Hi all,
The IEC60601-1-11 home healthcare standard states that a solid earth connection cannot be assumed in a home setting. Therefore, earth may not be used as a protective means.
Does this imply that EMC requirements must be met if the earth connection is absent? (We typically use earth for EMI mitigation.)
How about patient leakage current? If the earth connection is absent, is this considered a Normal Condition or Single Fault Condition? Typically, open earth is SFC for a device with a functional earth, but I'm not sure if this still applies for home healthcare, since they are saying we can't assume a solid earth connection.
Thanks,
Tim
Benjamin Weber
60601-1-11 does not allow class I equipment, so you shall not use the PE a means of protection. So the earth conductor can only be functional (FE) in a class II equipment. Interruption of the FE during leakage currwnt testing is NORMAL CONDITION in this case.
During EMC testing it depends, e.g. if your equipment has Essential Performance. If yes, you should consider FE interruption as normal condition, and test if your equipment still complies during immunity testing. SFC testing scenarios should be a result from risk management.
lauxt
Thank you Benjamin! This is certainly an interesting discussion.
I think my confusion is related to the idea of MOPP and how that relates to leakage current. Being relatively new to the field, I think of MOPP as the actual insulation system used to prevent transformer/isolation boundary breakdown. The leakage current concept seems separate from that.
Our "functional" earth connections actually help to reduce patient leakage current...so it almost seems like we should call it a "protective" earth connection, in that case. Is that true?
Just to make sure I'm not going insane, in a hospital/non-home healthcare setting, open earth is SINGLE FAULT CONDITION, right?
Our product is an external switch-mode power supply brick. As I mentioned, I'm new to this, without much guidance. My first estimation is that our product does not have Essential Performance, but I'm not sure
Thanks,
Tim
Benjamin Weber
If you define the earth conductor as PE, you end up with class I equipment. No way out. Definiton of class I: Protection against electrical shock depends on the protective earth and maybe insulation. Class II: Protection against electrical shock depends on reinforced or double insulation only.
The insulation concept assures that:
- short term / transient overvoltage do not result in an electrical shock
AND
- long term exposition / standard use does not result in an electrical shock.
To achieve both, you have to implement the required clearances, creepages and dielectric strengths (MOOP, MOPP) and you have to pass the relevant leakage current test. But you cannot say, that creepage, clearance and dielectric strength is only important for transient overvoltages. This will certainly help you, to be below the leakage current limits. I have seen a lot of devices, having the required creepages, clearances and dielectric strengths, but still failing the leakage currents tests (for example due to Y-caps not selected properly).
Your device is not an actual ME equipment. But if you want to market it to be combined with ME equipment, you should pass the relevant requirements of 60601-1 / -1-11. Patient leakage current is not really aplicable here, but I have seen other manufacturers of power supplys, who make the patient leakage current test between secondary out and mains in. And if you are going to sell it to ME equipment manufacturers who want to comply with -1-11, you should try to pass patient leakage current tests without the earth. FE interruption is considered normal condition during leakage current testing.
And yes you are right, in "normal" hospital / professional setting PE interruption is single fault condition.
Regarding essential performance: I don't think that a power supply like yours has EP. It does not have a clinical function, were the degradation or loss of that functions leads to an unacceptable risk. Take care to not confuse essential performance with basic safety!
EP: Your device does not do something, that it actually should do (ventilate the patient, give an accurate reading, give an alarm)!
BS: Your c´device does something, that it actually should not do (shick you, start to burn, crush a finger).
Regarding EMC testing: As I have written in my first post, you should apply your risk management process here. 60601-1-2:2014 explicitely requires you to do this! You have to evaluate, if there are EMC related problems, that could lead to an unacceptable risk. But to be honest, I am not sure, if SFC condition have to be tested during EMC and if FE interruption is an "official" normal condition to be tested?!? Maybe an EMC expert could help here!?!
lauxt
AND
Thanks again, Benjamin. It's interesting because the company I work for defines any power supply with an earth connection that is not tied directly to the output as Class II with FE. Maybe that's true for a hospital-use only power supply.
However, in the case of 60601-1-11, since this earth connection actually reduces the patient leakage current, it sounds like it's actually a protective earth. Typically, when I think of Class I PE, it's with the classic example:
-High voltage wire comes loose within chassis
-Wire comes in contact with earth referenced chassis
-Chassis shunts current to earth, blows fuse and protects
As you describe above, 2x MOPP is to ensure long term exposure (steady state leakage current, for example) does not cause physiological harm/electric shock. There's more to the story than a wire simply coming loose and causing a one-time dangerous situation!
Thank you for the explanation of essential performance.
Regarding EMC, I think we'll start by seeing how close/far we are to passing EMC with FE interrupted. We are planning to build these for a large medical device company, so maybe they can help us here, specifically with their acceptable risk. Any other insights greatly appreciated!
Thanks for help. It makes everything a little less scary! :)
Jaydub
For EMC tests, SFCs are not usually considered, unless Risk Analysis indicates they should be. It would be pretty unusual to perform EMC tests with an open PE or FE.
lauxt
Jaydub, thank you for that information.
Here is a functional earth configuration we ship on many of our power supplies. These are the two capacitors across the isolation boundary between the primary and secondary.
Benjamin Weber
In case you are not already aware: Regarding Y-caps bridging isolation barriers there are explicit requirements concerning component safety standard compliance of these Y-caps (IEC 60384-14 compliance is required, see IEC 60601-1, cl. 8.5.1.2 and 8.5.1.3)! And there is an very important note to the necessary number and type of Y-caps being in series to achieve the required dielectric strength (IEC 60601-1, Annex A, Note to subclause 8.5.1).
A IEC60601-1 - Battery Charger Requirement - Home Healthcare Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S IEC60601-1 Training recommendations Training - Internal, External, Online and Distance Learning 1
L Allowable AC mains fuse types for latest IEC60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC60601-1: Isolation for an applied part with working voltage of 1500Vp. IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
J IEC60601-2-18 - Endoscopic Surgical Monitor System Images IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N New EMC IEC60601-1-2 4.0 Ed. Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
B Figure A.20 of IEC60601-1 about definition of "Hand-held" & "Portable" IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K IEC60601 3rd Edition Insulation Thickness for electrosurgical devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Need for IEC60601-1 2nd edition certification? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
X IEC60601 Battery powered devices intended for measuring bio-impedance IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC60601-1 Battery Powered Body Worn Device Leakage Currents IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D Where to buy IEC60601-1 Symbols IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
L IEC60601-2-44 - Indicator Light Requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C Do all Class 1 Medical Devices (Electrical) have to comply with IEC60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
M Can I continue using IEC60601-1 Ed2 PSU for MOOP (Means Of Operator Protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
N High Integrity Components Definition according to clause 4.9 of IEC60601-1 3rd ED IEC 60601 - Medical Electrical Equipment Safety Standards Series 43
shimonv When to do the IEC60601-1 Safety Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
S OSHA and IEC61010 and IEC60601-1 - Varied Requirements Occupational Health & Safety Management Standards 1
J Which kind of medical equipment must comply with IEC60601-1-8? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J About the Intelligent Alarm System in IEC60601-1-8 (Anaesthetic Systems) IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
N Overhaul of a 20 mm threshold -3rd Ed of IEC60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N Number of changes in Amendment 1 of the 3rd Ed. of IEC60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
N Software Validation in accordance to IEC60601-1-4 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
A ILAC explanation - IEC60601-1 testing certified by an A2LA laboratory IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A USB Port Isolation as per IEC60601-1 - Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R USB Port (Signal I/O) Isolation as per IEC60601-1 - Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 40
N Ansi/aami/iso13485:2003 + Iso14971:2000 + Iec60601-1-4:2000 + Mdd93/42/eec IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
C IEC60601-1 3rd edition +ve vote - Medical Electrical Equipment Safety Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
O ASQ CQE Exam Experience, Home versus Exam Center ASQ - American Society for Quality 0
qualprod Assign a name to a home-made ERP system? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
R Class 1 Convenience kits for home teeth impression, NOT smile direct or other for alignment purposes Canada Medical Device Regulations 0
C Is it required to put"Rx only" on the home page of an app? Medical Device and FDA Regulations and Standards News 4
J Textiles needed for home monitoring device - any suggestions? Other Medical Device Related Standards 1
K AS9100D 8.4.1.1 external providers question - Walmart, Home Depot, our lawn care team. etc. AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 20
P 510(k) cleared, Rx only and home use 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
Marc Latest Posts in Right Sidebar on the Forum Listing Home Page Forum News and General Information 6
O FDA- CDRH - Design considerations for devices intended for home use US Food and Drug Administration (FDA) 1
D IEC 60601-1-11 - Home Care Device Minimal requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C Pressure requirement of home based Class II Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
Marc Google Home, Siri and Alexa in a Room Together Funny Stuff - Jokes and Humour 3
N Can I Get a Tax Break for Buying my First Home in 2016? Coffee Break and Water Cooler Discussions 5
N Anyone an expert with FHA Home Loans? Coffee Break and Water Cooler Discussions 15
Marc Has anyone here gotten Gigabyte Fiberoptics (Internet) to home? Coffee Break and Water Cooler Discussions 3
Marc BB King in hospice care at home World News 2
S Anyone know the self-selection study for home use medical device Other US Medical Device Regulations 2
T EMC for Medical Devices in Home Environment Other US Medical Device Regulations 4
M EMC Testing for a US-only, Home-Based Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
insect warfare New home for the R2 (Responsible Recycling) standard Miscellaneous Environmental Standards and EMS Related Discussions 2
R Applicable Medical Device Standards for Home Use Blood Pressure Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3