Hi There,

I am a CE newbie looking for some help. I work for a company that manufactures massage/chiropractor/medical tables, and in the past they have used the MDD to certify their products.

We now have a new product that is a nail "desk" (essentially a table that will be in nail salons used for manicures). It has some colored LED lights and a work lamp attached to it - these are both plugged into a power strip that plugs into the wall. The components are all individually CE certified.

My question is, does the Low Voltage Directive apply to this type of product? Or would it be only the General Product Safety Directive?

I'm assuming part of this will be somehow certifying that all the electrical components are safe when connected together, but is there also a standard that will require the desk itself has no safety issues?

I appreciate any help you can provide! Thank you!
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#3
Re: If/how to apply LVD in this case?

Hello and welcome to The Cove :bigwave:

Why do you think that the LVD doesn't apply?

Article 1
For the purposes of this Directive, ‘electrical equipment’ means any equipment designed for use with a voltage rating of between 50 and 1 000 V for alternating current and between 75 and 1 500 V for direct current, other than the equipment and phenomena listed in Annex II.
ANNEX II
Equipment and Phenomena outside the Scope of this Directive
Electrical equipment for use in an explosive atmosphere
Electrical equipment for radiology and medical purposes
Electrical parts for goods and passenger lifts
Electricity meters
Plugs and socket outlets for domestic use
Electric fence controllers
Radio-electrical interference
Specialised electrical equipment, for use on ships, aircraft or railways, which complies with the safety provisions drawn up
by international bodies in which the Member States participate.
The current LVD is in force until 20 April 2016 (a recast was approved in 2014).
 

CharlieUK

Quite Involved in Discussions
#4
Re: If/how to apply LVD in this case?

A few word of caution

The components are all individually CE certified.

Unfortunately, there is no such thing.. Items may be CE marked, but that is a Declaration by the manufacturer, that is all.
There are plenty of non-compliant products that have CE marks on them - weekly reports of market surveillance available http://ec.europa.eu/consumers/safety/rapex/alerts/main/index.cfm?event=main.listNotifications

As a minimum, I would recommend that you:
- obtain copies of the Declarations of Conformity for any 3rd party items you integrated into your product.
- where possible, source items that have 3rd party approvals marks (UL, TUV, VDE, Nemko, CSA etc.) AND verify on the certification bodies website that the mark is valid
- have someone competent perform an evaluation of a sample of the product

By way of example, I recently broke open a CE marked mains-USB charger and found that internal clearances were < 1mm and not >4mm as required by the standard

Charlie
 
J

jenlianne146

#5
Thanks for your responses.

Follow up question - so if a component (for example, LED lighting) has a CE mark, and the declaration of conformity says it is conforming to a certain directive and standard, do I ALSO have to do the testing to confirm that it conforms before I can include it on my product that I'm trying to CE mark?

If I have a component that is NOT CE marked that I want to use, is that even possible? Even if it is UL listed?

Overall, I'm still unsure if I need to use the standards specified for each individual component, or if there is a more general standard that will help me with verifying that the product is safe when everything is connected together. (For example, for UL certification we use UL 962 for commercial furnishings. But I don't see any kind of equivalent for EU standards.)

Thanks so much!! I am really struggling! :nopity:
 

Ronen E

Problem Solver
Staff member
Moderator
#6
Hi,

I'm not an expert on general electrical equipment, but I'd like to comment from the EC regulation perspective:

Follow up question - so if a component (for example, LED lighting) has a CE mark, and the declaration of conformity says it is conforming to a certain directive and standard, do I ALSO have to do the testing to confirm that it conforms before I can include it on my product that I'm trying to CE mark?
As Charlie has pointed out, your approach can be tiered. I would say that the level of scrutiny could be based on risk evaluation (not all components have the same criticality).

1. For low risk, a CE mark with a copy of the DoC from the component manufacturer might suffice. It's not much but is better than nothing.

2. Next risk level up (or when components are not CE marked), have reputable / accreditted 3rd party certifications on file and verify their validity.

3. For high risk / when no 3rd party certification is available (and you still want to use that specific component), test / inspect.

Above all that, I assume you conduct your own testing to ensure that your finished product functions properly and meets applicable requirements. This would provide an additional assurance level (albeit at system level, not component, thus less "sensitive").

If I have a component that is NOT CE marked that I want to use, is that even possible? Even if it is UL listed?
I'm not aware of a generic requirement to use only CE marked components. It is your responsibility to ensure that your finished product is compliant and safe. If you're using non-CE-marked components you have to account for that somehow (you actually have to account for the components you use anyway, to a certain extent, even if they're CE marked).

Cheers,
Ronen.
 

CharlieUK

Quite Involved in Discussions
#7
If I have a component that is NOT CE marked that I want to use, is that even possible? Even if it is UL listed?

Assuming you are using it within stated parameters, that can be better than having it CE marked. A UL listed/recognised product is subject to factory inspections and some checks to ensure ongoing compliance.
You can check that UL mark is current at http://database.ul.com/cgi-bin/XYV/template/LISEXT/1FRAME/index.htm

Charlie
 
Thread starter Similar threads Forum Replies Date
shimonv Worthwhileness for a small startups to apply for EUA (Emergency Use Authorization) for COVID-19 US Food and Drug Administration (FDA) 17
qualprod Records where apply and stop recording everything? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
B Operator protection - When to apply table 7, Dielectric strength test voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B How to apply external voltage to SIP/SOP IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
G Do HIPAA Rules Apply to a 3rd Party Logistics Shipper? Other US Medical Device Regulations 2
D Do purchasing controls apply to non-medical parts? ISO 13485:2016 - Medical Device Quality Management Systems 5
G Does pitch/increment/resolution of a ruled scale apply to measurement uncertainty as line item? Measurement Uncertainty (MU) 10
D Monitoring and measuring resources - Example of how section ISO 9001 7.1.5 would apply to a manufacture of software ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
C Does ITAR apply to Exporters only? Other ISO and International Standards and European Regulations 6
E Does CQI-23 (Molding Sys. Asses.) apply to RUBBER molding? IATF 16949 - Automotive Quality Systems Standard 2
M Informational From RAPS – UK NB Will Not Apply for EU MDR/IVDR Medical Device and FDA Regulations and Standards News 11
B Does EMC Directive 2014/30/EU cease to apply when the Machine Directive 2006/42/EC is applicable? CE Marking (Conformité Européene) / CB Scheme 4
E Discussion between co-worker on tolerance and uncertainty and how to apply it. Thoughts? 17025 ISO 17025 related Discussions 1
B How to assess which directives and standards apply - Brewing and distilling products CE Marking (Conformité Européene) / CB Scheme 1
N Student trying to apply Lean Six Sigma on a Construction project in my thesis Lean in Manufacturing and Service Industries 3
F Device Standards - Managing Changes to existing standards that apply to our devices Other Medical Device Related Standards 1
A ISO 9001:2015 - How to apply 7.1.5 to an organization performing only design ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
F What directives/regulations apply to Analyte Specific Reagents (ASRs) in EU EU Medical Device Regulations 0
F Can an individual apply for an MDEL in Canada, as an establishment? Canada Medical Device Regulations 6
N Apply control limits to a non-normal distribution Statistical Analysis Tools, Techniques and SPC 13
M IATF Clause 8.4.2.2 - What procedures does this apply to? IATF 16949 - Automotive Quality Systems Standard 1
S Optical coherence tomography (OCT) device - Do we need to apply IDE? Other US Medical Device Regulations 0
R Communication of CSR's throughout the organization and who they apply to IATF 16949 - Automotive Quality Systems Standard 4
M Apply for a job when no opening is posted? Career and Occupation Discussions 64
M AS9100 D - How to apply 8.1.1-8.1.4 to a coating service AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
M Does ISO 10993 Apply to Packaging? (e.g. Travel Case) Other Medical Device Related Standards 4
C Optic Patchcord Cables - Does 8.3 (Design) Apply Here? ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
M Does the Scope of 21 CFR Part 820.72 (Equipment) apply to Design? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
D Does IATF 16949:2016 Clause 7.3 (now 8.3) Apply to "Grandfathered" Products IATF 16949 - Automotive Quality Systems Standard 4
Moncia Tall Manufacturer - VDA Audit - Which parts of the audit apply to us? IATF 16949 - Automotive Quality Systems Standard 4
Q How to apply 21 CFR Part 11 and/or cGMP in the Life Sciences industry Qualification and Validation (including 21 CFR Part 11) 3
V When and How to apply the Bayes Success Run approach for Sample Size Determination Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
A 3 scenarios of accident and I want to apply OHSAS18001 Occupational Health & Safety Management Standards 2
I ISO 26262 standard may apply to automotive IC suppliers IATF 16949 - Automotive Quality Systems Standard 1
GStough Looking to Apply for IRCA Certification Using 2nd Party Audits Career and Occupation Discussions 2
M Does AS9120 (aerospace first article inspection) apply to all customers AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
A How soon can I apply for ISO 9001:2015 certification? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
J Who needs to apply for a CLIA waiver 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D Does Record Control apply to the company business processes as well? Records and Data - Quality, Legal and Other Evidence 6
P When can we apply DOE (Design of Experiments)? Six Sigma 8
L AS9100 7.5.2 Special Processes - Clean Room Apply? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
M Not sure what standards apply to non-electrical patient bed EU Medical Device Regulations 6
A Shall I apply for this job position (Quality Manager)? Career and Occupation Discussions 5
V Can we apply bracketing & matrixing simultaneously for stability studies @ ICH Q1D US Food and Drug Administration (FDA) 1
N Which regulations apply to Medical Stand Alone Software in Europe? ISO 13485:2016 - Medical Device Quality Management Systems 3
W Where to apply a Class I Medical Device Barcode in Canada Other Medical Device Related Standards 5
A When to apply Markov Chain Analysis over Weibull Analysis Reliability Analysis - Predictions, Testing and Standards 2
Sean Kelley Do CSR's (Customer Specific Requirements) apply from the customer of your customer? Customer and Company Specific Requirements 1
L Does IEC 61010-1 apply for Surgical Microscopes? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R RoHS for Medical Devices - Does the directive apply to non-EEE accessories? RoHS, REACH, ELV, IMDS and Restricted Substances 1

Similar threads

Top Bottom