We have a "Contract Sterilizer" to gamma irradiate our medical devices. When I asked him for his ISO 13485 certificate, I saw that their BSI certificate mentions the scope to comply per EN ISO 11137-1:2006.

The EN ISO 11137-1 is just one of the 3 parts of the EN ISO 11137 standard and has 3 parts -

Part 1 - Requirements for development, validation and routine control of a sterilization process for medical devices

Part 2 - Establishing the sterilization dose

Part 3 - Guidance on dosimetric aspects

So if the ISO 13485 certificate has the scope of just ISO 11137-1, would that be construed that Parts 2 and 3 are also covered ?

Or would the scope should have been just EN ISO 11137 and not mention the Parts ?

Thank you !
 
Elsmar Forum Sponsor
M

MIREGMGR

#2
Re: If this ISO 13485 certificate acceptable?

It's their job to establish and maintain a capable system...QMS and radiation hardware.

It's your job to design, validate and document your sterilization requirement for your particular product(s). You then will give them your spec, and they'll sterilize to that spec.

They might help you, as subject matter experts, but it's your responsibility.

So, they comply with -1, you comply with -1, -2 and -3.
 
Last edited by a moderator:
Thread starter Similar threads Forum Replies Date
F Irradiation Sterilization Validation Maximum Dose - VDmax 25 method (ISO 11137) EU Medical Device Regulations 2
R ISO 11137 Gamma sterilization validation, dose setting, dose audit Other Medical Device Related Standards 1
I Certification in Sterilization for ISO 11135 and ISO 11137 Professional Certifications and Degrees 3
M ISO 11137 - Verification dose audit for VD max method Other Medical Device Regulations World-Wide 7
J Does anyone have a summary of changes made to ISO 11137-2? Other Medical Device Related Standards 3
F ISO 11137 - Flow Chart | Decision Tree sterilization Other Medical Device Related Standards 4
K Very small batch validation ISO 11137 (VDmax25) ISO 13485:2016 - Medical Device Quality Management Systems 6
S Controlled Environment for Manufacturing Sterile Product - ISO 11137 Other Medical Device Related Standards 2
C Auditing Checklist for ISO 11135 and ISO 11137 (regulations for sterilization) Other Medical Device Related Standards 9
M Radiation (Gamma or E-beam) Dose Audit SOP - Quarterly dose audits based on ISO 11137 Other Medical Device Related Standards 8
E ISO 11137 gamma sterilized device dose audit - Lead time? Other ISO and International Standards and European Regulations 7
M Gamma Sterility Validation - ISO 11137-2, 4.2 vs. FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
C ISO 11137:1995 Sterilisation of healthcare products - Amd1:2001 ISO 13485:2016 - Medical Device Quality Management Systems 5
G ISO 17023 2017-11 - Suggestions for good books ISO 17025 related Discussions 0
B Timeframe for updating QMS / transitioning from ISO 14971:2012 to ISO 14971:2019 ISO 14971 - Medical Device Risk Management 3
M ISO 9001:2015 and AS6081:2012 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
C Implementation ISO 9001: 2015 ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
C ISO/ IEC 17021 Resource requirement (need help) Document Control Systems, Procedures, Forms and Templates 5
J Possible to get ISO 13485 certified with only OEM Product? ISO 13485:2016 - Medical Device Quality Management Systems 4
D ISO 14971:2019 vs MDR Annex 1, Requirement #4 - "Manufacturers shall inform users of any residual risks" ISO 14971 - Medical Device Risk Management 2
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 0
L Biological Assessment (ISO 10993-1) Other Medical Device Related Standards 1
M ISO 13485:2016 Complaint Definition Clarity Customer Complaints 2
eule del ayre Documented Information - Periodic Review of Documents? IATF 16949:2016 / ISO 9001:2015 IATF 16949 - Automotive Quality Systems Standard 34
J ISO 9001:2015, ISO 14001 & OHSAS18000 (IMS) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
G ISO 14001 - 6.1.3 Compliance Obligations ISO 14001:2015 Specific Discussions 1
D Rules for Paper Forms outside of an eQMS - 3 Questions (ISO 13485) Document Control Systems, Procedures, Forms and Templates 9
S Qualification question - ISO 13485 Reliability Analysis - Predictions, Testing and Standards 0
K ISO 13485 clause 8.5.2 'Any necessary CA shall be taken without undue delay' ISO 13485:2016 - Medical Device Quality Management Systems 10
Aymaneh ISO 11607-1: 2019 main changes Other Medical Device Related Standards 2
G National Structural Steel Specification 7th Edition - Do I now have to be audited against ISO 3843-3 as well as ISO 9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
J How much to charge for helping a startup company with initial ISO 13485 certification? Consultants and Consulting 3
J ISO 13485 System 'soft start' - How to best reflect this in initial audits, management review minutes and other records? ISO 13485:2016 - Medical Device Quality Management Systems 3
L How to understand the clause 6 Planning of ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
D ISO 13485 - 7.3.6 Design and development verification - Do most folks create a separate SOP? ISO 13485:2016 - Medical Device Quality Management Systems 4
B ISO 8536-4 Contamination Index ISO 13485:2016 - Medical Device Quality Management Systems 0
S Practical Implementation of ISO 14971 ISO 14971 - Medical Device Risk Management 6
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
L Process changes and biocompatibility (ISO 10993-1) Other Medical Device Related Standards 1
J Recommendations for online ISO 19011 training? Training - Internal, External, Online and Distance Learning 6
D ISO 13485 8.2.1 and 8.2.2 - Customer Feedback and Customer Complaints ISO 13485:2016 - Medical Device Quality Management Systems 5
S Requirements to obtain ISO 50001 Certification ISO 14001:2015 Specific Discussions 2
A ISO 11135:2014, B.1.4, BI resistance x product bioburden ISO 13485:2016 - Medical Device Quality Management Systems 6
Sravan Manchikanti How to interpret '8.3 Control of nonconforming product' for SaMD device while implementing ISO 13485 & MDSAP ISO 13485:2016 - Medical Device Quality Management Systems 4
J Sister-company providing parts is only ISO 9001 registered IATF 16949 - Automotive Quality Systems Standard 7
M Getting started in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 21
G Copy of withdrawn ISO 9001:1994 Quality Management Standard ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
P ISO 80369 and endoscope port/lumen Other ISO and International Standards and European Regulations 4

Similar threads

Top Bottom