K IEC 62304 - Testing Independance IEC 62304 - Medical Device Software Life Cycle Processes 5
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Exclusions or justification for non-applicability of IEC standards Reliability Analysis - Predictions, Testing and Standards 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
T ISO/IEC 17065 certification scheme Help Other ISO and International Standards and European Regulations 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
K IEC 62304 - Functional and performance requirements for SOUP items IEC 62304 - Medical Device Software Life Cycle Processes 2

Similar threads

Top Bottom