IEC/EN 60601-1 Certificate required before CE Marking?
LievenDW
I am not an IEC/EN 60601 expert. I do know medical devices must comply with EN 60601-1 (and other harmonized standards) and that a conformity assessment is required before affixing the CE mark. But can a manufacturer rely on their own test results and documentation to claim compliance with EN 60601-1, or do they need a certificate or report from a certification body before affixing the CE mark?
Marcelo
Re: 60601-1 certificate required before CE marking?
NO they don't, compliance to standards is voluntary (although it's the easiest way to show compliance with related essential requirements - these are mandatory)
You can use your own results and documentation, although a third party test house or certification may be viewed by some as more "reliable".
yodon
Just to add on a bit... that 3rd party test house needs to be an accredited lab or else the reviewer may not accept the conclusions. If you do submit your own data, expect a more lengthy review process. That's been what I've seen.
Peter Selvey
For all standards (not just IEC 60601-1) the following minimum requirements can be derived from the law (MDD):
- a report showing compliance with the standard, or pointing to alternate solutions for each and every requirement in a harmonized standard not applied;
- the report should be prepared by person(s) that are appropriately qualified and experience in the scope of tests
- the reports should show results of tests, calculations etc (in sufficient detail to to be audited by the notified body)
- there should be records of traceability for any equipment used to make measurements
It is not necessary for a third party lab or certified report, and it is very common not to use a lab, especially for the less popular standards.
For popular standards like IEC 60601-1 it is usually efficient to use a third party lab as they have the experience, equipment and are set up for the reporting. However, IEC 60601-1 these days is very broad and includes subjective areas. It can be more efficient to limit the test lab to objective, well established tests and exclude parts of the standard which are highly subjective (e.g. PEMS). The excluded parts then need to be covered by report(s) prepared by the manufacturer, and for completeness there should be a cover report which ties the sub-reports together and provides an overall statement of compliance with the standard.
Z CB Certificate and IEC 60601-1 Test Report requirements - Korea Product Registration Other Medical Device Regulations World-Wide 6
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 1
F Battery overload in IEC 60601-1 vs 62133-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Mandatory tests for IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
E Clause 8.9.1.2 of AMD2:2020 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P IEC 60601-1-2 RF wireless technology IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-2-8 cl.201.10.1.2.101 and cl.201.10.1.2.104 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G Risk Management for IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
J IEC 60601 Applied Part - Probes intended to be covered by sheaths IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S IEC 60601-1-11 Requirements related to Storage and Distribution IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
shimonv IEC 60601-1-2:2014+A1:2020 (Ed. 4.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
G IEC 60601 Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W IEC 60601-1 clause 8.8.3 - Dielectric Strength IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
B IEC 60601-1 Clause 14.8 - Architecture IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R IEC 60601-2-57 light source classification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
L Top opening of enclsoure against fire in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C EN 45502 vs IEC 60601-1-8 Other Medical Device Related Standards 0
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2