IEC/EN 60601 Testing and CE Marking for a Non Medical Device

Eamon

Involved In Discussions
#1
Hello.

I am working at a medical device company.

We are planning to release some products similar to some of our existing medical products, for non-medical use (i.e. for research and education). These products are a set of physiological (mostly electrophysiological) sensors, plus a module that passes the signal from these sensors while isolating them from the (potentially mains-connected) data acquisition system.

The main purpose and the initial motivation for non-medical labelling is to simplify things in the US market, where it's hard to imagine how to pass the device through 510(k) since it has no medical purpose in and of itself. The US market also has rules that ostensibly require some of these kinds of equipment to be sold only under the authority of a health care practitioner, a restriction that is not appropriate for this market.

Since the product in question has electrical connection to a live subject (applied part) and since it includes isolation between the applied part and the mains connected part, we think the most relevant standard to apply is IEC 60601-1. This is as opposed to IEC 61010, which might be another possibility. However, the device is engineered to 60601-1 standards, and if for no other reason than reassuring customers, we would like to show testing to 60601-1. Also, there are competing products which show testing to 60601-1 but are not marketed as medical products.

I would be interested in any helpful comments on this situation, but my immediate questions pertain to CE marking:
  • If it is tested to IEC or EN 60601-1, and if the text explaining the CE mark specifies this, does this necessarily imply that it is considered a medical device and therefore falls under MDD?
  • If it doesn't have to fall under MDD, what European regulatory scheme would it fall under (having dealt with nothing but medical devices, I know nothing about non-medical CE marking)?
  • What else regarding CE marking might I need to look into, in this situation?
Thanks in advance to anyone who can help me with answers, or give me pointers on where to look for them.

Eamon Egan
 
Elsmar Forum Sponsor

Marcelo

Inactive Registered Visitor
#2
If it is tested to IEC or EN 60601-1, and if the text explaining the CE mark specifies this, does this necessarily imply that it is considered a medical device and therefore falls under MDD?
The way that the new approach directives such as the MDD are designed is to rely on the information supplied by the manufacturer regarding the products.

So, MDD is applicable to medical devices, and

medical device means any instrument, apparatus, appliance, software, material or other article, whether used alone or in combination, including the software intended by its manufacturer to be used specifically for diagnostic and/or therapeutic purposes and necessary for its proper application, intended by the manufacturer.....
So, in principle, if you do not intend your device to be used for medical purposes, yor device does not fall under the MDD.

This same "directive design" has a built-in function of using standards to provide conformance. Although harmonized standards can be used to provide presumption of conformity, in fact you can use any standard that you see fit (but they do not give presumption so you need some kind of justification).

This is to say that being in conformance with a standard does not mean the product fall under a directive.....so no, there´s no implication here....in truth, it´s the other way, equipment under a directive CAN be in conformance with a particular standard harmonized to this directive.

This is all true from a theoretical point of view, but keep please in mind that interpretations on this could vary.

If it doesn't have to fall under MDD, what European regulatory scheme would it fall under (having dealt with nothing but medical devices, I know nothing about non-medical CE marking)?
You need to take a look at the directives of the EU, in lists such as the one from the New Approach website and the EU website (more complete).

Things for a fresh start: EMC directive, RoHS, WEEE, General product safety directive, Low Voltage directive

What else regarding CE marking might I need to look into, in this situation?
I think it would be best to have an official position, for example, from a notified body or the regulatory authority of your country, due to this being an unusual situation.
 
Last edited:
T

TamTom

#3
Hello,

CE marking is independ from medical devices.

Every "machine" whit "moving" parts, must have such a marking.

I attached, the EU legal text for this, it is not easy to read.

Regards,

TamTom
 

Attachments

chris1price

Trusted Information Resource
#4
Hi

We are in a similar position of having a "Medical Device" which is not covered by the MDD. I would suggest the EMC and LVD directives as a start. Check the Machinery Directive is applicable as well.

Chris
 

Eamon

Involved In Discussions
#5
Thanks to all who replied. I will read through the LV and and EMC directives to see how they are structured, and to be sure I understand how 60601 compliance can generate a presumption of conformity to these directives (versus the MDD).

- Eamon Egan
 

Marcelo

Inactive Registered Visitor
#6
I understand how 60601 compliance can generate a presumption of conformity to these directives
IEC 60601 cannot give presumption of conformity to those directives as it´s not harmonized under them.

But that does not mean you cannot use it to show conformity to the essential principles of those directives.
 
Last edited:

Eamon

Involved In Discussions
#7
IEC 60601 cannot give presumption of conformity to those directives as it´s not harmonized under them.

But that does not mean you cannot use it to shoe conformity to the essential principles of those directives.
Thank you for this distinction.

What I take this to mean is that the Technical File should contain some justificatory language as to why our 60601 test results results imply conformity with LVD and EMC. For example, list the different areas, note that the requirements are at least as stringent...

Does this sound like a reasonable approach?

- Eamon
 

Marcelo

Inactive Registered Visitor
#8
What I take this to mean is that the Technical File should contain some justificatory language as to why our 60601 test results results imply conformity with LVD and EMC. For example, list the different areas, note that the requirements are at least as stringent...

Does this sound like a reasonable approach?
Yes, that´s what i think you need to do. Note that another good justification is that, even not being intented to be used as a medical device, IEC 60601 seems to be the closest standard to use in terms of engineering design (at least from what you told us)...and surely is more stringest in general than other electrical devices standards.
 
S

schwatzi

#9
Hi I was wanting to know if you knew any websites that could help me, I was wanting to set up IEC 60601 testing and wanted to know the equipment required for leakage current measurements and HV tests etc, I do have other equipment as I work in a Test lab.
 
Thread starter Similar threads Forum Replies Date
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Testing lab in India IEC 60601-2-25 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S CTS/CSE Waveforms Testing - Partial Recognition of IEC 60601-2-25 Standard by FDA IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K Is Compliance Testing of IEC 60601-2-63 mandatory for Europe - Dental X-Ray machine CE Marking (Conformité Européene) / CB Scheme 3
L IEC 60601-1-11 IP22 Compliance - Ingress protection testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S IEC 60601-1-2 Radiated E-Field Emission Testing - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S IEC 60601-1 Electrical safety testing - Dielectric Strength Test - 8.8.3 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1 Electrical Safety testing - Class 1 ME externally powered - Earth resistance testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Testing my ECG device as per IEC 60601-2-25 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
O EMC testing per IEC 60601-1-2 for considering new cables CE Marking (Conformité Européene) / CB Scheme 1
O EMC testing under RED and IEC 60601-1-2 CE Marking (Conformité Européene) / CB Scheme 1
D IEC 60601-2-44: 202.101 Immunity Testing of Essential Performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
K IEC 60601-1 Testing for Multiple Socket Outlet (MSO) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601-1-2 Testing Laboratories in the Middle East IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J IEC 60601 Medical Device Testing House Costs IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
T Testing Laboratories in Europe to IEC/EN 60601-2-47 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
M Parallel processing IEC 60601 testing and 510k submission 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D Radiated Immunity Testing as per IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Reducing cost of IEC 60601 3rd edition testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
T How get prepared for IEC 60601 series testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W Is IEC 60601-1-2 compliance/testing required for battery powered Class I Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-1 Cl. 14 - Report changes in Software after Testing to Test Lab? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
T How to fulfill IEC 60601-1-2 EMC Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S IEC 60601-1-2 EMC Testing Optional in US? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
Q Sample Size justification for IEC 60601-1 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 60601-1 testing for device with its own 2V battery power supply necessary? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
J European Laboratories which can do IEC/EN 60601-2-50 Testing? (Close to Turkey) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K Identifying Required Testing to comply with IEC 60601 EU Medical Device Regulations 4
S Are you struggling with your IEC 60601-1 3rd edition testing lab? IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
S Issues with Dielectric Strength Testing - IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
J "Testing procedure: TMP", "WMT", "SMT" and "RMT" - IEC 60601-1 TRF nomenclature IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Current Limit in IEC 60601 Dielectric Strength Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601 Testing - Section 13.2.7 - Impairment of Cooling IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
M Brazilian Requirements for Medical Device Production Testing to IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S EN (IEC) 60601-1 ? Is in-house testing definitely OK? IEC 60601 - Medical Electrical Equipment Safety Standards Series 28
A Do HF accessories require EMC testing according to IEC 60601-1-2? IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
D Battery Powered Disposable Pump - What IEC 60601 testing is required, if any? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Does a 12V DC device need EMC (IEC 60601-1-2) Testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B IEC 60601 3rd Edition - Test Report Form UL uses for testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
N Insulation Barrier Testing - According IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 22
J Clarification in UL Mark in Critical Component - IEC 60601 testing for CE mark in UL IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601-1 Testing required for low-voltage devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Predicate Device - Question on Biocompatibility and IEC 60601-1 Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
D Excessive Temperature Testing to IEC 60601-2-37 - Internal Probes IEC 60601 - Medical Electrical Equipment Safety Standards Series 20
D IEC 60601-2-37 - No Dielectric testing specifications? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3

Similar threads

Top Bottom