IEC 62368-1 / Electrical Energy Source vs. Power Source?

#1
Hi,

I'm coming from very little safety background, and am trying to wrap my head around the practical differences in IEC 62368-1 of a Power Source vs. a Electrical Energy Source.

I'm most confused in a case where we have a device that is powered from AC / DC adapter (5V, 1A), charges from a USB input, and has an internal LiPo battery (3.7V, 2A).

In some cases, I've seen the USB input & battery classified as an ES1 and PS1 source. How could this be correct?

My uninitiated reading of the standards would be that the AC Adapter / DC output would be the electrical energy source. I would think the LiPo battery would then be the Power Source?

I've looked a number of EN 62368 test reports, and I've seen a really wide divergence for these types of embedded devices in what exact component is classified as the ES or PS.

Does anyone have some high-level guidance on where the ES begins and ends, and then where the PS begins and ends?
 
Elsmar Forum Sponsor
#2
Hello.

IEC 62368-1 defines energy sources depening on the nature of the hazard or type of injury that they may cuause to the user.

So, it is stated various types:

Electrical source is related to risk of electrical shock
Power source is related to risk of fire.


Other energy sources could be:
-Mechanical energy sources
-Radiation type source
-Chemical energy source
-Thermal energy source

Regarding to your specific question:

Electrical sources are classified taking in account only avalable voltage and current of the circuit under consideration.

Power sources are classified taking in account the total available power that the relevant circuit may supply.

They are independent classifications. For example: Your battery could be at same time ES1 ( more or less similar to SELV) and PS1 if the available power that it can supply is less than certain limit (simplifyng a lot, the limit would be 15W for PS1), but could be ES1 and PS2/PS3 if the total amount of available power that can supply is higher).

Similar procedure can be applied to the AC/DC adapter or any other energy source.

Best regards.
 

CharlieUK

Quite Involved in Discussions
#3
The battery output can be PS1 due to protection circuitry, but there are additional requirements in Annex - for example a Lithium Ion battery must have a Fire Enclosure due to the electrolyte - though this Fire Enclosure might already be provided by the battery case
 
Thread starter Similar threads Forum Replies Date
K IEC 62304 compliance - Code reviews as part of verification strategy IEC 62304 - Medical Device Software Life Cycle Processes 0
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 62471 vs IEC/TR 62778 CE Marking (Conformité Européene) / CB Scheme 0
B IEC 61010, Touch currents related to burns at higher frequencies (cl. 6.3, A.3) Other Medical Device Related Standards 0
M IEC 61000-4-2: Requirements for coating declared as insulating IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 7
B IEC 61010-1 - Voltage / Dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M When is EN 62133-1 (new version i.e. IEC 62133:2017) mandatory in EU Other ISO and International Standards and European Regulations 0
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
adir88 Amendment 1 of IEC 62366-1 - Details yet? IEC 62366 - Medical Device Usability Engineering 0
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
B Has anyone done an IEC 60601-1 gap analysis to IEC 60335? Medical Device and FDA Regulations and Standards News 4
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
K Primary Operating Functions (POF) in IEC 62366-1:2015. IEC 62366 - Medical Device Usability Engineering 5
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 62304 Class A Project IEC 62304 - Medical Device Software Life Cycle Processes 15
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 4
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
MDD_QNA QR Code Standard ISO/IEC 15417:2007 - Does anyone use it? Other Medical Device Related Standards 3
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Universal Power Cord with Adapters - IVD IEC 61010-2-101 CE Marking (Conformité Européene) / CB Scheme 6
V Applicability of IEC 62366-1 usability to resorbable bone substitutes IEC 62366 - Medical Device Usability Engineering 9
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 3
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
V IS/ISO/IEC 17025:2017 Clause 7, sub clause 7.11 Control of data and information management ISO 17025 related Discussions 1
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
A IEC 60601-1 and IEC 60601-2-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601-1-2 EMC Tables IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Bookmarking my security protected IEC 60601-1 .pdf file IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
P IEC 62304 - evaluation of integration and system testing IEC 62304 - Medical Device Software Life Cycle Processes 4
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P Risk acceptability alignment between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 6
Similar threads


















































Top Bottom