Hi all,
EN 61326:2013 specifies dated standards for test protocols (e.g. IEC 61000-3-2:2005 - Limits for harmonic current emissions), but my test house has used newer standards (e.g. EN 61000-3-2:2014). They state "This standard specifies test methods by dated reference; however, on this project updated versions of the test methods may have been applied to maintain current with best EMC practices."
I understand their reasoning, but has any ever had any problems with this as it is technically a non-conformance to the standard?
Thanks all.
CharlieUK
It's quite common - that way the lab only has to be concerned with one actual "test" standard (EN61000-3-x and EN61000-4-x series)
However....since the James Elliott case the European Commission has stated that undated referenced standards are no longer acceptable - because they allow the underlying requirement of a given Harmonised Standard to be changed without review of the Harmonised Standard - and now Harmonised Standards are essentially legal documents, you cannot have legal certainty from a document that can be changed (because the referenced basic standard has changed)
It might not matter, but if you are concerned, then you should instruct the lab to use the dated references in the Harmonised Standard - you are the customer afterall !!
Hi Charlie,
Thanks for the advice; I also got some feedback from UKAS to advise that they could see no issue with using the latest version of a testing standard to maintain state of the art.
CharlieUK
It could be argued that it maintains "state of the art" , but the European Commission are not citing Harmonised Standards without dated references - as Harmonised Standards must be "fully applied" to get a presumption of conformity, you should be careful when deviating from a standard that includes dated references
Also, remember that UKAS are accrediting the ability of the lab to test against certain standards and how to address deviations from them - they do NOT assess or accredit CE marking
C IEC 61326 Functional Performance CE Marking (Conformité Européene) / CB Scheme 1
B IEC 61010-1 - Voltage / Dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M When is EN 62133-1 (new version i.e. IEC 62133:2017) mandatory in EU Other ISO and International Standards and European Regulations 0
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
adir88 Amendment 1 of IEC 62366-1 - Details yet? IEC 62366 - Medical Device Usability Engineering 0
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
K Primary Operating Functions (POF) in IEC 62366-1:2015. IEC 62366 - Medical Device Usability Engineering 5
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 62304 Class A Project IEC 62304 - Medical Device Software Life Cycle Processes 15
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 4
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
MDD_QNA QR Code Standard ISO/IEC 15417:2007 - Does anyone use it? Other Medical Device Related Standards 3
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Universal Power Cord with Adapters - IVD IEC 61010-2-101 CE Marking (Conformité Européene) / CB Scheme 6
V Applicability of IEC 62366-1 usability to resorbable bone substitutes IEC 62366 - Medical Device Usability Engineering 9
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 3
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
A IEC 60601-1 and IEC 60601-2-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601-1-2 EMC Tables IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Bookmarking my security protected IEC 60601-1 .pdf file IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
P IEC 62304 - evaluation of integration and system testing IEC 62304 - Medical Device Software Life Cycle Processes 4
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P Risk acceptability alignment between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 6
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
L IEC 62368-1 / Electrical Energy Source vs. Power Source? CE Marking (Conformité Européene) / CB Scheme 2
D Required Checklist Showing Compliance to IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 11
E ECG device without ECG REPORTS - IEC 60601-2-25: 2011 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C IEC 60601-1-8, difference between table 4 and annex D IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Scope of application for IEC 60601-1-11 Medical electrical equipment — Part 1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601 7.1.3 Durability of Markings - Methylated Spirit or Ethanol? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C ABS plastic : which IEC 60601-1 tests to redo? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Q Isolation/Insulation Diagrams - IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1