Michael.K.
Registered
Hi,
I am reviewing a specific standard for dental equipment and encountered a requirement for the primary circuit to withstand a 4kV impulse, while ensuring the voltage on the secondary does not exceed 1kV. This test, as defined in clause 201, appears similar to the surge test in IEC 60601, which typically involves applying 1.5kV between line (L) and neutral (N).
Could you please clarify the following points:
Clause 201 (IEC 80601-2-60)
The transient impulse voltage of max. 1 kV shall be checked by testing as follows: Apply a
test voltage of 4 kV of 1,2/50 μs waveform to the primary circuit (mains). Verify that a limit of 1 kV
is not exceeded in the secondary circuit. The waveform has to be in accordance with
IEC 61180-1. The generator must comply with IEC 61180-2 (internal resistance of 2 Ω).
I am reviewing a specific standard for dental equipment and encountered a requirement for the primary circuit to withstand a 4kV impulse, while ensuring the voltage on the secondary does not exceed 1kV. This test, as defined in clause 201, appears similar to the surge test in IEC 60601, which typically involves applying 1.5kV between line (L) and neutral (N).
Could you please clarify the following points:
- If the power supply is certified to IEC 60601-1, would it comply with clause 201?
- Are there specific procedures or considerations I should be aware of when conducting the 4kV impulse test on dental equipment?
Clause 201 (IEC 80601-2-60)
The transient impulse voltage of max. 1 kV shall be checked by testing as follows: Apply a
test voltage of 4 kV of 1,2/50 μs waveform to the primary circuit (mains). Verify that a limit of 1 kV
is not exceeded in the secondary circuit. The waveform has to be in accordance with
IEC 61180-1. The generator must comply with IEC 61180-2 (internal resistance of 2 Ω).