Hi there,

I recently started looking into electrical safety and the IEC 60601-1:2005 standard and managed to come across this thread.

Previously, my predecessor had an external testing body tested the same product in accordance to the IEC 60601-1:1988 standard. The test report has roughly the same main points as the checklist from MECA but with slightly more details for each main point.

My question is: Would completing this checklist be sufficient to claim the medical device is IEC60601 compliant? as opposed to going through an external testing body?

Thanks in advance!
 
M

MIREGMGR

#14
Would completing this checklist be sufficient to claim the medical device is IEC60601 compliant? as opposed to going through an external testing body?
What market and regulatory system are you aiming at, and what is the reason for compliance?

Some FDA classification codes define IEC 60601-1 as a relevant Recognized Consensus Standard. A 510(k) premarket notification for a product in one of those classes will be most easily accepted if it includes evidence of conformance. Third party testing is an easy path to that evidence, but depending on the class I believe in-house testing is viable as well. In-house testing will require much more documentation, though, since you'll have to establish your calibrations and other measurement equipment validations, your technique conformance to measurement method standards, your personnel competence, etc.

Another common reason for compliance in the US market is that a number of inspectional jurisdictions...primarily large cities...won't approve the installation of electrical equipment unless it's Listed, and the Listing must be obtained via determination of standard conformance by a major test lab.

We also have heard of hospitals that have been told by their insurors in regard to potential major imaging system purchases that only Listed systems would be eligible for inclusion in their insurance coverage. I've heard of another instance where end users felt that they had to buy Listed equipment in regard to their JCAHO compliance stance, though I don't fully understand that one.

If your goal is to be able to support a marketing claim, at least in the US market, I'd think that in-house testing, competently carried out with appropriate documentation of course, would be fully sufficient.
 

RSantos

Registered Visitor
#15
I would expect your registrar to want 60601-1 testing, especially the EMC testing, to be performed by an ISO 17025 accredited lab. Those labs are accredited to specific test methods (always check their scope of accreditation) plus the quality system requirements. It's a lot of work ($$) setting a 17025 system in place, plus continued competency must be proven by round-robin testing and witnessing of testing.
 
F

freewind

#16
Thank you both for your reply.

I am currently working towards gaining registration with TGA. I believe ultimately, the path will lead to registering with FDA. But till then, as a manufacturer in Australia, TGA comes first, and electrical safety compliance is a must.

I guess we will try to perform in-house testing for now and keep our fingers crossed.

By the way is anyone a consultant in Australia?
 

Stijloor

Staff member
Super Moderator
#17
Thank you both for your reply.

I am currently working towards gaining registration with TGA. I believe ultimately, the path will lead to registering with FDA. But till then, as a manufacturer in Australia, TGA comes first, and electrical safety compliance is a must.

I guess we will try to perform in-house testing for now and keep our fingers crossed.

By the way is anyone a consultant in Australia?
Freewind,

JaneB is. She is a very nice person!

Stijloor.
 
Last edited:
Thread starter Similar threads Forum Replies Date
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC/TR 80002-1:2009 vs. AAMI TIR32:2004 Other Medical Device Related Standards 3
E Test report to certify compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 2
B 3-phase supply, overvoltage catogory, line-to-neutral and working voltage (IEC 61010-1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B European Harmonized vs IEC standard IEC 62366 - Medical Device Usability Engineering 2
E Software maintenance Process Software maintenance Process to IEC 6204? IEC 62304 - Medical Device Software Life Cycle Processes 3
A ISO/DIS 15223-1:2020 - Country of manufacture label (IEC 60417 No. 6049) - Which national law requires this symbol? Other Medical Device Related Standards 0
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B Reference to IEC 60950-1 in Amendment 2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
Le Chiffre Online training available for ISO/IEC 17021-1: Requirements for bodies providing audit and certification of management systems Training - Internal, External, Online and Distance Learning 3
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 1
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T Relationship between ISO 9001 and ISO – IEC BS EN 870079- 34 2020 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
T Help with BS EN ISO - IEC 80079-34 2020 (Explosive atmospheres QMS) Other ISO and International Standards and European Regulations 0
E IEC 62353 normalization IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
G Adopting old product - compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 9
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
M IEC 61010-1 requirements for motors Other Medical Device Related Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
A IEC 62304 safety classification, External Controls and off-label use related risks IEC 62304 - Medical Device Software Life Cycle Processes 5
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C ISO/IEC 17021-1 clause 7.1.2 - Determination of competence criteria Document Control Systems, Procedures, Forms and Templates 2
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C ISO/ IEC 17021 Resource requirement (need help) Document Control Systems, Procedures, Forms and Templates 5
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
S IEC 62304 software costs and time Medical Device and FDA Regulations and Standards News 3
S IEC 62304 - Software verification cost IEC 62304 - Medical Device Software Life Cycle Processes 3
Sravan Manchikanti Software Risk Management & probability of occurrence as per IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
M IEC 62304 Software changes - Minor labeling changes on the GUI IEC 62304 - Medical Device Software Life Cycle Processes 3
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K IEC 62304 - Testing Independance IEC 62304 - Medical Device Software Life Cycle Processes 5
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Exclusions or justification for non-applicability of IEC standards Reliability Analysis - Predictions, Testing and Standards 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7