I am working with a medical drill system consisting of two main subsystems.
1.The Footswitch which receives AC power from a wall outlet and through a Certified internal power supply converts the AC power to DC power (36V)
2.The handpiece (drill) which connects to the Footswitch and receives the DC power signals to commutate its motor.
While at out 60601 test house we have been told that any accessible metal parts on the handpiece need to pass the testing laid out in 8.6.4a. To me, this doesn't make sense. I can understand why the Footswitch would need to pass this test as there is a reasonable scenario where Mains voltage could come in contact with its metal enclosure. I do not see how this could possibly happen on the handpiece since it is downstream from the AC Mains voltage and only receives DC voltage outputted from the certified internal power supply.
Could someone help me understand why our Handpiece needs to meet this requirement? Is the rationale that it doesn't matter how the handpiece gets energized? We have run some tests on very similar predicate devices and they are not able to meet the 100mOhm requirement. I am wondering why.
Thank you.
1.The Footswitch which receives AC power from a wall outlet and through a Certified internal power supply converts the AC power to DC power (36V)
2.The handpiece (drill) which connects to the Footswitch and receives the DC power signals to commutate its motor.
While at out 60601 test house we have been told that any accessible metal parts on the handpiece need to pass the testing laid out in 8.6.4a. To me, this doesn't make sense. I can understand why the Footswitch would need to pass this test as there is a reasonable scenario where Mains voltage could come in contact with its metal enclosure. I do not see how this could possibly happen on the handpiece since it is downstream from the AC Mains voltage and only receives DC voltage outputted from the certified internal power supply.
Could someone help me understand why our Handpiece needs to meet this requirement? Is the rationale that it doesn't matter how the handpiece gets energized? We have run some tests on very similar predicate devices and they are not able to meet the 100mOhm requirement. I am wondering why.
Thank you.