IEC60601-1: Impedance and current-carrying capability (8.6.4)

cjlark

Registered
I am working with a medical drill system consisting of two main subsystems.

1.The Footswitch which receives AC power from a wall outlet and through a Certified internal power supply converts the AC power to DC power (36V)
2.The handpiece (drill) which connects to the Footswitch and receives the DC power signals to commutate its motor.

While at out 60601 test house we have been told that any accessible metal parts on the handpiece need to pass the testing laid out in 8.6.4a. To me, this doesn't make sense. I can understand why the Footswitch would need to pass this test as there is a reasonable scenario where Mains voltage could come in contact with its metal enclosure. I do not see how this could possibly happen on the handpiece since it is downstream from the AC Mains voltage and only receives DC voltage outputted from the certified internal power supply.

Could someone help me understand why our Handpiece needs to meet this requirement? Is the rationale that it doesn't matter how the handpiece gets energized? We have run some tests on very similar predicate devices and they are not able to meet the 100mOhm requirement. I am wondering why.

Thank you.
 

Peter Selvey

Leader
Super Moderator
According to the standard the test is not applicable.

Clause 8.6 is for protective earthing which if the definitions are read closely are intended for low impedance to the earth of the mains supply for protective purpose. Any part which is 2MOPP/2MOOP away from mains parts is out of scope for 8.6.

However, it may be that connection of the frame to the 0V line of the 36V circuit is used for safety, for example if the motor circuit shorts to the frame of the handpiece, this is picked up by overcurrent protection and the output shuts down with no safety issues. This is similar to protective earthing and tested in a similar way, by inspecting the cable AWG, checking for reliable connections and if needed testing with a current that represents the maximum current of the power supply plus a safety margin. The main difference is in the current used for analysis/testing. For example if the power supply limits the current to 2A, verify the wiring/connections/test are based on say 5A rather than 25A.

Also, it should be noted that in the final installation, there may be installation and periodic tests. Sometimes it is easier to connect everything metal to PE terminal with 16AWG wiring just in case, since the person performing the test won't know the internal construction and whether it needs to be protectively earthed or not. Otherwise, the instructions for use/installation need to clearly state where the protective earth test needs to be done.
 
Thread starter Similar threads Forum Replies Date
X IEC60601 Battery powered devices intended for measuring bio-impedance IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
Z How to interpret and apply IEC60601-2-54, clause 203.6.4.3.104.3 Accuracy of X-RAY TUBE VOLTAGE, second para. IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
O Insulation Diagram IEC60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C IEC60601-1 and Laptop in OR IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J IEC60601-1: HiPot Test Method for Internally Powered Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
H About IEC60601 EU Medical Device Regulations 0
C IEC60601-1-2 :2020 EMC Reliability Analysis - Predictions, Testing and Standards 2
W Classes and Applied Part IEC60601 Electromedical Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S IEC60601-1 Training recommendations Training - Internal, External, Online and Distance Learning 1
L IEC60601-1-11 Home Healthcare + IEC60601-1-2 EMC (Functional Earth allowed?) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
L Allowable AC mains fuse types for latest IEC60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC60601-1: Isolation for an applied part with working voltage of 1500Vp. IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
J IEC60601-2-18 - Endoscopic Surgical Monitor System Images IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N New EMC IEC60601-1-2 4.0 Ed. Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
B Figure A.20 of IEC60601-1 about definition of "Hand-held" & "Portable" IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC60601-1 - Battery Charger Requirement - Home Healthcare Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K IEC60601 3rd Edition Insulation Thickness for electrosurgical devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Need for IEC60601-1 2nd edition certification? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC60601-1 Battery Powered Body Worn Device Leakage Currents IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D Where to buy IEC60601-1 Symbols IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
L IEC60601-2-44 - Indicator Light Requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C Do all Class 1 Medical Devices (Electrical) have to comply with IEC60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
M Can I continue using IEC60601-1 Ed2 PSU for MOOP (Means Of Operator Protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
N High Integrity Components Definition according to clause 4.9 of IEC60601-1 3rd ED IEC 60601 - Medical Electrical Equipment Safety Standards Series 43
shimonv When to do the IEC60601-1 Safety Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
S OSHA and IEC61010 and IEC60601-1 - Varied Requirements Occupational Health & Safety Management Standards 1
J Which kind of medical equipment must comply with IEC60601-1-8? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J About the Intelligent Alarm System in IEC60601-1-8 (Anaesthetic Systems) IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
N Overhaul of a 20 mm threshold -3rd Ed of IEC60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N Number of changes in Amendment 1 of the 3rd Ed. of IEC60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
N Software Validation in accordance to IEC60601-1-4 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
A ILAC explanation - IEC60601-1 testing certified by an A2LA laboratory IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A USB Port Isolation as per IEC60601-1 - Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R USB Port (Signal I/O) Isolation as per IEC60601-1 - Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 40
N Ansi/aami/iso13485:2003 + Iso14971:2000 + Iec60601-1-4:2000 + Mdd93/42/eec IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
C IEC60601-1 3rd edition +ve vote - Medical Electrical Equipment Safety Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
B Protective impedance as MOOP/MOPP - Requirements? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Informational USFDA – Medical Devices; Anesthesiology Devices; Classification of the Ventilatory Electrical Impedance Tomograph Medical Device and FDA Regulations and Standards News 0
G IEC 61010 - Single Fault Condition - Protective Impedance Implementation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R What is the meaning of "Highest RATED Current" in Protective Earth Impedance Test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
K Leakage current tests for battery powered device with IEC 62368 compliant charger IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H 60601 - Patient leakage current via the patient connection test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Using current gauge to verify masters General Measurement Device and Calibration Topics 1
K Alternatives to Metal Foil for touch current measurement of enclosures made of insulating material. IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Help Understanding Current Role; Distributor, Manufacturer, or Importer EU Medical Device Regulations 4
B Leakage current measurement without isolating transformer IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B Precision Current Loop for Current Probe Calibration Reliability Analysis - Predictions, Testing and Standards 2
N Auxiliary current limit with coil cell powered device IEC 60601 - Medical Electrical Equipment Safety Standards Series 4

Similar threads

Top Bottom