IEC EN 60601-1-8: What defines a visual alarm?
craiglab
We make a doppler device for monitoring cerebral blood flow. We do not have any alarms - audible or visual - but we do have a pop-up at power up that advises of some precautions, like not to exceed a certain (ultrasound) power in a certain application, and to always use the lowest amount of power (ALARA). The user has to cancel the pop-up to continue.
Would this be considered a visual alarm? It only occurs the one time.
If when adjusting a parameter (like power level) we decided to pop-up a similar message, is that an alarm?
Peter Selvey
It's a good question.
The definitions of "alarm system" and "alarm condition" in IEC 60601-1-8 are not really tight enough to exclude an information message, since it is still to indicate a potential hazard.
It appears the scope statement (Clause 1.1) that the decision to use an "alarm system" is up to the manufacturer. Some particular standards specify that an alarm must be provided, also there are some MDD essential requirements that specifically reference alarms. In all other cases the risk management process will decide if an alarm is needed.
Just for additional justification, although nowhere stated, the principle of IEC 60601-1-8 alarms is to bring the users attention to the equipment generating the alarm, in an environment where there may be many medical devices or where the user may have their attention on other things.
In your case the user is already siting in front of the device, looking at the screen, and has just made a change in setting that might trigger the message. Thus, there is no need to try and grab the user's attention to your device, such as by providing a lamp, audible sound etc.
So perhaps an improved definition of an "alarm system" in IEC 60601-1-8 would indicate that such systems are specifically intended to get the user's attention from a distance, through the use of audible and visual signals.
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601-1 and IEC 60601-2-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3