I need your advice and guidance please!
We are designing a medical device battery powered , hand-held, home health care.
The medical device should be charged from stationary battery charger.
the connector pin of stationary battery charger is exposed.
I think medical device should be complied with IEC 60601-1, IEC60601-11
so it should be IP22
Then,
What standard is applied to battery charger?
IEC 60601-1, IEC 60601-1-11?
I would like to know IP class requirement for the stationary battery charger.
I would appreciate very much if anyone could help me!
Mark Meer
The battery-charger is (presumably) non-transit-operable. Therefore, by IEC60601-1-11 (presuming the device is charged in the home environment), the ingress protection requirement is IPX1 (i.e. dripping water shall have no harmful effect).
You should (if you don't already) have documented in your risk analysis consideration for ingress into the charger. If this is the case, I would presume you've already got a fuse or some other control to mitigate against potential shorts or electrical hazards due to fluid ingress.
Bottom line:
- Non-transit-operable (and not body-worn or hand-held): IP21
- Risk management & controls supports "no harmful effect" due to ingress
X IEC60601 Battery powered devices intended for measuring bio-impedance IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M IEC60601-1 Battery Powered Body Worn Device Leakage Currents IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
Z How to interpret and apply IEC60601-2-54, clause 203.6.4.3.104.3 Accuracy of X-RAY TUBE VOLTAGE, second para. IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
O Insulation Diagram IEC60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C IEC60601-1: Impedance and current-carrying capability (8.6.4) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C IEC60601-1 and Laptop in OR IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J IEC60601-1: HiPot Test Method for Internally Powered Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
H About IEC60601 EU Medical Device Regulations 0
C IEC60601-1-2 :2020 EMC Reliability Analysis - Predictions, Testing and Standards 2
W Classes and Applied Part IEC60601 Electromedical Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S IEC60601-1 Training recommendations Training - Internal, External, Online and Distance Learning 1
L IEC60601-1-11 Home Healthcare + IEC60601-1-2 EMC (Functional Earth allowed?) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
L Allowable AC mains fuse types for latest IEC60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC60601-1: Isolation for an applied part with working voltage of 1500Vp. IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
J IEC60601-2-18 - Endoscopic Surgical Monitor System Images IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N New EMC IEC60601-1-2 4.0 Ed. Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
B Figure A.20 of IEC60601-1 about definition of "Hand-held" & "Portable" IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K IEC60601 3rd Edition Insulation Thickness for electrosurgical devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Need for IEC60601-1 2nd edition certification? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Where to buy IEC60601-1 Symbols IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
L IEC60601-2-44 - Indicator Light Requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C Do all Class 1 Medical Devices (Electrical) have to comply with IEC60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
M Can I continue using IEC60601-1 Ed2 PSU for MOOP (Means Of Operator Protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
N High Integrity Components Definition according to clause 4.9 of IEC60601-1 3rd ED IEC 60601 - Medical Electrical Equipment Safety Standards Series 43
shimonv When to do the IEC60601-1 Safety Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
S OSHA and IEC61010 and IEC60601-1 - Varied Requirements Occupational Health & Safety Management Standards 1
J Which kind of medical equipment must comply with IEC60601-1-8? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J About the Intelligent Alarm System in IEC60601-1-8 (Anaesthetic Systems) IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
N Overhaul of a 20 mm threshold -3rd Ed of IEC60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N Number of changes in Amendment 1 of the 3rd Ed. of IEC60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
N Software Validation in accordance to IEC60601-1-4 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
A ILAC explanation - IEC60601-1 testing certified by an A2LA laboratory IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A USB Port Isolation as per IEC60601-1 - Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R USB Port (Signal I/O) Isolation as per IEC60601-1 - Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 40
N Ansi/aami/iso13485:2003 + Iso14971:2000 + Iec60601-1-4:2000 + Mdd93/42/eec IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
C IEC60601-1 3rd edition +ve vote - Medical Electrical Equipment Safety Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
K Leakage current tests for battery powered device with IEC 62368 compliant charger IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H Battery Labeling Requirements US Medical Device Regulations 0
X Separation of battery powered FB type applied part IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
X Insulation between opposite terminals of battery IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
F Interpretation of 60601-1 Figure J.5 (Isolation, example 5) for a rechargeable, battery operated device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L Single-cell lithium ion battery regulations IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M MOPs of power socket for battery powered device IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
F Battery overload in IEC 60601-1 vs 62133-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
G User Ejecting Lithium Battery from Device Other Medical Device and Orthopedic Related Topics 4
I Lithium Battery removal ISO 14971 - Medical Device Risk Management 1
F Lithium polymer battery, venting and CE Marking CE Marking (Conformité Européene) / CB Scheme 4
D LiPo battery minimum requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Z Is Op Amp Common-Mode Rejection Ratio (CMRR) test applicable to battery operated devices? Other Medical Device Related Standards 0
L EN 62368 (In an Annex - M.4.2.1) Secondary Lithium Battery Charging Safeguards CE Marking (Conformité Européene) / CB Scheme 2